Hello Everyone,
I just wanted to give everyone a heads up that I have generated a first set of LUTs for the pfRICH to be handed over to the software group for integration into EICrecon.
There are two tables, one for combined electron, pi, k, p and one for only hadron ID.
The electron+hadron table has a few pathologies where a probability overflowed or reported NAN instead of registering as 0 – these I can fix by hand. I also realize that the electron+hadron table does not split kaon probabilities between kaon and proton for sub-threshold kaons.
For this round, I did not look at different charge signs, omitted B-field and track smearing, and did not bin in Phi. These will be treated in subsequent rounds.
Thanks.
Brian
|
## Eta Bins: [-3.8, -3.5], [-3.5, -3.0], [-3.0, -2.8], [-2.8, -2.0], [-2.0,
-1.5]
## Momentum Bins [GeV]: 0.5 to 1.0 in 0.1 wide steps : then 1.0 to 15.0 in
1.0 wide steps
## NOTE: Particles thrown flat within Eta and Momentum Bins - results
reported at bin center
## NOTE: No charge scan - only thrown pi+, K+, and proton - assume charge
symmetry at this point
## NOTE: Particles thrown flat in Phi in this run - Phi value is a dummy
## Format: PID Charge Momentum Theta Phi prob-pion prob-kaon prob-proton
211 1 0.55 3.0896 0.00 1.000000 0.000000 0.000000
211 1 0.65 3.0896 0.00 1.000000 0.000000 0.000000
211 1 0.75 3.0896 0.00 1.000000 0.000000 0.000000
211 1 0.85 3.0896 0.00 1.000000 0.000000 0.000000
211 1 0.95 3.0896 0.00 1.000000 0.000000 0.000000
211 1 1.50 3.0896 0.00 1.000000 0.000000 0.000000
211 1 2.50 3.0896 0.00 0.913067 0.086933 0.000000
211 1 3.50 3.0896 0.00 0.533790 0.466210 0.000000
211 1 4.50 3.0896 0.00 0.508638 0.491362 0.000000
211 1 5.50 3.0896 0.00 0.512167 0.487833 0.000000
211 1 6.50 3.0896 0.00 0.822850 0.177150 0.000000
211 1 7.50 3.0896 0.00 0.978178 0.021822 0.000000
211 1 8.50 3.0896 0.00 0.934052 0.065948 0.000000
211 1 9.50 3.0896 0.00 0.901361 0.098639 0.000000
211 1 10.50 3.0896 0.00 0.833604 0.166396 0.000000
211 1 11.50 3.0896 0.00 0.792231 0.207769 0.000000
211 1 12.50 3.0896 0.00 0.766686 0.233314 0.000000
211 1 13.50 3.0896 0.00 0.704268 0.295732 0.000000
211 1 14.50 3.0896 0.00 0.694652 0.305348 0.000000
211 1 0.55 3.0641 0.00 1.000000 0.000000 0.000000
211 1 0.65 3.0641 0.00 1.000000 0.000000 0.000000
211 1 0.75 3.0641 0.00 1.000000 0.000000 0.000000
211 1 0.85 3.0641 0.00 1.000000 0.000000 0.000000
211 1 0.95 3.0641 0.00 1.000000 0.000000 0.000000
211 1 1.50 3.0641 0.00 1.000000 0.000000 0.000000
211 1 2.50 3.0641 0.00 0.993040 0.006960 0.000000
211 1 3.50 3.0641 0.00 0.989529 0.010471 0.000000
211 1 4.50 3.0641 0.00 0.668152 0.331848 0.000000
211 1 5.50 3.0641 0.00 0.553053 0.446947 0.000000
211 1 6.50 3.0641 0.00 0.915839 0.084161 0.000000
211 1 7.50 3.0641 0.00 0.971126 0.028874 0.000000
211 1 8.50 3.0641 0.00 0.951838 0.048162 0.000000
211 1 9.50 3.0641 0.00 0.908038 0.091962 0.000000
211 1 10.50 3.0641 0.00 0.862748 0.137252 0.000000
211 1 11.50 3.0641 0.00 0.821518 0.178482 0.000000
211 1 12.50 3.0641 0.00 0.770998 0.229002 0.000000
211 1 13.50 3.0641 0.00 0.743841 0.256159 0.000000
211 1 14.50 3.0641 0.00 0.710247 0.289753 0.000000
211 1 0.55 3.0317 0.00 1.000000 0.000000 0.000000
211 1 0.65 3.0317 0.00 1.000000 0.000000 0.000000
211 1 0.75 3.0317 0.00 1.000000 0.000000 0.000000
211 1 0.85 3.0317 0.00 1.000000 0.000000 0.000000
211 1 0.95 3.0317 0.00 1.000000 0.000000 0.000000
211 1 1.50 3.0317 0.00 1.000000 0.000000 0.000000
211 1 2.50 3.0317 0.00 0.523138 0.476862 0.000000
211 1 3.50 3.0317 0.00 0.652509 0.347491 0.000000
211 1 4.50 3.0317 0.00 0.504645 0.495355 0.000000
211 1 5.50 3.0317 0.00 0.534424 0.465576 0.000000
211 1 6.50 3.0317 0.00 0.799718 0.200282 0.000000
211 1 7.50 3.0317 0.00 0.979515 0.020485 0.000000
211 1 8.50 3.0317 0.00 0.949815 0.050185 0.000000
211 1 9.50 3.0317 0.00 0.894310 0.105690 0.000000
211 1 10.50 3.0317 0.00 0.864103 0.135897 0.000000
211 1 11.50 3.0317 0.00 0.809631 0.190369 0.000000
211 1 12.50 3.0317 0.00 0.770917 0.229083 0.000000
211 1 13.50 3.0317 0.00 0.733919 0.266081 0.000000
211 1 14.50 3.0317 0.00 0.711416 0.288584 0.000000
211 1 0.55 2.9607 0.00 1.000000 0.000000 0.000000
211 1 0.65 2.9607 0.00 1.000000 0.000000 0.000000
211 1 0.75 2.9607 0.00 1.000000 0.000000 0.000000
211 1 0.85 2.9607 0.00 1.000000 0.000000 0.000000
211 1 0.95 2.9607 0.00 1.000000 0.000000 0.000000
211 1 1.50 2.9607 0.00 1.000000 0.000000 0.000000
211 1 2.50 2.9607 0.00 0.535094 0.464906 0.000000
211 1 3.50 2.9607 0.00 0.564560 0.435440 0.000000
211 1 4.50 2.9607 0.00 0.557450 0.442550 0.000000
211 1 5.50 2.9607 0.00 0.532661 0.467339 0.000000
211 1 6.50 2.9607 0.00 0.840121 0.159879 0.000000
211 1 7.50 2.9607 0.00 0.972253 0.027747 0.000000
211 1 8.50 2.9607 0.00 0.934024 0.065976 0.000000
211 1 9.50 2.9607 0.00 0.915990 0.084010 0.000000
211 1 10.50 2.9607 0.00 0.862753 0.137247 0.000000
211 1 11.50 2.9607 0.00 0.811276 0.188724 0.000000
211 1 12.50 2.9607 0.00 0.779138 0.220862 0.000000
211 1 13.50 2.9607 0.00 0.739504 0.260496 0.000000
211 1 14.50 2.9607 0.00 0.706773 0.293227 0.000000
211 1 0.55 2.7975 0.00 1.000000 0.000000 0.000000
211 1 0.65 2.7975 0.00 1.000000 0.000000 0.000000
211 1 0.75 2.7975 0.00 1.000000 0.000000 0.000000
211 1 0.85 2.7975 0.00 1.000000 0.000000 0.000000
211 1 0.95 2.7975 0.00 1.000000 0.000000 0.000000
211 1 1.50 2.7975 0.00 1.000000 0.000000 0.000000
211 1 2.50 2.7975 0.00 0.953423 0.046577 0.000000
211 1 3.50 2.7975 0.00 0.707530 0.292470 0.000000
211 1 4.50 2.7975 0.00 0.519600 0.480400 0.000000
211 1 5.50 2.7975 0.00 0.507280 0.492720 0.000000
211 1 6.50 2.7975 0.00 0.685008 0.314992 0.000000
211 1 7.50 2.7975 0.00 0.948843 0.051157 0.000000
211 1 8.50 2.7975 0.00 0.921132 0.078868 0.000000
211 1 9.50 2.7975 0.00 0.885478 0.114522 0.000000
211 1 10.50 2.7975 0.00 0.828059 0.171941 0.000000
211 1 11.50 2.7975 0.00 0.779961 0.220039 0.000000
211 1 12.50 2.7975 0.00 0.755918 0.244082 0.000000
211 1 13.50 2.7975 0.00 0.730497 0.269503 0.000000
211 1 14.50 2.7975 0.00 0.703692 0.296308 0.000000
321 1 0.55 3.0896 0.00 0.000000 0.500000 0.500000
321 1 0.65 3.0896 0.00 0.000000 0.500000 0.500000
321 1 0.75 3.0896 0.00 0.000000 0.500000 0.500000
321 1 0.85 3.0896 0.00 0.000000 0.500000 0.500000
321 1 0.95 3.0896 0.00 0.000000 0.500000 0.500000
321 1 1.50 3.0896 0.00 0.000000 1.000000 0.000000
321 1 2.50 3.0896 0.00 0.000000 1.000000 0.000000
321 1 3.50 3.0896 0.00 0.105568 0.894432 0.000000
321 1 4.50 3.0896 0.00 0.313310 0.625592 0.061098
321 1 5.50 3.0896 0.00 0.314877 0.577898 0.107225
321 1 6.50 3.0896 0.00 0.027119 0.523884 0.448996
321 1 7.50 3.0896 0.00 0.028167 0.477655 0.494178
321 1 8.50 3.0896 0.00 0.074478 0.433141 0.492381
321 1 9.50 3.0896 0.00 0.124474 0.379123 0.496403
321 1 10.50 3.0896 0.00 0.171563 0.337001 0.491436
321 1 11.50 3.0896 0.00 0.194550 0.431152 0.374298
321 1 12.50 3.0896 0.00 0.260400 0.733195 0.006405
321 1 13.50 3.0896 0.00 0.283041 0.672535 0.044424
321 1 14.50 3.0896 0.00 0.305348 0.584070 0.110583
321 1 0.55 3.0641 0.00 0.000000 0.500000 0.500000
321 1 0.65 3.0641 0.00 0.000000 0.500000 0.500000
321 1 0.75 3.0641 0.00 0.000000 0.500000 0.500000
321 1 0.85 3.0641 0.00 0.000000 0.500000 0.500000
321 1 0.95 3.0641 0.00 0.000000 0.500000 0.500000
321 1 1.50 3.0641 0.00 0.000000 1.000000 0.000000
321 1 2.50 3.0641 0.00 0.000000 1.000000 0.000000
321 1 3.50 3.0641 0.00 0.000000 0.999947 0.000053
321 1 4.50 3.0641 0.00 0.010322 0.988633 0.001044
321 1 5.50 3.0641 0.00 0.144150 0.853392 0.002459
321 1 6.50 3.0641 0.00 0.005871 0.892307 0.101822
321 1 7.50 3.0641 0.00 0.016154 0.831234 0.152612
321 1 8.50 3.0641 0.00 0.044167 0.470828 0.485005
321 1 9.50 3.0641 0.00 0.089300 0.910141 0.000559
321 1 10.50 3.0641 0.00 0.135785 0.460467 0.403748
321 1 11.50 3.0641 0.00 0.179961 0.465753 0.354286
321 1 12.50 3.0641 0.00 0.223621 0.444961 0.331418
321 1 13.50 3.0641 0.00 0.253578 0.666619 0.079803
321 1 14.50 3.0641 0.00 0.285174 0.646985 0.067841
321 1 0.55 3.0317 0.00 0.000000 0.500000 0.500000
321 1 0.65 3.0317 0.00 0.000000 0.500000 0.500000
321 1 0.75 3.0317 0.00 0.000000 0.500000 0.500000
321 1 0.85 3.0317 0.00 0.000000 0.500000 0.500000
321 1 0.95 3.0317 0.00 0.000000 0.500000 0.500000
321 1 1.50 3.0317 0.00 0.000000 1.000000 0.000000
321 1 2.50 3.0317 0.00 0.087649 0.912351 0.000000
321 1 3.50 3.0317 0.00 0.002417 0.997583 0.000000
321 1 4.50 3.0317 0.00 0.353110 0.433606 0.213283
321 1 5.50 3.0317 0.00 0.198768 0.392095 0.409137
321 1 6.50 3.0317 0.00 0.022773 0.754387 0.222840
321 1 7.50 3.0317 0.00 0.019332 0.484594 0.496074
321 1 8.50 3.0317 0.00 0.054288 0.451782 0.493930
321 1 9.50 3.0317 0.00 0.089832 0.439857 0.470311
321 1 10.50 3.0317 0.00 0.128456 0.425372 0.446172
321 1 11.50 3.0317 0.00 0.177344 0.483100 0.339557
321 1 12.50 3.0317 0.00 0.220943 0.674416 0.104641
321 1 13.50 3.0317 0.00 0.260014 0.700591 0.039396
321 1 14.50 3.0317 0.00 0.286295 0.639882 0.073823
321 1 0.55 2.9607 0.00 0.000000 0.500000 0.500000
321 1 0.65 2.9607 0.00 0.000000 0.500000 0.500000
321 1 0.75 2.9607 0.00 0.000000 0.500000 0.500000
321 1 0.85 2.9607 0.00 0.000000 0.500000 0.500000
321 1 0.95 2.9607 0.00 0.000000 0.500000 0.500000
321 1 1.50 2.9607 0.00 0.000000 1.000000 0.000000
321 1 2.50 2.9607 0.00 0.044665 0.955335 0.000000
321 1 3.50 2.9607 0.00 0.031117 0.755676 0.213206
321 1 4.50 2.9607 0.00 0.086438 0.612032 0.301531
321 1 5.50 2.9607 0.00 0.204899 0.402501 0.392600
321 1 6.50 2.9607 0.00 0.012641 0.788764 0.198595
321 1 7.50 2.9607 0.00 0.014253 0.511724 0.474023
321 1 8.50 2.9607 0.00 0.039137 0.467739 0.493124
321 1 9.50 2.9607 0.00 0.089351 0.662173 0.248476
321 1 10.50 2.9607 0.00 0.137247 0.383215 0.479538
321 1 11.50 2.9607 0.00 0.161448 0.488200 0.350351
321 1 12.50 2.9607 0.00 0.226336 0.746155 0.027509
321 1 13.50 2.9607 0.00 0.244724 0.699095 0.056181
321 1 14.50 2.9607 0.00 0.284163 0.663231 0.052606
321 1 0.55 2.7975 0.00 0.000000 0.500000 0.500000
321 1 0.65 2.7975 0.00 0.000000 0.500000 0.500000
321 1 0.75 2.7975 0.00 0.000000 0.500000 0.500000
321 1 0.85 2.7975 0.00 0.000000 0.500000 0.500000
321 1 0.95 2.7975 0.00 0.000000 0.500000 0.500000
321 1 1.50 2.7975 0.00 0.000000 1.000000 0.000000
321 1 2.50 2.7975 0.00 0.000000 1.000000 0.000000
321 1 3.50 2.7975 0.00 0.000652 0.632225 0.367123
321 1 4.50 2.7975 0.00 0.224453 0.754908 0.020639
321 1 5.50 2.7975 0.00 0.354108 0.512142 0.133750
321 1 6.50 2.7975 0.00 0.061183 0.737272 0.201545
321 1 7.50 2.7975 0.00 0.025293 0.595062 0.379645
321 1 8.50 2.7975 0.00 0.056459 0.514237 0.429304
321 1 9.50 2.7975 0.00 0.101552 0.402599 0.495849
321 1 10.50 2.7975 0.00 0.157445 0.380135 0.462420
321 1 11.50 2.7975 0.00 0.193004 0.459224 0.347772
321 1 12.50 2.7975 0.00 0.239221 0.375523 0.385255
321 1 13.50 2.7975 0.00 0.270651 0.626212 0.103136
321 1 14.50 2.7975 0.00 0.301466 0.609837 0.088697
2212 1 0.55 3.0896 0.00 0.000000 0.500000 0.500000
2212 1 0.65 3.0896 0.00 0.000000 0.500000 0.500000
2212 1 0.75 3.0896 0.00 0.000000 0.500000 0.500000
2212 1 0.85 3.0896 0.00 0.000000 0.500000 0.500000
2212 1 0.95 3.0896 0.00 0.000000 0.500000 0.500000
2212 1 1.50 3.0896 0.00 0.000000 0.000000 1.000000
2212 1 2.50 3.0896 0.00 0.000000 0.000000 1.000000
2212 1 3.50 3.0896 0.00 0.000000 0.215492 0.784508
2212 1 4.50 3.0896 0.00 0.000000 0.356057 0.643943
2212 1 5.50 3.0896 0.00 0.000000 0.279865 0.720135
2212 1 6.50 3.0896 0.00 0.000000 0.170710 0.829290
2212 1 7.50 3.0896 0.00 0.000000 0.337676 0.662324
2212 1 8.50 3.0896 0.00 0.000000 0.340406 0.659594
2212 1 9.50 3.0896 0.00 0.000000 0.396933 0.603067
2212 1 10.50 3.0896 0.00 0.000000 0.364309 0.635691
2212 1 11.50 3.0896 0.00 0.000000 0.117536 0.882464
2212 1 12.50 3.0896 0.00 0.000000 0.018587 0.981413
2212 1 13.50 3.0896 0.00 0.000000 0.061699 0.938301
2212 1 14.50 3.0896 0.00 0.000000 0.074531 0.925469
2212 1 0.55 3.0641 0.00 0.000000 0.500000 0.500000
2212 1 0.65 3.0641 0.00 0.000000 0.500000 0.500000
2212 1 0.75 3.0641 0.00 0.000000 0.500000 0.500000
2212 1 0.85 3.0641 0.00 0.000000 0.500000 0.500000
2212 1 0.95 3.0641 0.00 0.000000 0.500000 0.500000
2212 1 1.50 3.0641 0.00 0.000000 0.000000 1.000000
2212 1 2.50 3.0641 0.00 0.000000 0.000000 1.000000
2212 1 3.50 3.0641 0.00 0.000000 0.000000 1.000000
2212 1 4.50 3.0641 0.00 -0.000000 0.000047 0.999953
2212 1 5.50 3.0641 0.00 0.000000 0.103774 0.896226
2212 1 6.50 3.0641 0.00 0.000000 0.000456 0.999544
2212 1 7.50 3.0641 0.00 0.000000 0.000559 0.999441
2212 1 8.50 3.0641 0.00 0.000000 0.272378 0.727622
2212 1 9.50 3.0641 0.00 0.000000 0.000064 0.999936
2212 1 10.50 3.0641 0.00 0.000000 0.111275 0.888725
2212 1 11.50 3.0641 0.00 0.000000 0.086713 0.913287
2212 1 12.50 3.0641 0.00 0.000000 0.097629 0.902371
2212 1 13.50 3.0641 0.00 0.000000 0.034072 0.965928
2212 1 14.50 3.0641 0.00 0.000000 0.056010 0.943990
2212 1 0.55 3.0317 0.00 0.000000 0.500000 0.500000
2212 1 0.65 3.0317 0.00 0.000000 0.500000 0.500000
2212 1 0.75 3.0317 0.00 0.000000 0.500000 0.500000
2212 1 0.85 3.0317 0.00 0.000000 0.500000 0.500000
2212 1 0.95 3.0317 0.00 0.000000 0.500000 0.500000
2212 1 1.50 3.0317 0.00 0.000000 0.000000 1.000000
2212 1 2.50 3.0317 0.00 0.000000 0.000000 1.000000
2212 1 3.50 3.0317 0.00 0.000000 0.000001 0.999999
2212 1 4.50 3.0317 0.00 0.000000 0.270621 0.729379
2212 1 5.50 3.0317 0.00 0.000000 0.226689 0.773311
2212 1 6.50 3.0317 0.00 0.000000 0.013747 0.986253
2212 1 7.50 3.0317 0.00 0.000000 0.364686 0.635314
2212 1 8.50 3.0317 0.00 0.000000 0.350703 0.649297
2212 1 9.50 3.0317 0.00 0.000000 0.231271 0.768729
2212 1 10.50 3.0317 0.00 0.000000 0.178792 0.821208
2212 1 11.50 3.0317 0.00 0.000000 0.082347 0.917653
2212 1 12.50 3.0317 0.00 0.000000 0.025601 0.974399
2212 1 13.50 3.0317 0.00 0.000000 0.036058 0.963942
2212 1 14.50 3.0317 0.00 0.000000 0.055422 0.944578
2212 1 0.55 2.9607 0.00 0.000000 0.500000 0.500000
2212 1 0.65 2.9607 0.00 0.000000 0.500000 0.500000
2212 1 0.75 2.9607 0.00 0.000000 0.500000 0.500000
2212 1 0.85 2.9607 0.00 0.000000 0.500000 0.500000
2212 1 0.95 2.9607 0.00 0.000000 0.500000 0.500000
2212 1 1.50 2.9607 0.00 0.000000 0.000000 1.000000
2212 1 2.50 2.9607 0.00 0.000000 0.000000 1.000000
2212 1 3.50 2.9607 0.00 0.000000 0.008286 0.991714
2212 1 4.50 2.9607 0.00 0.000000 0.072136 0.927864
2212 1 5.50 2.9607 0.00 0.000000 0.212571 0.787429
2212 1 6.50 2.9607 0.00 0.000000 0.006525 0.993475
2212 1 7.50 2.9607 0.00 0.000000 0.193940 0.806060
2212 1 8.50 2.9607 0.00 0.000000 0.347159 0.652841
2212 1 9.50 2.9607 0.00 0.000000 0.012588 0.987412
2212 1 10.50 2.9607 0.00 0.000000 0.284845 0.715155
2212 1 11.50 2.9607 0.00 0.000000 0.088161 0.911839
2212 1 12.50 2.9607 0.00 0.000000 0.013920 0.986080
2212 1 13.50 2.9607 0.00 0.000000 0.033373 0.966627
2212 1 14.50 2.9607 0.00 0.000000 0.058958 0.941042
2212 1 0.55 2.7975 0.00 0.000000 0.500000 0.500000
2212 1 0.65 2.7975 0.00 0.000000 0.500000 0.500000
2212 1 0.75 2.7975 0.00 0.000000 0.500000 0.500000
2212 1 0.85 2.7975 0.00 0.000000 0.500000 0.500000
2212 1 0.95 2.7975 0.00 0.000000 0.500000 0.500000
2212 1 1.50 2.7975 0.00 0.000000 0.000000 1.000000
2212 1 2.50 2.7975 0.00 0.000000 0.000000 1.000000
2212 1 3.50 2.7975 0.00 0.000000 0.015306 0.984694
2212 1 4.50 2.7975 0.00 0.000000 0.154326 0.845674
2212 1 5.50 2.7975 0.00 0.000000 0.393572 0.606428
2212 1 6.50 2.7975 0.00 0.000000 0.028915 0.971085
2212 1 7.50 2.7975 0.00 0.000000 0.043326 0.956674
2212 1 8.50 2.7975 0.00 0.000000 0.113475 0.886525
2212 1 9.50 2.7975 0.00 0.000000 0.393411 0.606589
2212 1 10.50 2.7975 0.00 0.000000 0.237372 0.762628
2212 1 11.50 2.7975 0.00 0.000000 0.102436 0.897564
2212 1 12.50 2.7975 0.00 0.000000 0.149122 0.850878
2212 1 13.50 2.7975 0.00 0.000000 0.045062 0.954938
2212 1 14.50 2.7975 0.00 0.000000 0.063493 0.936507
## pfRICH PID LUT - version 0.1
## Eta Bins: [-3.8, -3.5], [-3.5, -3.0], [-3.0, -2.8], [-2.8, -2.0], [-2.0,
-1.5]
## Momentum Bins [GeV]: [0.05, 0.1] : then 0.1 to 1.0 in 0.1 wide steps :
then 1.0 to 15.0 in 1.0 wide steps
## NOTE: Particles thrown flat within Eta and Momentum Bins - results
reported at bin center
## NOTE: No charge scan - only thrown electrons, pi+, K+, and proton - assume
charge symmetry at this point
## NOTE: Particles thrown flat in Phi in this run - Phi value is a dummy
## Format: PID Charge Momentum Theta Phi prob-electron prob-pion prob-kaon
prob-proton
11 -1 0.08 3.0896 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.15 3.0896 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.25 3.0896 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.35 3.0896 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.45 3.0896 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.55 3.0896 0.00 0.820912 0.179088 0.000000 0.000000
11 -1 0.65 3.0896 0.00 0.816904 0.183096 0.000000 0.000000
11 -1 0.75 3.0896 0.00 0.695519 0.304481 0.000000 0.000000
11 -1 0.85 3.0896 0.00 0.968533 0.031467 0.000000 0.000000
11 -1 0.95 3.0896 0.00 0.570381 0.429619 0.000000 0.000000
11 -1 1.50 3.0896 0.00 0.594750 0.405250 0.000000 0.000000
11 -1 2.50 3.0896 0.00 0.844381 0.155619 0.000000 0.000000
11 -1 3.50 3.0896 0.00 0.774869 0.225131 0.000000 0.000000
11 -1 4.50 3.0896 0.00 0.676494 0.323506 0.000000 0.000000
11 -1 5.50 3.0896 0.00 0.627163 0.372837 0.000000 0.000000
11 -1 6.50 3.0896 0.00 0.582670 0.417330 0.000000 0.000000
11 -1 7.50 3.0896 0.00 0.565104 0.434896 0.000000 0.000000
11 -1 8.50 3.0896 0.00 0.548254 0.451746 0.000000 0.000000
11 -1 9.50 3.0896 0.00 0.539713 0.460287 0.000000 0.000000
11 -1 10.50 3.0896 0.00 0.532581 0.467419 0.000000 0.000000
11 -1 11.50 3.0896 0.00 0.529443 0.470557 0.000000 0.000000
11 -1 12.50 3.0896 0.00 0.522247 0.477753 0.000000 0.000000
11 -1 13.50 3.0896 0.00 0.519769 0.480231 0.000000 0.000000
11 -1 14.50 3.0896 0.00 0.517555 0.482445 0.000000 0.000000
11 -1 0.08 3.0641 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.15 3.0641 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.25 3.0641 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.35 3.0641 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.45 3.0641 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.55 3.0641 0.00 0.996657 0.003343 0.000000 0.000000
11 -1 0.65 3.0641 0.00 0.999980 0.000020 0.000000 0.000000
11 -1 0.75 3.0641 0.00 0.991958 0.008042 0.000000 0.000000
11 -1 0.85 3.0641 0.00 0.985046 0.014954 0.000000 0.000000
11 -1 0.95 3.0641 0.00 0.998802 0.001198 0.000000 0.000000
11 -1 1.50 3.0641 0.00 0.644962 0.355038 0.000000 0.000000
11 -1 2.50 3.0641 0.00 0.885331 0.114669 0.000000 0.000000
11 -1 3.50 3.0641 0.00 0.798450 0.201550 0.000000 0.000000
11 -1 4.50 3.0641 0.00 0.703394 0.296606 0.000000 0.000000
11 -1 5.50 3.0641 0.00 0.642122 0.357878 0.000000 0.000000
11 -1 6.50 3.0641 0.00 0.600882 0.399118 0.000000 0.000000
11 -1 7.50 3.0641 0.00 0.573328 0.426672 0.000000 0.000000
11 -1 8.50 3.0641 0.00 0.558285 0.441715 0.000000 0.000000
11 -1 9.50 3.0641 0.00 0.546135 0.453865 0.000000 0.000000
11 -1 10.50 3.0641 0.00 0.537807 0.462193 0.000000 0.000000
11 -1 11.50 3.0641 0.00 0.531525 0.468475 0.000000 0.000000
11 -1 12.50 3.0641 0.00 0.526199 0.473801 0.000000 0.000000
11 -1 13.50 3.0641 0.00 0.522887 0.477113 0.000000 0.000000
11 -1 14.50 3.0641 0.00 0.519599 0.480401 0.000000 0.000000
11 -1 0.08 3.0317 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.15 3.0317 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.25 3.0317 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.35 3.0317 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.45 3.0317 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.55 3.0317 0.00 0.584644 0.415356 0.000000 0.000000
11 -1 0.65 3.0317 0.00 0.534399 0.465601 0.000000 0.000000
11 -1 0.75 3.0317 0.00 0.587790 0.412210 0.000000 0.000000
11 -1 0.85 3.0317 0.00 0.643793 0.356207 0.000000 0.000000
11 -1 0.95 3.0317 0.00 0.510761 0.489239 0.000000 0.000000
11 -1 1.50 3.0317 0.00 0.771504 0.228496 0.000000 0.000000
11 -1 2.50 3.0317 0.00 0.872828 0.127172 0.000000 0.000000
11 -1 3.50 3.0317 0.00 0.799781 0.200219 0.000000 0.000000
11 -1 4.50 3.0317 0.00 0.695511 0.304489 0.000000 0.000000
11 -1 5.50 3.0317 0.00 0.638672 0.361328 0.000000 0.000000
11 -1 6.50 3.0317 0.00 0.594468 0.405532 0.000000 0.000000
11 -1 7.50 3.0317 0.00 0.570281 0.429719 0.000000 0.000000
11 -1 8.50 3.0317 0.00 0.554746 0.445254 0.000000 0.000000
11 -1 9.50 3.0317 0.00 0.545980 0.454020 0.000000 0.000000
11 -1 10.50 3.0317 0.00 0.538913 0.461087 0.000000 0.000000
11 -1 11.50 3.0317 0.00 0.531874 0.468126 0.000000 0.000000
11 -1 12.50 3.0317 0.00 0.526531 0.473469 0.000000 0.000000
11 -1 13.50 3.0317 0.00 0.522194 0.477806 0.000000 0.000000
11 -1 14.50 3.0317 0.00 0.519484 0.480516 0.000000 0.000000
11 -1 0.08 2.9607 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.15 2.9607 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.25 2.9607 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.35 2.9607 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.45 2.9607 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.55 2.9607 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.65 2.9607 0.00 0.517811 0.482189 0.000000 0.000000
11 -1 0.75 2.9607 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.85 2.9607 0.00 0.736561 0.263439 0.000000 0.000000
11 -1 0.95 2.9607 0.00 0.545495 0.454505 0.000000 0.000000
11 -1 1.50 2.9607 0.00 0.999990 0.000010 0.000000 0.000000
11 -1 2.50 2.9607 0.00 0.880670 0.119330 0.000000 0.000000
11 -1 3.50 2.9607 0.00 0.810360 0.189640 0.000000 0.000000
11 -1 4.50 2.9607 0.00 0.706990 0.293010 0.000000 0.000000
11 -1 5.50 2.9607 0.00 0.638739 0.361261 0.000000 0.000000
11 -1 6.50 2.9607 0.00 0.601370 0.398630 0.000000 0.000000
11 -1 7.50 2.9607 0.00 0.575178 0.424822 0.000000 0.000000
11 -1 8.50 2.9607 0.00 0.560903 0.439097 0.000000 0.000000
11 -1 9.50 2.9607 0.00 0.546099 0.453901 0.000000 0.000000
11 -1 10.50 2.9607 0.00 0.537575 0.462425 0.000000 0.000000
11 -1 11.50 2.9607 0.00 0.534210 0.465790 0.000000 0.000000
11 -1 12.50 2.9607 0.00 0.525851 0.474149 0.000000 0.000000
11 -1 13.50 2.9607 0.00 0.523914 0.476086 0.000000 0.000000
11 -1 14.50 2.9607 0.00 0.519721 0.480279 0.000000 0.000000
11 -1 0.08 2.7975 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.15 2.7975 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.25 2.7975 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.35 2.7975 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.45 2.7975 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.55 2.7975 0.00 0.944870 0.055130 0.000000 0.000000
11 -1 0.65 2.7975 0.00 0.535976 0.464024 0.000000 0.000000
11 -1 0.75 2.7975 0.00 0.828477 0.171523 0.000000 0.000000
11 -1 0.85 2.7975 0.00 0.985078 0.014922 0.000000 0.000000
11 -1 0.95 2.7975 0.00 0.769982 0.230018 0.000000 0.000000
11 -1 1.50 2.7975 0.00 0.999787 0.000213 0.000000 0.000000
11 -1 2.50 2.7975 0.00 0.899691 0.100309 0.000000 0.000000
11 -1 3.50 2.7975 0.00 0.786474 0.213526 0.000000 0.000000
11 -1 4.50 2.7975 0.00 0.688065 0.311935 0.000000 0.000000
11 -1 5.50 2.7975 0.00 0.627914 0.372086 0.000000 0.000000
11 -1 6.50 2.7975 0.00 0.591617 0.408383 0.000000 0.000000
11 -1 7.50 2.7975 0.00 0.567638 0.432362 0.000000 0.000000
11 -1 8.50 2.7975 0.00 0.554591 0.445409 0.000000 0.000000
11 -1 9.50 2.7975 0.00 0.543559 0.456441 0.000000 0.000000
11 -1 10.50 2.7975 0.00 0.534616 0.465384 0.000000 0.000000
11 -1 11.50 2.7975 0.00 0.530040 0.469960 0.000000 0.000000
11 -1 12.50 2.7975 0.00 0.524437 0.475563 0.000000 0.000000
11 -1 13.50 2.7975 0.00 0.521079 0.478921 0.000000 0.000000
11 -1 14.50 2.7975 0.00 0.517948 0.482052 0.000000 0.000000
211 1 0.08 3.0896 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.15 3.0896 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.25 3.0896 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.35 3.0896 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.45 3.0896 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.55 3.0896 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.65 3.0896 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.75 3.0896 0.00 0.000005 0.999995 0.000000 0.000000
211 1 0.85 3.0896 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.95 3.0896 0.00 0.018873 0.981127 0.000000 0.000000
211 1 1.50 3.0896 0.00 0.070770 0.929230 0.000000 0.000000
211 1 2.50 3.0896 0.00 0.066668 0.846399 0.086933 0.000000
211 1 3.50 3.0896 0.00 0.234630 0.299160 0.466210 0.000000
211 1 4.50 3.0896 0.00 0.313117 0.195520 0.491362 0.000000
211 1 5.50 3.0896 0.00 0.376271 0.135896 0.487833 0.000000
211 1 6.50 3.0896 0.00 0.406275 0.416574 0.177150 0.000000
211 1 7.50 3.0896 0.00 0.433104 0.545074 0.021822 0.000000
211 1 8.50 3.0896 0.00 0.439765 0.494287 0.065948 0.000000
211 1 9.50 3.0896 0.00 0.457122 0.444239 0.098639 0.000000
211 1 10.50 3.0896 0.00 0.465794 0.367810 0.166396 0.000000
211 1 11.50 3.0896 0.00 0.474269 0.317962 0.207769 0.000000
211 1 12.50 3.0896 0.00 0.475902 0.290784 0.233314 0.000000
211 1 13.50 3.0896 0.00 0.478393 0.225875 0.295732 0.000000
211 1 14.50 3.0896 0.00 0.483350 0.211303 0.305348 0.000000
211 1 0.08 3.0641 0.00 0.000000 1.000000 0.000000
1249929749711127462130814941442627454093949552504892583888254073307136.000000
211 1 0.15 3.0641 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.25 3.0641 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.35 3.0641 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.45 3.0641 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.55 3.0641 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.65 3.0641 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.75 3.0641 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.85 3.0641 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.95 3.0641 0.00 0.000000 1.000000 0.000000 0.000000
211 1 1.50 3.0641 0.00 0.029063 0.970937 0.000000 0.000000
211 1 2.50 3.0641 0.00 0.049665 0.943375 0.006960 0.000000
211 1 3.50 3.0641 0.00 0.189927 0.799602 0.010471 0.000000
211 1 4.50 3.0641 0.00 0.293040 0.375112 0.331848 0.000000
211 1 5.50 3.0641 0.00 0.361304 0.191749 0.446947 0.000000
211 1 6.50 3.0641 0.00 0.396557 0.519282 0.084161 0.000000
211 1 7.50 3.0641 0.00 0.423920 0.547205 0.028874 0.000000
211 1 8.50 3.0641 0.00 0.440241 0.511597 0.048162 0.000000
211 1 9.50 3.0641 0.00 0.452412 0.455627 0.091962 0.000000
211 1 10.50 3.0641 0.00 0.462423 0.400326 0.137252 0.000000
211 1 11.50 3.0641 0.00 0.467885 0.353633 0.178482 0.000000
211 1 12.50 3.0641 0.00 0.472988 0.298010 0.229002 0.000000
211 1 13.50 3.0641 0.00 0.476829 0.267012 0.256159 0.000000
211 1 14.50 3.0641 0.00 0.480039 0.230209 0.289753 0.000000
211 1 0.08 3.0317 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.15 3.0317 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.25 3.0317 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.35 3.0317 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.45 3.0317 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.55 3.0317 0.00 0.000004 0.999996 0.000000 0.000000
211 1 0.65 3.0317 0.00 0.012970 0.987030 0.000000 0.000000
211 1 0.75 3.0317 0.00 0.001036 0.998964 0.000000 0.000000
211 1 0.85 3.0317 0.00 0.000344 0.999656 0.000000 0.000000
211 1 0.95 3.0317 0.00 0.203091 0.796909 0.000000 0.000000
211 1 1.50 3.0317 0.00 0.005403 0.994597 0.000000 -nan
211 1 2.50 3.0317 0.00 0.059318 0.463820 0.476862 -nan
211 1 3.50 3.0317 0.00 0.186991 0.465519 0.347491 -nan
211 1 4.50 3.0317 0.00 0.301182 0.203464 0.495355 0.000000
211 1 5.50 3.0317 0.00 0.357021 0.177403 0.465576 0.000000
211 1 6.50 3.0317 0.00 0.404423 0.395295 0.200282 0.000000
211 1 7.50 3.0317 0.00 0.425404 0.554111 0.020485 0.000000
211 1 8.50 3.0317 0.00 0.440781 0.509033 0.050185 0.000000
211 1 9.50 3.0317 0.00 0.457880 0.436430 0.105690 0.000000
211 1 10.50 3.0317 0.00 0.463613 0.400490 0.135897 0.000000
211 1 11.50 3.0317 0.00 0.469810 0.339821 0.190369 0.000000
211 1 12.50 3.0317 0.00 0.473469 0.297448 0.229083 0.000000
211 1 13.50 3.0317 0.00 0.476985 0.256934 0.266081 0.000000
211 1 14.50 3.0317 0.00 0.480516 0.230900 0.288584 0.000000
211 1 0.08 2.9607 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.15 2.9607 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.25 2.9607 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.35 2.9607 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.45 2.9607 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.55 2.9607 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.65 2.9607 0.00 0.050804 0.949196 0.000000 0.000000
211 1 0.75 2.9607 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.85 2.9607 0.00 0.000003 0.999997 0.000000 0.000000
211 1 0.95 2.9607 0.00 0.038888 0.961112 0.000000 0.000000
211 1 1.50 2.9607 0.00 0.000001 0.999999 0.000000 0.000000
211 1 2.50 2.9607 0.00 0.052103 0.482991 0.464906 0.000000
211 1 3.50 2.9607 0.00 0.189640 0.374920 0.435440 0.000000
211 1 4.50 2.9607 0.00 0.291794 0.265655 0.442550 0.000000
211 1 5.50 2.9607 0.00 0.361261 0.171401 0.467339 0.000000
211 1 6.50 2.9607 0.00 0.392612 0.447508 0.159879 0.000000
211 1 7.50 2.9607 0.00 0.423403 0.548850 0.027747 0.000000
211 1 8.50 2.9607 0.00 0.440998 0.493026 0.065976 0.000000
211 1 9.50 2.9607 0.00 0.451533 0.464457 0.084010 0.000000
211 1 10.50 2.9607 0.00 0.461962 0.400790 0.137247 0.000000
211 1 11.50 2.9607 0.00 0.467518 0.343759 0.188724 0.000000
211 1 12.50 2.9607 0.00 0.472515 0.306623 0.220862 0.000000
211 1 13.50 2.9607 0.00 0.476238 0.263266 0.260496 0.000000
211 1 14.50 2.9607 0.00 0.480398 0.226375 0.293227 0.000000
211 1 0.08 2.7975 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.15 2.7975 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.25 2.7975 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.35 2.7975 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.45 2.7975 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.55 2.7975 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.65 2.7975 0.00 0.013260 0.986740 0.000000 0.000000
211 1 0.75 2.7975 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.85 2.7975 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.95 2.7975 0.00 0.000017 0.999983 0.000000 0.000000
211 1 1.50 2.7975 0.00 0.000003 0.999997 0.000000 0.000000
211 1 2.50 2.7975 0.00 0.064794 0.888629 0.046577 0.000000
211 1 3.50 2.7975 0.00 0.216132 0.491398 0.292470 0.000000
211 1 4.50 2.7975 0.00 0.304614 0.214986 0.480400 0.000000
211 1 5.50 2.7975 0.00 0.370643 0.136637 0.492720 0.000000
211 1 6.50 2.7975 0.00 0.402322 0.282686 0.314992 0.000000
211 1 7.50 2.7975 0.00 0.430427 0.518416 0.051157 0.000000
211 1 8.50 2.7975 0.00 0.446339 0.474793 0.078868 0.000000
211 1 9.50 2.7975 0.00 0.460345 0.425133 0.114522 0.000000
211 1 10.50 2.7975 0.00 0.464386 0.363673 0.171941 0.000000
211 1 11.50 2.7975 0.00 0.470976 0.308985 0.220039 0.000000
211 1 12.50 2.7975 0.00 0.474858 0.281060 0.244082 0.000000
211 1 13.50 2.7975 0.00 0.479157 0.251340 0.269503 0.000000
211 1 14.50 2.7975 0.00 0.482906 0.220787 0.296308 0.000000
321 1 0.08 3.0896 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.15 3.0896 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.25 3.0896 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.35 3.0896 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.45 3.0896 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.55 3.0896 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.65 3.0896 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.75 3.0896 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.85 3.0896 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.95 3.0896 0.00 0.000000 0.000000 1.000000 0.000000
321 1 1.50 3.0896 0.00 0.000000 0.000000 1.000000 0.000000
321 1 2.50 3.0896 0.00 0.000000 0.000000 1.000000 0.000000
321 1 3.50 3.0896 0.00 0.000000 0.105568 0.894432 0.000000
321 1 4.50 3.0896 0.00 0.000000 0.313310 0.625592 0.061098
321 1 5.50 3.0896 0.00 0.000000 0.314877 0.577898 0.107225
321 1 6.50 3.0896 0.00 0.000000 0.027119 0.523884 0.448996
321 1 7.50 3.0896 0.00 0.000000 0.028167 0.477655 0.494178
321 1 8.50 3.0896 0.00 0.000000 0.074478 0.433141 0.492381
321 1 9.50 3.0896 0.00 0.000000 0.124474 0.379123 0.496403
321 1 10.50 3.0896 0.00 0.000000 0.171563 0.337001 0.491436
321 1 11.50 3.0896 0.00 0.000000 0.194550 0.431152 0.374298
321 1 12.50 3.0896 0.00 0.000000 0.260400 0.733195 0.006405
321 1 13.50 3.0896 0.00 0.000000 0.283041 0.672535 0.044424
321 1 14.50 3.0896 0.00 0.000000 0.305348 0.584070 0.110583
321 1 0.08 3.0641 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.15 3.0641 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.25 3.0641 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.35 3.0641 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.45 3.0641 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.55 3.0641 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.65 3.0641 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.75 3.0641 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.85 3.0641 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.95 3.0641 0.00 0.000000 0.000000 1.000000 0.000000
321 1 1.50 3.0641 0.00 0.000000 0.000000 1.000000 0.000000
321 1 2.50 3.0641 0.00 0.000000 0.000000 1.000000 0.000000
321 1 3.50 3.0641 0.00 0.000000 0.000000 0.999947 0.000053
321 1 4.50 3.0641 0.00 0.000000 0.010322 0.988633 0.001044
321 1 5.50 3.0641 0.00 0.000000 0.144150 0.853392 0.002459
321 1 6.50 3.0641 0.00 0.000000 0.005871 0.892307 0.101822
321 1 7.50 3.0641 0.00 0.000000 0.016154 0.831234 0.152612
321 1 8.50 3.0641 0.00 0.000000 0.044167 0.470828 0.485005
321 1 9.50 3.0641 0.00 0.000000 0.089300 0.910141 0.000559
321 1 10.50 3.0641 0.00 0.000000 0.135785 0.460467 0.403748
321 1 11.50 3.0641 0.00 0.000000 0.179961 0.465753 0.354286
321 1 12.50 3.0641 0.00 0.000000 0.223621 0.444961 0.331418
321 1 13.50 3.0641 0.00 0.000000 0.253578 0.666619 0.079803
321 1 14.50 3.0641 0.00 0.000000 0.285174 0.646985 0.067841
321 1 0.08 3.0317 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.15 3.0317 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.25 3.0317 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.35 3.0317 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.45 3.0317 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.55 3.0317 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.65 3.0317 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.75 3.0317 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.85 3.0317 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.95 3.0317 0.00 0.000000 0.000000 1.000000 0.000000
321 1 1.50 3.0317 0.00 -nan 0.000000 1.000000 0.000000
321 1 2.50 3.0317 0.00 -nan 0.087649 0.912351 0.000000
321 1 3.50 3.0317 0.00 0.000000 0.002417 0.997583 0.000000
321 1 4.50 3.0317 0.00 0.000000 0.353110 0.433606 0.213283
321 1 5.50 3.0317 0.00 0.000000 0.198768 0.392095 0.409137
321 1 6.50 3.0317 0.00 -nan 0.022773 0.754387 0.222840
321 1 7.50 3.0317 0.00 0.000000 0.019332 0.484594 0.496074
321 1 8.50 3.0317 0.00 0.000000 0.054288 0.451782 0.493930
321 1 9.50 3.0317 0.00 -nan 0.089832 0.439857 0.470311
321 1 10.50 3.0317 0.00 0.000000 0.128456 0.425372 0.446172
321 1 11.50 3.0317 0.00 0.000000 0.177344 0.483100 0.339557
321 1 12.50 3.0317 0.00 -nan 0.220943 0.674416 0.104641
321 1 13.50 3.0317 0.00 0.000000 0.260014 0.700591 0.039396
321 1 14.50 3.0317 0.00 0.000000 0.286295 0.639882 0.073823
321 1 0.08 2.9607 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.15 2.9607 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.25 2.9607 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.35 2.9607 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.45 2.9607 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.55 2.9607 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.65 2.9607 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.75 2.9607 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.85 2.9607 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.95 2.9607 0.00 0.000000 0.000000 1.000000 0.000000
321 1 1.50 2.9607 0.00 0.000000 0.000000 1.000000 0.000000
321 1 2.50 2.9607 0.00 0.000000 0.044665 0.955335 0.000000
321 1 3.50 2.9607 0.00 0.000000 0.031117 0.755676 0.213206
321 1 4.50 2.9607 0.00 0.000000 0.086438 0.612032 0.301531
321 1 5.50 2.9607 0.00 0.000000 0.204899 0.402501 0.392600
321 1 6.50 2.9607 0.00 0.000000 0.012641 0.788764 0.198595
321 1 7.50 2.9607 0.00 0.000000 0.014253 0.511724 0.474023
321 1 8.50 2.9607 0.00 0.000000 0.039137 0.467739 0.493124
321 1 9.50 2.9607 0.00 0.000000 0.089351 0.662173 0.248476
321 1 10.50 2.9607 0.00 0.000000 0.137247 0.383215 0.479538
321 1 11.50 2.9607 0.00 0.000000 0.161448 0.488200 0.350351
321 1 12.50 2.9607 0.00 0.000000 0.226336 0.746155 0.027509
321 1 13.50 2.9607 0.00 0.000000 0.244724 0.699095 0.056181
321 1 14.50 2.9607 0.00 0.000000 0.284163 0.663231 0.052606
321 1 0.08 2.7975 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.15 2.7975 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.25 2.7975 0.00
69442959116059723827408726541773148034926167480004203707445942068917968278953989398069248.000000
0.000000 1.000000 0.000000
321 1 0.35 2.7975 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.45 2.7975 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.55 2.7975 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.65 2.7975 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.75 2.7975 0.00 -nan 0.000000 1.000000 0.000000
321 1 0.85 2.7975 0.00 0.000000 0.000000 1.000000 0.000000
321 1 0.95 2.7975 0.00 0.000000 0.000000 1.000000 0.000000
321 1 1.50 2.7975 0.00 0.000000 0.000000 1.000000 0.000000
321 1 2.50 2.7975 0.00 0.000000 0.000000 1.000000 0.000000
321 1 3.50 2.7975 0.00 0.000000 0.000652 0.632225 0.367123
321 1 4.50 2.7975 0.00 0.000000 0.224453 0.754908 0.020639
321 1 5.50 2.7975 0.00 0.000000 0.354108 0.512142 0.133750
321 1 6.50 2.7975 0.00 0.000000 0.061183 0.737272 0.201545
321 1 7.50 2.7975 0.00 0.000000 0.025293 0.595062 0.379645
321 1 8.50 2.7975 0.00 0.000000 0.056459 0.514237 0.429304
321 1 9.50 2.7975 0.00 0.000000 0.101552 0.402599 0.495849
321 1 10.50 2.7975 0.00 0.000000 0.157445 0.380135 0.462420
321 1 11.50 2.7975 0.00 0.000000 0.193004 0.459224 0.347772
321 1 12.50 2.7975 0.00 0.000000 0.239221 0.375523 0.385255
321 1 13.50 2.7975 0.00 0.000000 0.270651 0.626212 0.103136
321 1 14.50 2.7975 0.00 0.000000 0.301466 0.609837 0.088697
2212 1 0.08 3.0896 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.15 3.0896 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.25 3.0896 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.35 3.0896 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.45 3.0896 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.55 3.0896 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.65 3.0896 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.75 3.0896 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.85 3.0896 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.95 3.0896 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 1.50 3.0896 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 2.50 3.0896 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 3.50 3.0896 0.00 0.000000 -nan 0.215492 0.784508
2212 1 4.50 3.0896 0.00 0.000000 0.000000 0.356057 0.643943
2212 1 5.50 3.0896 0.00 -nan 0.000000 0.279865 0.720135
2212 1 6.50 3.0896 0.00
291228430815245216126826311259994157723187085730428057435496189432102689672719990233501802759421312700063046891674345283518805058266417548756211577481647443577962695457986646399227972577011925883047494513916107454708755468280956232466632436806240338647711744.000000
0.000000 0.170710 0.829290
2212 1 7.50 3.0896 0.00 0.000000 0.000000 0.337676 0.662324
2212 1 8.50 3.0896 0.00 0.000000 0.000000 0.340406 0.659594
2212 1 9.50 3.0896 0.00
-5551995652432595002794134145027321992423168922297038759692874145608921441192821042388712750275933366182918278299397964785604592434480895633044934113273850231545749977723162713979818422462212076499678953974488658525808313556511301962866709531227059689223819922225683732702495848564148249587901138699550720.000000
0.000000 0.396933 0.603067
2212 1 10.50 3.0896 0.00 -nan 0.000000 0.364309 0.635691
2212 1 11.50 3.0896 0.00 0.000000 0.000000 0.117536 0.882464
2212 1 12.50 3.0896 0.00 0.000000 0.000000 0.018587 0.981413
2212 1 13.50 3.0896 0.00 -nan
6964111035017396037231577106626991378222735929273688436188259952495714893940780973033789526806547822676535144599412168621836115878807376057564633864301830447935587020166639282107203944864523142901993558350106626246472648304969011907970710924976717824000.000000
0.061699 0.938301
2212 1 14.50 3.0896 0.00 -nan 0.000000 0.074531 0.925469
2212 1 0.08 3.0641 0.00 -nan 0.000000 0.000000 1.000000
2212 1 0.15 3.0641 0.00 -nan 0.000000 0.000000 1.000000
2212 1 0.25 3.0641 0.00 -nan 0.000000 0.000000 1.000000
2212 1 0.35 3.0641 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.45 3.0641 0.00 -nan 0.000000 0.000000 1.000000
2212 1 0.55 3.0641 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.65 3.0641 0.00 -nan 0.000000 0.000000 1.000000
2212 1 0.75 3.0641 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.85 3.0641 0.00 -nan 0.000000 0.000000 1.000000
2212 1 0.95 3.0641 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 1.50 3.0641 0.00 -nan 0.000000 0.000000 1.000000
2212 1 2.50 3.0641 0.00 -nan 0.000000 0.000000 1.000000
2212 1 3.50 3.0641 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 4.50 3.0641 0.00 0.000000 0.000000 0.000047 0.999953
2212 1 5.50 3.0641 0.00 -nan 0.000000 0.103774 0.896226
2212 1 6.50 3.0641 0.00 -nan 0.000000 0.000456 0.999544
2212 1 7.50 3.0641 0.00 -nan 0.000000 0.000559 0.999441
2212 1 8.50 3.0641 0.00 -nan 0.000000 0.272378 0.727622
2212 1 9.50 3.0641 0.00 0.000000 0.000000 0.000064 0.999936
2212 1 10.50 3.0641 0.00 0.000000 0.000000 0.111275 0.888725
2212 1 11.50 3.0641 0.00 0.000000 0.000000 0.086713 0.913287
2212 1 12.50 3.0641 0.00 0.000000 0.000000 0.097629 0.902371
2212 1 13.50 3.0641 0.00 0.000000 0.000000 0.034072 0.965928
2212 1 14.50 3.0641 0.00 0.000000 0.000000 0.056010 0.943990
2212 1 0.08 3.0317 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.15 3.0317 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.25 3.0317 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.35 3.0317 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.45 3.0317 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.55 3.0317 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.65 3.0317 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.75 3.0317 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.85 3.0317 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.95 3.0317 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 1.50 3.0317 0.00 -nan -nan 0.000000 1.000000
2212 1 2.50 3.0317 0.00 -nan -nan 0.000000 1.000000
2212 1 3.50 3.0317 0.00 -nan 0.000000 0.000001 0.999999
2212 1 4.50 3.0317 0.00 -nan 0.000000 0.270621 0.729379
2212 1 5.50 3.0317 0.00 -nan 0.000000 0.226689 0.773311
2212 1 6.50 3.0317 0.00 -nan -nan 0.013747 0.986253
2212 1 7.50 3.0317 0.00 -nan 0.000000 0.364686 0.635314
2212 1 8.50 3.0317 0.00 -nan 0.000000 0.350703 0.649297
2212 1 9.50 3.0317 0.00 -nan 0.000000 0.231271 0.768729
2212 1 10.50 3.0317 0.00 -nan 0.000000 0.178792 0.821208
2212 1 11.50 3.0317 0.00 0.000000 0.000000 0.082347 0.917653
2212 1 12.50 3.0317 0.00 0.000000 0.000000 0.025601 0.974399
2212 1 13.50 3.0317 0.00 0.000000 0.000000 0.036058 0.963942
2212 1 14.50 3.0317 0.00 0.000000 0.000000 0.055422 0.944578
2212 1 0.08 2.9607 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.15 2.9607 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.25 2.9607 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.35 2.9607 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.45 2.9607 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.55 2.9607 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.65 2.9607 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.75 2.9607 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.85 2.9607 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.95 2.9607 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 1.50 2.9607 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 2.50 2.9607 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 3.50 2.9607 0.00 0.000000 0.000000 0.008286 0.991714
2212 1 4.50 2.9607 0.00 0.000000 0.000000 0.072136 0.927864
2212 1 5.50 2.9607 0.00 0.000000 0.000000 0.212571 0.787429
2212 1 6.50 2.9607 0.00 0.000000 0.000000 0.006525 0.993475
2212 1 7.50 2.9607 0.00 0.000000 0.000000 0.193940 0.806060
2212 1 8.50 2.9607 0.00 0.000000 0.000000 0.347159 0.652841
2212 1 9.50 2.9607 0.00 0.000000 0.000000 0.012588 0.987412
2212 1 10.50 2.9607 0.00 0.000000 0.000000 0.284845 0.715155
2212 1 11.50 2.9607 0.00 0.000000 0.000000 0.088161 0.911839
2212 1 12.50 2.9607 0.00 0.000000 0.000000 0.013920 0.986080
2212 1 13.50 2.9607 0.00 0.000000 0.000000 0.033373 0.966627
2212 1 14.50 2.9607 0.00 0.000000 0.000000 0.058958 0.941042
2212 1 0.08 2.7975 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.15 2.7975 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.25 2.7975 0.00
4925932131729113811527445543052507304724555478781745866084776443836991933513728.000000
6964111060505883648814636797759483422419217999945146194807587321846308962022758052398293766302717352438789044824475646981168732695585647732291981390594181942597645441737710463475395998647705734212360887535031722068361148205641801685551158090990662713344.000000
0.000000 1.000000
2212 1 0.35 2.7975 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.45 2.7975 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.55 2.7975 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.65 2.7975 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.75 2.7975 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.85 2.7975 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 0.95 2.7975 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 1.50 2.7975 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 2.50 2.7975 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 3.50 2.7975 0.00 0.000000 0.000000 0.015306 0.984694
2212 1 4.50 2.7975 0.00 0.000000 0.000000 0.154326 0.845674
2212 1 5.50 2.7975 0.00 0.000000 0.000000 0.393572 0.606428
2212 1 6.50 2.7975 0.00 0.000000 0.000000 0.028915 0.971085
2212 1 7.50 2.7975 0.00 0.000000 0.000000 0.043326 0.956674
2212 1 8.50 2.7975 0.00 0.000000 0.000000 0.113475 0.886525
2212 1 9.50 2.7975 0.00 0.000000 0.000000 0.393411 0.606589
2212 1 10.50 2.7975 0.00 0.000000 0.000000 0.237372 0.762628
2212 1 11.50 2.7975 0.00 0.000000 0.000000 0.102436 0.897564
2212 1 12.50 2.7975 0.00 0.000000 0.000000 0.149122 0.850878
2212 1 13.50 2.7975 0.00 0.000000 0.000000 0.045062 0.954938
2212 1 14.50 2.7975 0.00 0.000000 0.000000 0.063493 0.936507
- [Eic-projdet-pfrich-software-l] First Set of pfRICH LUTs, Page, Brian, 03/06/2024
Archive powered by MHonArc 2.6.24.