[Sphenix-magnet-l] voltage taps/current-limiting resistance on the extension ?
- From: Kin Yip <kinyip AT bnl.gov>
- To: "sphenix-magnet-l AT lists.bnl.gov" <sphenix-magnet-l AT lists.bnl.gov>
- Subject: [Sphenix-magnet-l] voltage taps/current-limiting resistance on the extension ?
- Date: Wed, 6 Jul 2016 12:28:48 -0400
Hi, Thanks Bob for doing the calculation. Bob asked me the issue of NEW voltage taps and any current-limiting resistance for the extension that we're going to build ? Do we already have new voltage taps and current-limiting resistance for the extension ? I heard people mention it but I'm not sure whether they're already built-in. I understand that they're easy to add when we do the assembly but it'll be a shame if we somehow forget about them. ... Paul just told me that he didn't include them ... but maybe, I'll look for Piyush (or who ??) to make sure that we have the taps/current-llimiting resistance when we do the assembly ... Kin -------- Forwarded Message -------- Subject: RE: Closing this ticket this way ?? Date: Wed, 6 Jul 2016 10:56:12 -0400 From: Lambiase, Robert <lambiase AT bnl.gov> To: Kovach, Paul E <kovach AT bnl.gov> CC: Yip, Kin <kinyip AT bnl.gov> This time with the attachment.
Bob
--------------------
Hi Paul,
I’d still like the sketch I asked for below, But I can say the lead design is very conservative. · With a current of 4.7kA and a time constant of 33 seconds, the Miits requirement is 4.72 x 33 / 2 = 364 Miits · The smallest cross section of the stabilizing copper is 46mm x 12.7mm. This gives a cross section area of 4.6 x 1.27 = 5.84cm2. · The cross section of RHIC 30 strand superconductor (dipole cable) is 0.1166cm x 0.973cm. Two are being used, so the area is 0.227cm2. I remember that the 23 strand cable had a critical current of 11kA in 0T field, so two 30 strand cables might be overkill. It won’t hurt anything to have two. · According to the calculation, 364Miits will bring the temperature up to about 20K. That is a big hunk of copper you have there.
I have given Kin the basis for my calculations.
Bob
------------
Hi Paul,
Because I’ve been away from this for a bit, could provide a little more information?
It would be ideal for me if you had a sketch which showed each segment from the coil to the outside world.
Each of the segments in this sequence would show what superconductor is used (where superconductor is used), the cross sectional area of the stabilizer, where the voltage taps are, and whether each of these segments were in place for the 100 Amp test.
This would give me a better idea of all the pieces are, allow me to correlate the links to the 100 Amp test data, and make sure all the inputs to the quench detector are accounted for.
Thanks,
Bob
From:
Kovach, Paul E
The jpeg called “sphenix extension lead bus” is the copper busses that carry the superconductor through the extension. They are the same dimensions as the copper leads in the Ansaldo lead box and have the superconductor soldered along their length. They are 46mm x 15mm cross section. The flags on the end are 46mm x 50mm lap area as in the valve box. The JPEG’s called “sphenix lead bus jumpers” and “sphenix lead bus” are the copper jumpers that connect the lead bus extension to the valve box leads at the far end. Their cross section is 46mm x 12.7mm. The JPEG called “sphenix lead assy” shows the laminated copper flex leads at the magnet end of the extension connected to the lead extension bus. Their cross section is 46mm x 15mm. The JPEG called “sphenix copper flex bus” shows the flex leads connecting the extension leads to the magnet flags. What this means is that all the copper busses connecting the original magnet lead flags to the original lead box flags are 46mm x 15mm cross section except for the short jumpers at the lead box end of the extension which are 46 x 12.7mm cross section. All the lap joints are the same as the original Ansaldo joints (46mm x 50mm). The super conductor is doubled RHIC dipole superconductor. Roberto has assured us that there will be no temperature gradient along the leads.
From:
Kin Yip [mailto:kinyip AT bnl.gov]
Hi Paul, On 07/01/2016 03:00 PM, Kin Yip wrote:
|
Attachment:
Lead Calculation.docx