Skip to Content.
Sympa Menu

star-fst-l - [Star-fst-l] FST cooling system, EIT-D communication probllem

Please Wait...

star-fst-l AT lists.bnl.gov

Subject: Star-fst-l mailing list

List archive

Chronological Thread  
  • From: "Hu, Yu" <yuhu AT bnl.gov>
  • To: "Wilks, Gavin M" <gwilks3 AT uic.edu>, "star-fst-l AT lists.bnl.gov" <star-fst-l AT lists.bnl.gov>
  • Cc: "Betts, Wayne" <wbetts AT bnl.gov>, "Tribedy, Prithwish" <ptribedy AT bnl.gov>, "Videbaek, Christian" <cvidebaek AT bnl.gov>, "Videbaek, Flemming" <videbaek AT bnl.gov>, Zhenyu Ye <yezhenyu AT uic.edu>, Xu Sun <xusun AT impcas.ac.cn>, "Camarda, Timothy" <tcamarda AT bnl.gov>
  • Subject: [Star-fst-l] FST cooling system, EIT-D communication probllem
  • Date: Thu, 21 Mar 2024 20:46:27 +0000

Dear All,

As a short report for today. We went to STAR this afternoon, and with the help of Tim we identified the problem is really from the OMEGA EIT-D (Serial to Ethernet/Internet module). Tim has removed the broken module and will help us reinstall a newer one once it is delivered. (for more details of the problem, please see the history below this email)
We have turned off the power of the FST Cooling Rack. Please don't turn it on before the newer OMEGA EIT-D module is reinstalled. 

Thank you.
Best,
Yu

From: Hu, Yu <yuhu AT bnl.gov>
Sent: Wednesday, March 20, 2024 19:06
To: Wilks, Gavin M <gwilks3 AT uic.edu>; Camarda, Timothy <tcamarda AT bnl.gov>
Cc: Betts, Wayne <wbetts AT bnl.gov>; Tribedy, Prithwish <ptribedy AT bnl.gov>; Videbaek, Christian <cvidebaek AT bnl.gov>; Videbaek, Flemming <videbaek AT bnl.gov>; Zhenyu Ye <yezhenyu AT uic.edu>; Xu Sun <xusun AT impcas.ac.cn>
Subject: Re: [Star-ops-l] Power dip
 
Dear All,

Just to keep everyone on the same page. 

We went to STAR today, and power cycled the FST cooling rack. We still see the same problem: everything seems running well (from the cooling rack panel), but a couple of PVs are missing (shown 0) from the FST GUI. The 130.199.61.24 address has no response, while the 130.199.61.23 shows us the configure page correctly. 
Two possible problems: 1) the ethernet cable or switch is not working; 2) the network module is not working. We will go to STAR tomorrow with Tim, and switch the net cables to see which one is the real problem.
Then if the problem is the network module (Omega EIT-D), we will first check if it is because it is configured to the default setup, so the IP we set is not working anymore. Then, if none of this works, we will order a new one. 

I will post what we found after tomorrow's investigation. 

Thank you.

Best,
Yu

From: Gavin Wilks <gwilks3 AT uic.edu>
Sent: Tuesday, March 19, 2024 8:59
To: Hu, Yu <yuhu AT bnl.gov>
Cc: Betts, Wayne <wbetts AT bnl.gov>; Tribedy, Prithwish <ptribedy AT bnl.gov>; Videbaek, Christian <cvidebaek AT bnl.gov>; Videbaek, Flemming <videbaek AT bnl.gov>; Zhenyu Ye <yezhenyu AT uic.edu>; Xu Sun <xusun AT impcas.ac.cn>
Subject: Re: [Star-ops-l] Power dip
 
Hello Yu,

Thank you for your help.

I checked that the cooling connections for the 3 sets of patch panel boards were secure. I was informed at the end of February that the MCW was running through them. I saw no leaks. 

Also, at this point we can no longer visually inspect connections on the west TPC face since the west pole tip is closed. 

Best,
Gavin

On Mar 19, 2024, at 10:44, Hu, Yu <yuhu AT bnl.gov> wrote:

Hi Gavin, and All,

Thank you, I will be there tomorrow and power cycle the crate.
I remember I could reset the system by our own (in-person, manual page 12), I will give it a try. 
For power-cycle the crate, I need some help from our STAR engineers. 

Another thing, do you know if the cooling for the patch panels are connected and running? One loop was not connected at the beginning of the last years run, and crated a lot of problems last time. We certainly want to double check this year. 

Will post the details to this list and a summary to the ops mail.

Thank you.
Best,
Yu

Link to the manual:


From: Gavin Wilks <gwilks3 AT uic.edu>
Sent: Monday, March 18, 2024 22:30
To: Hu, Yu <yuhu AT bnl.gov>
Cc: Betts, Wayne <wbetts AT bnl.gov>; Tribedy, Prithwish <ptribedy AT bnl.gov>; Videbaek, Christian <cvidebaek AT bnl.gov>; Videbaek, Flemming <videbaek AT bnl.gov>; Zhenyu Ye <yezhenyu AT uic.edu>; Xu Sun <xusun AT impcas.ac.cn>
Subject: Re: [Star-ops-l] Power dip
 
Hello Yu,

Yes, the FST cooling is off now. I have not checked anything new regarding the communication problem. 

If you have time, it would be great if you could try power-cycling the crate and resetting the interlock. 

Thanks,
Gavin 

On Mar 19, 2024, at 00:24, Gavin Wilks <gwilks3 AT uic.edu> wrote:

Hello Yu,

The cooling

On Mon, Mar 18, 2024 at 22:32 Hu, Yu <yuhu AT bnl.gov> wrote:
Hi Gavin,

I was wondering, is the FST cooling off now? And is there anything else you checked since the last time we discussed the FST Cooling module (Omega unit) communication problem? 
If it is ok, I might try to power-cycle the crate (will also need help from Christian to reset the rack interlock since there was a power dip) at some point this week. I'll be on-site for the collaboration meeting for this week, hope we can figure this out. 


Thank you.
Best,
Yu

From: Star-ops-l <star-ops-l-bounces AT lists.bnl.gov> on behalf of videbaek via Star-ops-l <star-ops-l AT lists.bnl.gov>
Sent: Monday, March 18, 2024 16:06
To: STAR experiment operations mailing list <star-ops-l AT lists.bnl.gov>
Cc: Videbaek, Flemming <videbaek AT bnl.gov>
Subject: Re: [Star-ops-l] Power dip
 
Hi Alexei,

could you check tomorow and see if the FST cooling is up running?
It was left that way when Gavin left.

best Flemming


On 2024-03-18 09:04, Lebedev, Alexei via Star-ops-l wrote:
> Dear All,
>  Today we had a power dip in RHIC about 8AM. We lost Modified Chill
> Water and TPC water. I put MCW in by-pass mode to preserve power on
> platform. I switched OFF only  TCD crate, all others were OK. About 15
> min ago MCW and TPC water are restarted.
>  Now all by-passes removed to normal and TCD crate is ON.
>  Best regards, Alexei
> _______________________________________________
> Star-ops-l mailing list
> Star-ops-l AT lists.bnl.gov
> https://lists.bnl.gov/mailman/listinfo/star-ops-l

-- 
Flemming Videbaek
senior scientist, emeritus
videbaek @ bnl.gov
Brookhaven National Lab
Physics Department
Bldg 510D
Upton, NY 11973

phone: 631-344-4106
cell     :  631-681-1596
_______________________________________________
Star-ops-l mailing list
Star-ops-l AT lists.bnl.gov
https://lists.bnl.gov/mailman/listinfo/star-ops-l




Archive powered by MHonArc 2.6.24.

Top of Page