Hi Abhisek, Very nice. So a spec on the sigma of the normalized MPV
distribution < 10% should be possible, and as you point out
some of this can be removed by adjusting the voltages. Did you have a change to look at a measure of the width of
the Landau fit to the cosmics? I think this might be a more
relevant spec, as it would tell us something about the variation
of the individual SiPM gains within a tower. John On 5/4/2017 12:02 PM, Abhisek Sen
wrote:
Hi John,
As we discussed in the yesterday's hcal meeting, I
looked into tower variations in the cosmic data from last test
beam. It turns out the 12-13% variation that I quoted was
without subtracting the geometry variation. Once I normalized
the MPVs with respect to the simulation, geometry effect was
taken out. Variation is ~6-7% among 16 towers (raw, no fit). I
attached the plot below. As I mentioned, it can be further
improved by adjusting the bias voltages.
Cheers,
Abhisek
On Mon, May 1, 2017 at 11:02 AM John Lajoie
<lajoie AT iastate.edu> wrote:
Dear HCAL'ers I have been charged by PM to come up with 1-2 "Key
Performance Parameters" (KPPs) for the HCAL. For those
of you not familiar with the DOE-speak on this, KPP's
are low-level performance parameters that are necessary
for the subsystem to ultimately perform its required
role in the physics goals of the experiment. You have to
be able to demonstrate that KPP's have been met prior to
declaring the project complete, so they can't depend on
lots of analysis and calibrations. As an example, you couldn't say "measure hadrons
with a stochiastic term less than 100%/sqrt(E)". That
will require analysis and calibrations that might take a
while after the first data is taken. So what should we use as KPP's for the HCAL? I have
a few ideas, but I would like to hear what others have
to say. For example, we could use something like: -> Tower cosmic ray gain calibrations within 10% tower-to-tower variation (Is this good enough? What did we achieve in the test beam?) Keep in mind we only want 1-2 - NOT a laundry list. We
will be held to every KPP we set down, so they should be
few and well thought-out. Things like the fraction of
live channels have already been called out by Ed at a
higher level, so I am looking for HCAL-specific KPP's. Let me know what you think Regards, --
John Lajoie Professor of Physics Iowa State University
(515) 294-6952 lajoie AT iastate.edu Contact me: john.lajoie sPHENIX-HCal-l mailing list sPHENIX-HCal-l AT lists.bnl.gov https://lists.bnl.gov/mailman/listinfo/sphenix-hcal-l --
Contact me: john.lajoie |
-
[Sphenix-hcal-l] KPP's for HCAL,
John Lajoie, 05/01/2017
-
Re: [Sphenix-hcal-l] KPP's for HCAL,
Abhisek Sen, 05/04/2017
-
Re: [Sphenix-hcal-l] KPP's for HCAL,
John Lajoie, 05/04/2017
-
Re: [Sphenix-hcal-l] KPP's for HCAL,
Abhisek Sen, 05/04/2017
-
Re: [Sphenix-hcal-l] KPP's for HCAL,
John Lajoie, 05/04/2017
-
Re: [Sphenix-hcal-l] KPP's for HCAL,
Barbara Jacak, 05/04/2017
- Re: [Sphenix-hcal-l] KPP's for HCAL, John Lajoie, 05/04/2017
-
Re: [Sphenix-hcal-l] KPP's for HCAL,
Barbara Jacak, 05/04/2017
-
Re: [Sphenix-hcal-l] KPP's for HCAL,
John Lajoie, 05/04/2017
-
Re: [Sphenix-hcal-l] KPP's for HCAL,
Abhisek Sen, 05/04/2017
-
Re: [Sphenix-hcal-l] KPP's for HCAL,
John Lajoie, 05/04/2017
-
Re: [Sphenix-hcal-l] KPP's for HCAL,
Abhisek Sen, 05/04/2017
Archive powered by MHonArc 2.6.24.