Skip to Content.
Sympa Menu

sphenix-electronics-l - [Sphenix-electronics-l] Comparison of IV break down with Vop

sphenix-electronics-l AT lists.bnl.gov

Subject: sPHENIX discussion of electronics

List archive

Chronological Thread  
  • From: eric mannel <mannel AT bnl.gov>
  • To: sphenix-electronics-l AT lists.bnl.gov
  • Subject: [Sphenix-electronics-l] Comparison of IV break down with Vop
  • Date: Wed, 19 Feb 2020 19:17:15 -0500

I updated my slides from today to include a plot of the difference between the breakdown voltage as measured using the Debrecen test stand and software and the reported Vop from Hamamatsu.  Based on previously reported discussions with Hamamatsu, we expect that the difference should be 4.0V.  As one can see, the difference is 3.15V.  This may be related to way we define the breakdown voltage using the IV curve as compared to Hamamatsu method and is being investigated, but will take a few days to complete. As indicated on the slides, we have always seen some discrepancies in the measurement data for deliveries 3, 4 and 5 that are yet to be resolved. For the other 7 deliveries, the spread is of order 30mV.

My presentation is here: https://indico.bnl.gov/event/7674/

Eric

--
Eric Mannel, Ph.D.
PHENIX Group
Dept of Physics
Brookhaven National Lab.
631/344-7626 (Office)
914/659-3235 (Mobile)




  • [Sphenix-electronics-l] Comparison of IV break down with Vop, eric mannel, 02/19/2020

Archive powered by MHonArc 2.6.24.

Top of Page