Dear Glenn,
I go to our LogView and plot the temperatures that I’ve been plotting (see recovery.gif). It seems that those thermal sensors went back to their nominal temperatures within about an hour or so.
We haven’t done 1A/s so far and we’ll try to do it today. Yesterday, we did see the gas-cooled voltage rising from 2 mV to > 20mV. But as I mentioned in the email, the Cryo was reading wrong current etc. and they were trying to make changes yesterday. They will try to optimize today.
John Haggerty and Frank Toldo placed a hall probe on a shaft attached to the EMCAL. This is what he told me:
“BTW, my Hall probe showed 1.25 T at 4620 A, and the map at that location gives 1.39 T, so I see a field that's 10% low, although we have not calibrated the probe. “
{ Power Supply needs to give 4620 A in order to produce 4596 A in the DCCT/Magnet. Some current went to the dump resistor …}
Kin
From: Glenn Young <glennyoung82251 AT gmail.com>
Dear Kin, Ah, this is very exciting. How long did it take to recool after the 'quench' from full current, where you say the temperature went to 35K
Is there any noticeable change in helium consumption when ramping at the faster rate, 1.0 vs 0.5 A/s ? I am hoping the lead temperature at the vapor cooled leads stays within good limits then?
Is there anywhere an NMR probe to measure field, or an installed Hall probe in the bore? Thanks, Glenn
On Mon, Oct 24, 2022 at 6:57 PM Yip, Kin <kinyip AT bnl.gov> wrote:
|
Attachment:
recovery.gif
Description: recovery.gif
Attachment:
GC-voltages.gif
Description: GC-voltages.gif
-
[Sphenix-magnet-l] Magnet reached top current (4596 A) ... Tiny magnetic field at the top platform ... He-Gas-flow needs to be adjusted,
Yip, Kin, 10/24/2022
- Re: [Sphenix-magnet-l] Magnet reached top current (4596 A) ... Tiny magnetic field at the top platform ... He-Gas-flow needs to be adjusted, Sondheim, Walter E, 10/24/2022
- Re: [Sphenix-magnet-l] Magnet reached top current (4596 A) ... Tiny magnetic field at the top platform ... He-Gas-flow needs to be adjusted, Edward O'Brien, 10/24/2022
-
Message not available
-
Message not available
- [Sphenix-magnet-l] Cryo seems to recover within in an hour ..., Yip, Kin, 10/25/2022
-
Message not available
Archive powered by MHonArc 2.6.24.