sphenix-magnet-l - [Sphenix-magnet-l] some limits in the procedure ?

sphenix-magnet-l AT lists.bnl.gov

Subject: sPHENIX discussion of the superconducting solenoid

List archive

Chronological Thread  
  • From: Kin Yip <kinyip AT bnl.gov>
  • To: "Muratore, Joseph" <muratore AT bnl.gov>
  • Cc: "sphenix-magnet-l AT lists.bnl.gov" <sphenix-magnet-l AT lists.bnl.gov>
  • Subject: [Sphenix-magnet-l] some limits in the procedure ?
  • Date: Wed, 21 Jun 2017 15:27:23 -0400

Hi Joe,

Don Lynch suggested that you put down some limiting statements in the procedure such as :

No polarity change
No current higher than (say) 4850 A
Ramp rate < (say) 3.3 A/s

So that we won't go over these limits during the time of testing.    What do you think ??

The above nos. are just examples and I've based on the document below and your run plan.

In 
Volume 5.1 Design and Quality Control Report GENERAL MAIN COMPONENTS.pdf
\\bnl.gov\c-adnas\babar\Volume 5 DESIGN AND QUALITY CONTROL REPORT\5.1 GENERAL AND MAIN COMPONENTS

they did their "first" real test (inside BaBar detector),   apart from the slower ramp rate at the top like 0.5 A/s (from 4400 to 4610 A),

they also mentioned that they had tried at ~4830 (or 4826) A and held it for 14 hours [ pdf page 25 ].

They also measured that at 1.7A/s, the temp. rise was 0.1K ; at 2.7A/s, the temperature rise was 0.2 K :-)


Kin