Re: [Sphenix-hcal-l] Gain Factors in HCal Simulation
- From: "Huang, Jin" <jhuang AT bnl.gov>
- To: Megan Connors <meganeconnors AT gmail.com>, "Haggerty, John" <haggerty AT bnl.gov>, "sphenix-hcal-l AT lists.bnl.gov" <sphenix-hcal-l AT lists.bnl.gov>
- Subject: Re: [Sphenix-hcal-l] Gain Factors in HCal Simulation
- Date: Fri, 25 Mar 2022 00:50:37 +0000
Hi Megan and all
Yes, in simulation, photo-electron per ADU conversion factors of 32/5 and 16/5 are currently used for inner and outer HCal, respectively, following the original design at the test beam era:
> Now that factor is 32 for both inner and Outer electronics. My conclusion is that we need to update this factor to 32 for outer HCal in the outer HCal simulation.
First a caution : from the preamp block diagram below, changing high gain channel’s gain does not necessarily change the gain for the nominal operation at the low gain channel. Nonetheless, I think the general statement is correct, the photo-electron per ADU conversion factors should be (High gain of 32)/(single photon peak at high gain mode in ADU). Nonetheless, John probably can best comment on it.
One more thing: as you can see, some of the HCal digitization factors in simulation are based on guess work of the past. If we have measurement of the actual value, e.g. pedestal width, gains, photon yield, please make an pull request to have them updated.
Cheers
Jin
______________________________
Jin HUANG
Physicist, Ph.D. Brookhaven National Laboratory Physics Department, Bldg 510 C Upton, NY 11973-5000
Office: 631-344-5898 Cell: 757-604-9946 ______________________________
From: Megan Connors <meganeconnors AT gmail.com>
Hi Jin and John,
I'm hoping you can help clarify a question that came up during the calibration meeting today. Shuhang showed the slide below that includes a 5/16 factor. I believe the 5 comes from the single pixel peak of 5 ADC we measured. The 16 is the factor between high and low gain. The test beam paper we state that the factor is 16 for the outer HCal and 32 for inner. Now that factor is 32 for both inner and Outer electronics. My conclusion is that we need to update this factor to 32 for outer HCal in the outer HCal simulation. Does this all sound correct to you? Any other insight to these factors in the simulation are appreciated.
Best, -Megan |