Dear Test Beam Publication Committee (or TBPC): A very nice first draft. I have a number of questions and
comments that I outline below. (I second Jamie's comment that it would make the review process easier if line numbers were shown for each line in the next iteration.) One additional general comment - on *many* of the figures,
the text on the figure is way to small to be legible. I think the
TBPC will need to go through all these figures and make another
pass at organization and font size before these will be
publication-ready. Regards, JGL Comments: Abstract: The electromagnetic and hadronic calorimeter design... In the third sentence you mention the stochiastic resolution term
but not the constant. I think it would be better to say something
like : "To achieve the goals of the proposed physics program, an energy resolution of 15%/sqrt(E) and 100%/sqrt(E) are required for the electromagntic and hadronic calorimeters, respectively. In addition, the energy resolution function should have a small constant term, <5% for the electromagnetic calorimeter and <20% for the hadronic calorimeter." Body of Paper: line 6 - put "jets" in quotes when first introduced line 8 - the appears throughout the paper: comprises -> is comprised of "... sPHENIX is comprised of calorimeters and *a* tracking system..." line 13: is comprised of line 28-29: I don't understand this sentence. The requirement on projectivity of the EMCal is to avoid spreading the EM showers out between multiple towers and keeping the shower size contained. "enable accurate jet measurements" is ambiguous, and not accurate in this discussion lines 31-35. I do not understand this argument. Why is the
cluster size quoted in MeV, and how does this directly impact the
resolution? The argument we typically make has to do with the
size of the underlying event and its fluctuations. This needs to
be stated more clearly. line 45: is compacted by vibration line 58 : 11.2 degrees line 75/Figure 1c: This just shows the cutting wheel, but does
not support the claim that such a cutting procedure avoids
damaging the fibers, which is what I expected from the text.
Perhaps a closeup of the cut face that supports this assertion
would be more appropriate? Figure 1 (caption): there is no mention of vaccum or vibration to
compact the tungsten. The caption should be updated to include
relevant details, or something like "see text for details". As it
is I think the caption is incorrect by omission. Figure 3(a): This figure (and the text) are way too small to be
readable. line 100: Next, an acrylic... lines 107-122: This is described again in the test beam section
(almost verbatim), and is really more relevant there than in this
construction section. Delete these lines in favor of the
description in the test beam results section. Figure 4: Is (0,0) the center of the light guide? If so, why the
asymmetry? This caption requires more details to orient the reader
and explain the result. line 131: groups of four by manufacturer measured gain line 158: It would be good here to say something "...to represent
the sPHENIX magnet and cryostat with an equivalent number if
interaction lengths of material." line 165: "...wavelength-shifting (WLS) fiber ... line 169: do you need a reference for Kuraray fibers? Will it be
clear to the reader what this is? lines 200-205: "This process removed microscopic non-uniformities normally present on the surface of the extruded plastic, decreasing aging and improving the ability to withstand pressure without crazing. It also enhances the efficiency of light collection in tiles with embedded WLS fibers. The coated tiles are grooved, then WLS fibers are embedded..." line 244: show a decreasing trend line 265: This resulted in a total... line 268: Sixteen preamplifier boards... line 272: ...all five SiPMs... line 293: In each run, four... line 295: eight line 204-206: This sentence "It monitors..." seems incorrect.
While short-term gain changes from temperature can be monitored
with the LEDs you don't monitor leakage current with this. You
just look at the leakage current. line 331: they are line 345: This is the first place a blower is mentioned. Perhaps somewhere upstream of this sentence you need to say something about a blower being used to provide air cooling for the electronics. line 392: is comprised of lines 392-405: You need to be consistent here and capitalize both
parts of "Interface Board" and "Controller Board", or ideally
neither of these since the use as proper names is a bit stretched.
line 411: ... Two Controllers were used... line 440: for most sPHENIX R&D Figure 17: I think this Figure is confusing as we used an
internal clock for the test beam. It would be better to have a
modified figure that more accurately shows the electronics as used
for the test beam. line 451: RCDAQ allows the capture.... line 472: allows users to... line 485: You can't say the momentum spread depends on beam tune
and then quote a number? How about ".. beam tune, and for the
T1044 was approximately 3%". Also note this is quoted as 3% here,
but the EMCAL section corrects for dp/p of 2%. These need to be
made consistent. Figure 20: Not at all readable Figure 22: This will not reproduce well with all the detail in
the EMCal modules. line 575: Section 7? line 580: ...to produce a combined result. line 583: described in Section 5. line 588: a few percent contamination... Figure 25: I don't think this is too detailed to included line 627: hadrons and muons line 629: .. event, the calibrated EMCal tower energy line 637: The energy resolution... Figure 27: caption last sentence "stem from..." line 650: The energy resolution.... line 655: emphasizes improving... General comment on EMCAL results: In the text we quote resolutions with the beam momentum spread subtracted, but all the plots show it without the subtraction. This is confusing. The plots should be made consistent with what is quoted in the paper. (Figs 28, 29, 31, 32). Figure 29,32: (caption) last sentence. You should ideally repeat the full caption with each figure; the reader shouldn't have to bounce around. As a second-best option you could say "Figure details are the same as in Figure 28." line 709-711: This is thought to be due to.... line 735-736: I think we discussed this in the HCAL meeting at
some point. If this is e/pi, then a plot dividing the two curves
is appropriate, or if it is constant, quote a number. Conclusions: The conclusions section needs to be tightened up. Sentences like
"The calibrations are described in Section 3.4/" are useless in
this context. You quote the energy resolution for the different
calorimeter sections but you should also quote the combined result
as well. You should also state again in the conclusions that the
measured performance is consistent with sPHENIX requirements.
On 9/26/2016 4:55 PM, Ron Belmont
wrote:
Dear HCal and EMCal enthusiasts,
The T-1044 2016 paper is ready to be reviewed by the
experts prior to our release to the collaboration at
large. The draft is linked on the test beam publication
page (https://wiki.bnl.gov/sPHENIX/index.php/T-1044_publication)
and can be accessed directly here: https://wiki.bnl.gov/sPHENIX/images/e/eb/Beam-test-results.pdfPlease note that, as always, editing the Overleaf is not allowed without first consulting with the section chairs as listed on the test beam publication page. In principle we have no customary or official deadline
for comments at this point, in contrast with official
collaboration release, but we do want to release to the
collaboration as soon as possible, so please send comments
not more than two weeks from now (the sooner the better).
-- -----------------------------------------------------
Ron Belmont Postdoctoral Research Associate University of Colorado, Boulder ron.belmont AT colorado.edu ----------------------------------------------------- _______________________________________________ Sphenix-hcal-l mailing list Sphenix-hcal-l AT lists.bnl.gov https://lists.bnl.gov/mailman/listinfo/sphenix-hcal-l --
Contact me: john.lajoie |
-
Re: [Sphenix-hcal-l] T-1044 2016 paper ready for expert comments,
Jamie Nagle, 10/02/2016