[Sphenix-magnet-l] probably only 7 hours each time at 4.5 K
- From: "Yip, Kin" <kinyip AT bnl.gov>
- To: "sphenix-magnet-l AT lists.bnl.gov" <sphenix-magnet-l AT lists.bnl.gov>
- Subject: [Sphenix-magnet-l] probably only 7 hours each time at 4.5 K
- Date: Sat, 19 Mar 2016 22:15:37 +0000
Hi,
Thanks to Roberto’s and the Cryo group’s test this afternoon, you may read the powerpoint attached with 3 pictures and probably more revealing are his emails below . In the 2nd page of the powerpoint, you may see the situation when Roberto & Co. got the system to ~4.5 K.
Apparently, they can only keep the system at ~4.5K for 7 hours a day. Then, we need to re-liquefy and refill … I’m asking Roberto how much time this may take. ( It may be like an ERL operation. )
Kin
From: Than, Yatming (Roberto)
Kin, We just finished the test using the subcooler. I tried putting the solenoid return to the 9K return of the plant, from the 25K return, but pressure drop was too high and almost lifted the relief of the VTF. Our return shares the same return as the VTF. so we had the VTF at 19.7 psia.
So supplied the solenoid with the return to the 25K return of the plant.
Consumption rate was 7%/hr. We used 15% for initial cooldown. So starting from 80% drawing to 10%, we have 7 hours of 4.5K time. The level probe we will hav trouble reading stable. Roberto
From: Yip, Kin Hi,
How much “(low-field) test time” are we talking about ? Days or hours ? If it’s hours, I guess we need to coordinate very exactly (so as not to waste) …
Kin
From: Than, Yatming (Roberto)
Kin,
From yesterday afternoon test, I tried running direct from storage dewar feed. But the pressure differential was not enough to get the flow, so I will use subcooler. Not as efficient, so I will go through my storage dewar inventory quicker.
Also when I put it on the cold end of the plant, the heat leak is too large, so we the test time will be dictated by storage dewar inventory.
I will test the subcooler this afternoon/
Roberto
From:
sphenix-magnet-l-bounces AT lists.bnl.gov [sphenix-magnet-l-bounces AT lists.bnl.gov] on behalf of Yip, Kin [kinyip AT bnl.gov] Hi,
From the temperature change and my conversations with the Cryo group, I can see that late last night (9:30 pm or so), they’ve started to use 10+ K He gas to cool the system back to <15 K or so. This afternoon (around 2 pm-ish), they started to use the ~4 K He to cool.
Now, the outlet temperature is fluctuating around 7 K and whereas various thermometers in the magnet were showing around ~5 K or ~6 K. Hopefully, soon it’ll be close to 4-5 K …. I understand that they’ll let the coils soaked in the bath of ~4.5 K He so that hopefully the temperature deep inside the metallic coil will be all ~ 4.5K. Sunday or so, they’ll open the valves for the “phase separator” and then there will be a circulating loop (of liquid He) between the phase-separator and the magnet coil. This is my rough/probably inaccurate understanding of what it’s going on and what it’s going to happen. Thanks for taking the risk to read up to here, hahaha …… J
Kin |