Subject: sPHENIX discussion of the superconducting solenoid
List archive
Re: [Sphenix-magnet-l] Brieif Summary of today's Magnet meeting (Mar. 27, 2019) at 902A conference room
- From: "Phillips, David B" <phillips AT bnl.gov>
- To: "Yip, Kin" <kinyip AT bnl.gov>, "sphenix-magnet-l AT lists.bnl.gov" <sphenix-magnet-l AT lists.bnl.gov>
- Subject: Re: [Sphenix-magnet-l] Brieif Summary of today's Magnet meeting (Mar. 27, 2019) at 902A conference room
- Date: Thu, 28 Mar 2019 17:39:45 +0000
For item 7, I assume that there will be entry to the IR with the magnet at full field (except for anyone with medical implants affected by magnetic fields). My concern was for time delay waiting for “magnet off” before opening a pole tip (maybe this isn’t a concern).
Some details for item 5: Estimates of resistance of connections between the Power Supply & Magnet (water cooled bus and cables), I tried to be conservative with lengths and temperature of conductor so actual resistance may be slightly lower. There are 4 pairs of PHENIX water cooled bus (WCB) from 1008B to the IR (2 pairs 2” schedule 80 and 2 pairs 2” schedule 160 copper pipe). I considered two different paths to reach the magnet, #1 jump from the west wall of the IR to the magnet platform (basically follow the cryo lines), #2 go to the floor (below the rails) and up the side of the platform. All of these estimates of resistance include 535kmcm (12 in parallel) from the Power Supply to the Dump Resistor, from the Dump Resistor to the WCB, from the building WCB to the Platform WCB and from the platform WCB to the Magnet Leads (total resistance of these 535kmcm jumpers is 0.13mOhms). For path #1a using 2 pairs of existing schedule 160 bus (modified at the ends of course) from 1008B to the west wall of the IR and then one new pair of schedule 160 bus from the wall to the magnet the total resistance is 0.73mOhms. For path #1b using 1 pair of schedule 160 bus and 1 pair of schedule 80 bus from 1008b to the west wall of the IR and then one new pair of schedule 160 bus from the wall to the magnet the total resistance is 0.86mOhms. For path #2 using 2 pairs of existing schedule 160 bus (modified at the ends of course) from 1008B to the base of the platform and then one new pair of schedule 160 bus up the platform the total resistance is 0.92mOhms.
I assume we want to go with option #1a? Kin – can you tell us the ramp down time for 0.73mOhms? This resistance could be lowered by using a lower resistance bus from the West wall to the Magnet, or by adding more 535kmcm jumpers (in parallel). The resistance could be raised (within limits) by removing 535kmcm jumpers. If we want the option to add a resistor (to reduce ramp down time) we should have a plan and leave space for it now. I would like to start the modifications to the WCB in 1008B soon (during this RHIC Run), so we need to solidify a plan.
Attached is the proposed layout of the PS & Dump Resistor on the mezzanine in 1008B.
Dave
From: sPHENIX-magnet-l <sphenix-magnet-l-bounces AT lists.bnl.gov>
On Behalf Of Yip, Kin
Powerpoint (Carl Schultheiss) : https://indico.bnl.gov/event/5935/
connectors at the bottom of the doghouse/junction-box attached to the Valvebox, he proposed that we re-route some of the voltage tap signals from the
another 2 days to close it back. All in all, it may be a week of his time. He’ll send us an email when he can find a time to do it. This is OK as it’s not really that urgent.
Kin
|
Attachment:
PS Layout 2018-03-28.pdf
Description: PS Layout 2018-03-28.pdf
-
[Sphenix-magnet-l] Brieif Summary of today's Magnet meeting (Mar. 27, 2019) at 902A conference room,
Yip, Kin, 03/27/2019
- Re: [Sphenix-magnet-l] Brieif Summary of today's Magnet meeting (Mar. 27, 2019) at 902A conference room, Phillips, David B, 03/28/2019