Hi Xu, Zhenyu,
I have a question about the signal size vs. bias. On slide 13 in the latest, we see as before from the noise vs. bias plot that the detector reaches full depletion around 100 V. And we see from the "Cluster Signal vs. HV" that the signal is pretty constant above 100 V, at least for inner detector (outer ??). The inner case shows less than 2% increase in signal from 100 to 150 V, which seems quite reasonable (AFAIK; I could be wrong). Whereas the old plot that you showed, in https://indico.bnl.gov/event/8967/contributions/39814/attachments/29461/45813/FstModuleAssemblyandTest_BNLF2F_Jul2020.pdf slide 10, showed again a full depletion at 100 V from the capacitance (noise) plot, but at least 15-20% extra signal when raising the bias above 100 V. I never understood that result, and it seemed rather worrisome (if correct). Is there any reason to think there might have been a problem with that old measurement and we should discount it? And from the new results, would you agree the bias can be set around 130 V or so both for inner and outer and it should be good?
Sincerely,

Gerard


On 7/30/2020 9:14 PM, Xu Sun wrote:
Hi Flemming,

I have made some updates and you could find the updated slides in the following link:

https://drupal.star.bnl.gov/STAR/system/files/FstReviewAug2020_v9.pdf

Best,

Xu





Archive powered by MHonArc 2.6.24.

Top of Page