Re: [Sphenix-magnet-l] magnet current suddenly dropped ~100 A since 15:47 --- plan to fix the DCCT next Maintenance Day (June 20)
- From: "Yip, Kin" <kinyip AT bnl.gov>
- To: "sphenix-magnet-l AT lists.bnl.gov" <sphenix-magnet-l AT lists.bnl.gov>
- Subject: Re: [Sphenix-magnet-l] magnet current suddenly dropped ~100 A since 15:47 --- plan to fix the DCCT next Maintenance Day (June 20)
- Date: Tue, 4 Jun 2024 22:26:18 +0000
Hello, Almost forgot to let this mailing list know what we've found .... When I went to 1008B mezzanine to read the current from the DCCT
But the Cryo's gas flow is based on this Magnet-Current --- hard-wired (Brian told me). So, when this Magnet-Current dropped, the gas flow also dropped which resulted in the rises of voltage-tap (voltage) values. Perhaps, more importantly, not just other current readings have not changed, the Hall Probes that John Haggerty has installed inside the Magnet We can't/shouldn't fix the DCCT while the magnet is in top current/field. Pablo and his team are busy tomorrow for other works that have been planned. Kin On 6/4/24 00:42, Yip, Kin wrote: Hello, Carl, Pablo, and Winston,Just before I left, I noticed that the so-called "Magnet Current" dropped instantly by almost 90-100 A to only ~4500 A (instead of close to ~4596 A).This has been the case since 15:47 this afternoon.I show you the plot in the 180Hz data. But all the currents stay about the same, including qfg.8b-sphenix.B0.Winston: is it possible that a factor suddenly got changed ???Tomorrow, I'll go up to the 1008B mezzanine to take a look at the DCCT meter.KinPS:I checked John Haggerty's hall probes inside the Magnet bore which show no such step. |
Attachment:
20240604_084840.jpg
Description: 20240604_084840.jpg
Attachment:
Tue_Jun__4_090850_2024.2137952.png
Description: Tue_Jun__4_090850_2024.2137952.png
-
[Sphenix-magnet-l] magnet current suddenly dropped ~100 A since 15:47,
Yip, Kin, 06/04/2024
- Re: [Sphenix-magnet-l] magnet current suddenly dropped ~100 A since 15:47 --- plan to fix the DCCT next Maintenance Day (June 20), Yip, Kin, 06/04/2024
- Re: [Sphenix-magnet-l] magnet current suddenly dropped ~100 A since 15:47 --- plan to fix the DCCT next Maintenance Day (June 20), Yip, Kin, 06/04/2024