| Hello All, 
 As promised at Thursday’s meeting, I have attached the pfRICH LUTs (e/pi/k/p and pi/k/p) I sent to the S&C group. Also attached is a powerpoint detailing some spot checks done to look at momentum, eta, and phi dependencies, B-field effects, threshold settings, and momentum resolution effects. 
 Thanks. 
 Brian 
 
 | 
Attachment:
pfRICHmeeting_LUTpreview_24-02-29_v1.pptx
Description: pfRICHmeeting_LUTpreview_24-02-29_v1.pptx
## Eta Bins: [-3.8, -3.5], [-3.5, -3.0], [-3.0, -2.8], [-2.8, -2.0], [-2.0,
-1.5]
## Momentum Bins [GeV]: [0.05, 0.1] : then 0.1 to 1.0 in 0.1 wide steps :
then 1.0 to 15.0 in 1.0 wide steps
## NOTE: Particles thrown flat within Eta and Momentum Bins - results
reported at bin center
## NOTE: No charge scan - only thrown electrons, pi+, K+, and proton - assume
charge symmetry at this point
## NOTE: Particles thrown flat in Phi in this run - Phi value is a dummy
## NOTE: Rows with all zero probability indicate PID not possible
## Format: PID Charge Momentum Theta Phi prob-electron prob-pion prob-kaon
prob-proton
11 -1 0.08 3.0641 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.15 3.0641 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.25 3.0641 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.35 3.0641 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.45 3.0641 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.55 3.0641 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.65 3.0641 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.75 3.0641 0.00 0.995492 0.004508 0.000000 0.000000
11 -1 0.85 3.0641 0.00 0.999989 0.000011 0.000000 0.000000
11 -1 0.95 3.0641 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 1.50 3.0641 0.00 0.681747 0.318253 0.000000 0.000000
11 -1 2.50 3.0641 0.00 0.890893 0.109107 0.000000 0.000000
11 -1 3.50 3.0641 0.00 0.798450 0.201550 0.000000 0.000000
11 -1 4.50 3.0641 0.00 0.703394 0.296606 0.000000 0.000000
11 -1 5.50 3.0641 0.00 0.642122 0.357878 0.000000 0.000000
11 -1 6.50 3.0641 0.00 0.600882 0.399118 0.000000 0.000000
11 -1 7.50 3.0641 0.00 0.573328 0.426672 0.000000 0.000000
11 -1 8.50 3.0641 0.00 0.558285 0.441715 0.000000 0.000000
11 -1 9.50 3.0641 0.00 0.546135 0.453865 0.000000 0.000000
11 -1 10.50 3.0641 0.00 0.537807 0.462193 0.000000 0.000000
11 -1 11.50 3.0641 0.00 0.531525 0.468475 0.000000 0.000000
11 -1 12.50 3.0641 0.00 0.526199 0.473801 0.000000 0.000000
11 -1 13.50 3.0641 0.00 0.522887 0.477113 0.000000 0.000000
11 -1 14.50 3.0641 0.00 0.519599 0.480401 0.000000 0.000000
11 -1 0.08 3.0317 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.15 3.0317 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.25 3.0317 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.35 3.0317 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.45 3.0317 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.55 3.0317 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.65 3.0317 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.75 3.0317 0.00 0.991441 0.008559 0.000000 0.000000
11 -1 0.85 3.0317 0.00 0.999965 0.000035 0.000000 0.000000
11 -1 0.95 3.0317 0.00 0.999999 0.000001 0.000000 0.000000
11 -1 1.50 3.0317 0.00 0.670976 0.329024 0.000000 0.000000
11 -1 2.50 3.0317 0.00 0.875520 0.124480 0.000000 0.000000
11 -1 3.50 3.0317 0.00 0.799781 0.200219 0.000000 0.000000
11 -1 4.50 3.0317 0.00 0.695511 0.304489 0.000000 0.000000
11 -1 5.50 3.0317 0.00 0.638672 0.361328 0.000000 0.000000
11 -1 6.50 3.0317 0.00 0.594468 0.405532 0.000000 0.000000
11 -1 7.50 3.0317 0.00 0.570281 0.429719 0.000000 0.000000
11 -1 8.50 3.0317 0.00 0.554746 0.445254 0.000000 0.000000
11 -1 9.50 3.0317 0.00 0.545980 0.454020 0.000000 0.000000
11 -1 10.50 3.0317 0.00 0.538913 0.461087 0.000000 0.000000
11 -1 11.50 3.0317 0.00 0.531874 0.468126 0.000000 0.000000
11 -1 12.50 3.0317 0.00 0.526531 0.473469 0.000000 0.000000
11 -1 13.50 3.0317 0.00 0.522194 0.477806 0.000000 0.000000
11 -1 14.50 3.0317 0.00 0.519484 0.480516 0.000000 0.000000
11 -1 0.08 2.9607 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.15 2.9607 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.25 2.9607 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.35 2.9607 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.45 2.9607 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.55 2.9607 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.65 2.9607 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.75 2.9607 0.00 0.991409 0.008591 0.000000 0.000000
11 -1 0.85 2.9607 0.00 0.999678 0.000322 0.000000 0.000000
11 -1 0.95 2.9607 0.00 0.999999 0.000001 0.000000 0.000000
11 -1 1.50 2.9607 0.00 0.683759 0.316241 0.000000 0.000000
11 -1 2.50 2.9607 0.00 0.888959 0.111041 0.000000 0.000000
11 -1 3.50 2.9607 0.00 0.810360 0.189640 0.000000 0.000000
11 -1 4.50 2.9607 0.00 0.706990 0.293010 0.000000 0.000000
11 -1 5.50 2.9607 0.00 0.638739 0.361261 0.000000 0.000000
11 -1 6.50 2.9607 0.00 0.601370 0.398630 0.000000 0.000000
11 -1 7.50 2.9607 0.00 0.575178 0.424822 0.000000 0.000000
11 -1 8.50 2.9607 0.00 0.560903 0.439097 0.000000 0.000000
11 -1 9.50 2.9607 0.00 0.546099 0.453901 0.000000 0.000000
11 -1 10.50 2.9607 0.00 0.537575 0.462425 0.000000 0.000000
11 -1 11.50 2.9607 0.00 0.534210 0.465790 0.000000 0.000000
11 -1 12.50 2.9607 0.00 0.525851 0.474149 0.000000 0.000000
11 -1 13.50 2.9607 0.00 0.523914 0.476086 0.000000 0.000000
11 -1 14.50 2.9607 0.00 0.519721 0.480279 0.000000 0.000000
11 -1 0.08 2.7975 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.15 2.7975 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.25 2.7975 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.35 2.7975 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.45 2.7975 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.55 2.7975 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.65 2.7975 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 0.75 2.7975 0.00 0.994089 0.005911 0.000000 0.000000
11 -1 0.85 2.7975 0.00 0.999883 0.000117 0.000000 0.000000
11 -1 0.95 2.7975 0.00 1.000000 0.000000 0.000000 0.000000
11 -1 1.50 2.7975 0.00 0.664001 0.335999 0.000000 0.000000
11 -1 2.50 2.7975 0.00 0.893452 0.106548 0.000000 0.000000
11 -1 3.50 2.7975 0.00 0.786474 0.213526 0.000000 0.000000
11 -1 4.50 2.7975 0.00 0.688065 0.311935 0.000000 0.000000
11 -1 5.50 2.7975 0.00 0.627914 0.372086 0.000000 0.000000
11 -1 6.50 2.7975 0.00 0.591617 0.408383 0.000000 0.000000
11 -1 7.50 2.7975 0.00 0.567638 0.432362 0.000000 0.000000
11 -1 8.50 2.7975 0.00 0.554591 0.445409 0.000000 0.000000
11 -1 9.50 2.7975 0.00 0.543559 0.456441 0.000000 0.000000
11 -1 10.50 2.7975 0.00 0.534616 0.465384 0.000000 0.000000
11 -1 11.50 2.7975 0.00 0.530040 0.469960 0.000000 0.000000
11 -1 12.50 2.7975 0.00 0.524437 0.475563 0.000000 0.000000
11 -1 13.50 2.7975 0.00 0.521079 0.478921 0.000000 0.000000
11 -1 14.50 2.7975 0.00 0.517948 0.482052 0.000000 0.000000
211 1 0.08 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.15 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.25 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.35 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.45 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.55 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.65 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.75 3.0641 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.85 3.0641 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.95 3.0641 0.00 0.000000 1.000000 0.000000 0.000000
211 1 1.50 3.0641 0.00 0.016788 0.983212 0.000000 0.000000
211 1 2.50 3.0641 0.00 0.048739 0.951261 0.000000 0.000000
211 1 3.50 3.0641 0.00 0.189927 0.702266 0.107807 0.000000
211 1 4.50 3.0641 0.00 0.293040 0.698541 0.008419 0.000000
211 1 5.50 3.0641 0.00 0.361304 0.636617 0.002079 0.000000
211 1 6.50 3.0641 0.00 0.396557 0.598400 0.005044 0.000000
211 1 7.50 3.0641 0.00 0.423920 0.555457 0.020622 0.000000
211 1 8.50 3.0641 0.00 0.440241 0.511597 0.048162 0.000000
211 1 9.50 3.0641 0.00 0.452412 0.454278 0.093310 0.000000
211 1 10.50 3.0641 0.00 0.462423 0.400326 0.137252 0.000000
211 1 11.50 3.0641 0.00 0.467885 0.353633 0.178482 0.000000
211 1 12.50 3.0641 0.00 0.472988 0.298010 0.229002 0.000000
211 1 13.50 3.0641 0.00 0.476829 0.267012 0.256159 0.000000
211 1 14.50 3.0641 0.00 0.480039 0.230209 0.289753 0.000000
211 1 0.08 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.15 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.25 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.35 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.45 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.55 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.65 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.75 3.0317 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.85 3.0317 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.95 3.0317 0.00 0.000000 1.000000 0.000000 0.000000
211 1 1.50 3.0317 0.00 0.022759 0.977241 0.000000 0.000000
211 1 2.50 3.0317 0.00 0.058836 0.941164 0.000000 0.000000
211 1 3.50 3.0317 0.00 0.186991 0.678454 0.134556 0.000000
211 1 4.50 3.0317 0.00 0.301182 0.690263 0.008555 0.000000
211 1 5.50 3.0317 0.00 0.357021 0.638295 0.004685 0.000000
211 1 6.50 3.0317 0.00 0.404423 0.588301 0.007276 0.000000
211 1 7.50 3.0317 0.00 0.425404 0.547659 0.026937 0.000000
211 1 8.50 3.0317 0.00 0.440781 0.507026 0.052193 0.000000
211 1 9.50 3.0317 0.00 0.457880 0.439216 0.102904 0.000000
211 1 10.50 3.0317 0.00 0.463613 0.398973 0.137414 0.000000
211 1 11.50 3.0317 0.00 0.469810 0.338345 0.191845 0.000000
211 1 12.50 3.0317 0.00 0.473469 0.297448 0.229083 0.000000
211 1 13.50 3.0317 0.00 0.476985 0.256934 0.266081 0.000000
211 1 14.50 3.0317 0.00 0.480516 0.230900 0.288584 0.000000
211 1 0.08 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.15 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.25 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.35 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.45 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.55 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.65 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.75 2.9607 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.85 2.9607 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.95 2.9607 0.00 0.000000 1.000000 0.000000 0.000000
211 1 1.50 2.9607 0.00 0.017183 0.982817 0.000000 0.000000
211 1 2.50 2.9607 0.00 0.050698 0.949302 0.000000 0.000000
211 1 3.50 2.9607 0.00 0.189640 0.634629 0.175731 0.000000
211 1 4.50 2.9607 0.00 0.291794 0.698213 0.009993 0.000000
211 1 5.50 2.9607 0.00 0.361261 0.636471 0.002268 0.000000
211 1 6.50 2.9607 0.00 0.392612 0.600439 0.006949 0.000000
211 1 7.50 2.9607 0.00 0.423403 0.553172 0.023426 0.000000
211 1 8.50 2.9607 0.00 0.440998 0.497910 0.061092 0.000000
211 1 9.50 2.9607 0.00 0.451533 0.463149 0.085318 0.000000
211 1 10.50 2.9607 0.00 0.461962 0.400790 0.137247 0.000000
211 1 11.50 2.9607 0.00 0.467518 0.343759 0.188724 0.000000
211 1 12.50 2.9607 0.00 0.472515 0.306623 0.220862 0.000000
211 1 13.50 2.9607 0.00 0.476238 0.261936 0.261826 0.000000
211 1 14.50 2.9607 0.00 0.480398 0.227517 0.292085 0.000000
211 1 0.08 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.15 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.25 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.35 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.45 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.55 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.65 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
211 1 0.75 2.7975 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.85 2.7975 0.00 0.000000 1.000000 0.000000 0.000000
211 1 0.95 2.7975 0.00 0.000000 1.000000 0.000000 0.000000
211 1 1.50 2.7975 0.00 0.026217 0.973783 0.000000 0.000000
211 1 2.50 2.7975 0.00 0.065596 0.934404 0.000000 0.000000
211 1 3.50 2.7975 0.00 0.216132 0.641474 0.142394 0.000000
211 1 4.50 2.7975 0.00 0.304614 0.685826 0.009560 0.000000
211 1 5.50 2.7975 0.00 0.370643 0.624327 0.005030 0.000000
211 1 6.50 2.7975 0.00 0.402322 0.589077 0.008601 0.000000
211 1 7.50 2.7975 0.00 0.430427 0.536427 0.033146 0.000000
211 1 8.50 2.7975 0.00 0.446339 0.484181 0.069479 0.000000
211 1 9.50 2.7975 0.00 0.460345 0.423780 0.115875 0.000000
211 1 10.50 2.7975 0.00 0.464386 0.360969 0.174644 0.000000
211 1 11.50 2.7975 0.00 0.470976 0.308985 0.220039 0.000000
211 1 12.50 2.7975 0.00 0.474858 0.279840 0.245302 0.000000
211 1 13.50 2.7975 0.00 0.479157 0.250196 0.270648 0.000000
211 1 14.50 2.7975 0.00 0.482906 0.220787 0.296308 0.000000
321 1 0.08 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.15 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.25 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.35 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.45 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.55 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.65 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.75 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.85 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.95 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
321 1 1.50 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
321 1 2.50 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
321 1 3.50 3.0641 0.00 0.000000 0.000001 0.999999 0.000000
321 1 4.50 3.0641 0.00 0.000000 0.000000 1.000000 0.000000
321 1 5.50 3.0641 0.00 0.000000 0.000026 0.973126 0.026848
321 1 6.50 3.0641 0.00 0.000000 0.001520 0.995158 0.003322
321 1 7.50 3.0641 0.00 0.000000 0.015656 0.983754 0.000589
321 1 8.50 3.0641 0.00 0.000000 0.044167 0.955425 0.000409
321 1 9.50 3.0641 0.00 0.000000 0.089320 0.910046 0.000633
321 1 10.50 3.0641 0.00 0.000000 0.135785 0.861748 0.002467
321 1 11.50 3.0641 0.00 0.000000 0.179961 0.813668 0.006371
321 1 12.50 3.0641 0.00 0.000000 0.223621 0.760773 0.015606
321 1 13.50 3.0641 0.00 0.000000 0.253578 0.715215 0.031207
321 1 14.50 3.0641 0.00 0.000000 0.285174 0.659172 0.055654
321 1 0.08 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.15 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.25 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.35 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.45 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.55 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.65 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.75 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.85 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.95 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
321 1 1.50 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
321 1 2.50 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
321 1 3.50 3.0317 0.00 0.000000 0.000002 0.999998 0.000000
321 1 4.50 3.0317 0.00 0.000000 0.000001 0.999999 0.000000
321 1 5.50 3.0317 0.00 0.000000 0.000054 0.947748 0.052198
321 1 6.50 3.0317 0.00 0.000000 0.002753 0.991082 0.006165
321 1 7.50 3.0317 0.00 0.000000 0.019550 0.979593 0.000857
321 1 8.50 3.0317 0.00 0.000000 0.054251 0.945487 0.000262
321 1 9.50 3.0317 0.00 0.000000 0.089669 0.909613 0.000718
321 1 10.50 3.0317 0.00 0.000000 0.128506 0.867032 0.004462
321 1 11.50 3.0317 0.00 0.000000 0.177425 0.809841 0.012734
321 1 12.50 3.0317 0.00 0.000000 0.220943 0.760278 0.018779
321 1 13.50 3.0317 0.00 0.000000 0.260014 0.703137 0.036849
321 1 14.50 3.0317 0.00 0.000000 0.286295 0.654813 0.058893
321 1 0.08 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.15 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.25 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.35 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.45 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.55 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.65 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.75 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.85 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.95 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
321 1 1.50 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
321 1 2.50 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
321 1 3.50 2.9607 0.00 0.000000 0.000006 0.999994 0.000000
321 1 4.50 2.9607 0.00 0.000000 0.000000 1.000000 0.000000
321 1 5.50 2.9607 0.00 0.000000 0.000037 0.938005 0.061958
321 1 6.50 2.9607 0.00 0.000000 0.001532 0.994917 0.003552
321 1 7.50 2.9607 0.00 0.000000 0.013945 0.985360 0.000695
321 1 8.50 2.9607 0.00 0.000000 0.038657 0.961114 0.000229
321 1 9.50 2.9607 0.00 0.000000 0.089322 0.909667 0.001012
321 1 10.50 2.9607 0.00 0.000000 0.137247 0.860977 0.001776
321 1 11.50 2.9607 0.00 0.000000 0.161448 0.832257 0.006295
321 1 12.50 2.9607 0.00 0.000000 0.226336 0.758531 0.015133
321 1 13.50 2.9607 0.00 0.000000 0.244823 0.719954 0.035223
321 1 14.50 2.9607 0.00 0.000000 0.284113 0.660221 0.055666
321 1 0.08 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.15 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.25 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.35 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.45 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.55 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.65 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.75 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.85 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
321 1 0.95 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
321 1 1.50 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
321 1 2.50 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
321 1 3.50 2.7975 0.00 0.000000 0.000005 0.999995 0.000000
321 1 4.50 2.7975 0.00 0.000000 0.000001 0.999999 0.000000
321 1 5.50 2.7975 0.00 0.000000 0.000142 0.957201 0.042658
321 1 6.50 2.7975 0.00 0.000000 0.003496 0.988391 0.008114
321 1 7.50 2.7975 0.00 0.000000 0.023802 0.974646 0.001552
321 1 8.50 2.7975 0.00 0.000000 0.055747 0.943688 0.000566
321 1 9.50 2.7975 0.00 0.000000 0.101626 0.896064 0.002311
321 1 10.50 2.7975 0.00 0.000000 0.157611 0.837160 0.005230
321 1 11.50 2.7975 0.00 0.000000 0.193004 0.796842 0.010154
321 1 12.50 2.7975 0.00 0.000000 0.239251 0.741388 0.019361
321 1 13.50 2.7975 0.00 0.000000 0.270648 0.684750 0.044602
321 1 14.50 2.7975 0.00 0.000000 0.301466 0.625336 0.073198
2212 1 0.08 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.15 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.25 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.35 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.45 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.55 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.65 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.75 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.85 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.95 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 1.50 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 2.50 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 3.50 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 4.50 3.0641 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 5.50 3.0641 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 6.50 3.0641 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 7.50 3.0641 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 8.50 3.0641 0.00 0.000000 0.000000 0.000002 0.999998
2212 1 9.50 3.0641 0.00 0.000000 0.000000 0.000071 0.999929
2212 1 10.50 3.0641 0.00 0.000000 0.000000 0.000805 0.999195
2212 1 11.50 3.0641 0.00 0.000000 0.000000 0.003907 0.996093
2212 1 12.50 3.0641 0.00 0.000000 0.000000 0.016109 0.983891
2212 1 13.50 3.0641 0.00 0.000000 0.000000 0.029652 0.970348
2212 1 14.50 3.0641 0.00 0.000000 0.000000 0.055654 0.944346
2212 1 0.08 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.15 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.25 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.35 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.45 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.55 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.65 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.75 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.85 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.95 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 1.50 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 2.50 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 3.50 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 4.50 3.0317 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 5.50 3.0317 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 6.50 3.0317 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 7.50 3.0317 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 8.50 3.0317 0.00 0.000000 0.000000 0.000002 0.999998
2212 1 9.50 3.0317 0.00 0.000000 0.000000 0.000123 0.999877
2212 1 10.50 3.0317 0.00 0.000000 0.000000 0.000888 0.999112
2212 1 11.50 3.0317 0.00 0.000000 0.000000 0.006060 0.993940
2212 1 12.50 3.0317 0.00 0.000000 0.000000 0.016144 0.983856
2212 1 13.50 3.0317 0.00 0.000000 0.000000 0.036023 0.963977
2212 1 14.50 3.0317 0.00 0.000000 0.000000 0.054647 0.945353
2212 1 0.08 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.15 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.25 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.35 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.45 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.55 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.65 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.75 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.85 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.95 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 1.50 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 2.50 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 3.50 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 4.50 2.9607 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 5.50 2.9607 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 6.50 2.9607 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 7.50 2.9607 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 8.50 2.9607 0.00 0.000000 0.000000 0.000004 0.999996
2212 1 9.50 2.9607 0.00 0.000000 0.000000 0.000049 0.999951
2212 1 10.50 2.9607 0.00 0.000000 0.000000 0.000777 0.999223
2212 1 11.50 2.9607 0.00 0.000000 0.000000 0.005093 0.994907
2212 1 12.50 2.9607 0.00 0.000000 0.000000 0.013229 0.986771
2212 1 13.50 2.9607 0.00 0.000000 0.000000 0.032800 0.967200
2212 1 14.50 2.9607 0.00 0.000000 0.000000 0.057789 0.942211
2212 1 0.08 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.15 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.25 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.35 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.45 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.55 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.65 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.75 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.85 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 0.95 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 1.50 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 2.50 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 3.50 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 4.50 2.7975 0.00 0.000000 0.000000 0.000000 0.000000
2212 1 5.50 2.7975 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 6.50 2.7975 0.00 0.000000 0.000000 0.000000 1.000000
2212 1 7.50 2.7975 0.00 0.000000 0.000000 0.000001 0.999999
2212 1 8.50 2.7975 0.00 0.000000 0.000000 0.000012 0.999988
2212 1 9.50 2.7975 0.00 0.000000 0.000000 0.000295 0.999705
2212 1 10.50 2.7975 0.00 0.000000 0.000000 0.003253 0.996747
2212 1 11.50 2.7975 0.00 0.000000 0.000000 0.012285 0.987715
2212 1 12.50 2.7975 0.00 0.000000 0.000000 0.023402 0.976598
2212 1 13.50 2.7975 0.00 0.000000 0.000000 0.039440 0.960560
2212 1 14.50 2.7975 0.00 0.000000 0.000000 0.062346 0.937654
## pfRICH PID LUT - version 0.1
## Eta Bins: [-3.8, -3.5], [-3.5, -3.0], [-3.0, -2.8], [-2.8, -2.0], [-2.0,
-1.5]
## Momentum Bins [GeV]: [0.05, 0.1] : then 0.1 to 1.0 in 0.1 wide steps :
then 1.0 to 15.0 in 1.0 wide steps
## NOTE: Particles thrown flat within Eta and Momentum Bins - results
reported at bin center
## NOTE: No charge scan - only thrown electrons, pi+, K+, and proton - assume
charge symmetry at this point
## NOTE: Particles thrown flat in Phi in this run - Phi value is a dummy
## NOTE: Rows with all zero probability indicate PID not possible
## Format: PID Charge Momentum Theta Phi prob-electron prob-pion prob-kaon
prob-proton
211 1 0.75 3.0641 0.00 1.000000 0.000000 0.000000
211 1 0.85 3.0641 0.00 1.000000 0.000000 0.000000
211 1 0.95 3.0641 0.00 1.000000 0.000000 0.000000
211 1 1.50 3.0641 0.00 1.000000 0.000000 0.000000
211 1 2.50 3.0641 0.00 1.000000 0.000000 0.000000
211 1 3.50 3.0641 0.00 0.892193 0.107807 0.000000
211 1 4.50 3.0641 0.00 0.991581 0.008419 0.000000
211 1 5.50 3.0641 0.00 0.997921 0.002079 0.000000
211 1 6.50 3.0641 0.00 0.994956 0.005044 0.000000
211 1 7.50 3.0641 0.00 0.979378 0.020622 0.000000
211 1 8.50 3.0641 0.00 0.951838 0.048162 0.000000
211 1 9.50 3.0641 0.00 0.906690 0.093310 0.000000
211 1 10.50 3.0641 0.00 0.862748 0.137252 0.000000
211 1 11.50 3.0641 0.00 0.821518 0.178482 0.000000
211 1 12.50 3.0641 0.00 0.770998 0.229002 0.000000
211 1 13.50 3.0641 0.00 0.743841 0.256159 0.000000
211 1 14.50 3.0641 0.00 0.710247 0.289753 0.000000
211 1 0.75 3.0317 0.00 1.000000 0.000000 0.000000
211 1 0.85 3.0317 0.00 1.000000 0.000000 0.000000
211 1 0.95 3.0317 0.00 1.000000 0.000000 0.000000
211 1 1.50 3.0317 0.00 1.000000 0.000000 0.000000
211 1 2.50 3.0317 0.00 1.000000 0.000000 0.000000
211 1 3.50 3.0317 0.00 0.865444 0.134556 0.000000
211 1 4.50 3.0317 0.00 0.991445 0.008555 0.000000
211 1 5.50 3.0317 0.00 0.995315 0.004685 0.000000
211 1 6.50 3.0317 0.00 0.992724 0.007276 0.000000
211 1 7.50 3.0317 0.00 0.973063 0.026937 0.000000
211 1 8.50 3.0317 0.00 0.947807 0.052193 0.000000
211 1 9.50 3.0317 0.00 0.897096 0.102904 0.000000
211 1 10.50 3.0317 0.00 0.862586 0.137414 0.000000
211 1 11.50 3.0317 0.00 0.808155 0.191845 0.000000
211 1 12.50 3.0317 0.00 0.770917 0.229083 0.000000
211 1 13.50 3.0317 0.00 0.733919 0.266081 0.000000
211 1 14.50 3.0317 0.00 0.711416 0.288584 0.000000
211 1 0.75 2.9607 0.00 1.000000 0.000000 0.000000
211 1 0.85 2.9607 0.00 1.000000 0.000000 0.000000
211 1 0.95 2.9607 0.00 1.000000 0.000000 0.000000
211 1 1.50 2.9607 0.00 1.000000 0.000000 0.000000
211 1 2.50 2.9607 0.00 1.000000 0.000000 0.000000
211 1 3.50 2.9607 0.00 0.824269 0.175731 0.000000
211 1 4.50 2.9607 0.00 0.990007 0.009993 0.000000
211 1 5.50 2.9607 0.00 0.997732 0.002268 0.000000
211 1 6.50 2.9607 0.00 0.993051 0.006949 0.000000
211 1 7.50 2.9607 0.00 0.976574 0.023426 0.000000
211 1 8.50 2.9607 0.00 0.938908 0.061092 0.000000
211 1 9.50 2.9607 0.00 0.914682 0.085318 0.000000
211 1 10.50 2.9607 0.00 0.862753 0.137247 0.000000
211 1 11.50 2.9607 0.00 0.811276 0.188724 0.000000
211 1 12.50 2.9607 0.00 0.779138 0.220862 0.000000
211 1 13.50 2.9607 0.00 0.738174 0.261826 0.000000
211 1 14.50 2.9607 0.00 0.707915 0.292085 0.000000
211 1 0.75 2.7975 0.00 1.000000 0.000000 0.000000
211 1 0.85 2.7975 0.00 1.000000 0.000000 0.000000
211 1 0.95 2.7975 0.00 1.000000 0.000000 0.000000
211 1 1.50 2.7975 0.00 1.000000 0.000000 0.000000
211 1 2.50 2.7975 0.00 1.000000 0.000000 0.000000
211 1 3.50 2.7975 0.00 0.857606 0.142394 0.000000
211 1 4.50 2.7975 0.00 0.990440 0.009560 0.000000
211 1 5.50 2.7975 0.00 0.994970 0.005030 0.000000
211 1 6.50 2.7975 0.00 0.991399 0.008601 0.000000
211 1 7.50 2.7975 0.00 0.966854 0.033146 0.000000
211 1 8.50 2.7975 0.00 0.930521 0.069479 0.000000
211 1 9.50 2.7975 0.00 0.884125 0.115875 0.000000
211 1 10.50 2.7975 0.00 0.825356 0.174644 0.000000
211 1 11.50 2.7975 0.00 0.779961 0.220039 0.000000
211 1 12.50 2.7975 0.00 0.754698 0.245302 0.000000
211 1 13.50 2.7975 0.00 0.729352 0.270648 0.000000
211 1 14.50 2.7975 0.00 0.703692 0.296308 0.000000
321 1 0.75 3.0641 0.00 0.000000 0.000000 0.000000
321 1 0.85 3.0641 0.00 0.000000 0.000000 0.000000
321 1 0.95 3.0641 0.00 0.000000 0.000000 0.000000
321 1 1.50 3.0641 0.00 0.000000 0.000000 0.000000
321 1 2.50 3.0641 0.00 0.000000 0.000000 0.000000
321 1 3.50 3.0641 0.00 0.000001 0.999999 0.000000
321 1 4.50 3.0641 0.00 0.000000 1.000000 0.000000
321 1 5.50 3.0641 0.00 0.000026 0.973126 0.026848
321 1 6.50 3.0641 0.00 0.001520 0.995158 0.003322
321 1 7.50 3.0641 0.00 0.015656 0.983754 0.000589
321 1 8.50 3.0641 0.00 0.044167 0.955425 0.000409
321 1 9.50 3.0641 0.00 0.089320 0.910046 0.000633
321 1 10.50 3.0641 0.00 0.135785 0.861748 0.002467
321 1 11.50 3.0641 0.00 0.179961 0.813668 0.006371
321 1 12.50 3.0641 0.00 0.223621 0.760773 0.015606
321 1 13.50 3.0641 0.00 0.253578 0.715215 0.031207
321 1 14.50 3.0641 0.00 0.285174 0.659172 0.055654
321 1 0.75 3.0317 0.00 0.000000 0.000000 0.000000
321 1 0.85 3.0317 0.00 0.000000 0.000000 0.000000
321 1 0.95 3.0317 0.00 0.000000 0.000000 0.000000
321 1 1.50 3.0317 0.00 0.000000 0.000000 0.000000
321 1 2.50 3.0317 0.00 0.000000 0.000000 0.000000
321 1 3.50 3.0317 0.00 0.000002 0.999998 0.000000
321 1 4.50 3.0317 0.00 0.000001 0.999999 0.000000
321 1 5.50 3.0317 0.00 0.000054 0.947748 0.052198
321 1 6.50 3.0317 0.00 0.002753 0.991082 0.006165
321 1 7.50 3.0317 0.00 0.019550 0.979593 0.000857
321 1 8.50 3.0317 0.00 0.054251 0.945487 0.000262
321 1 9.50 3.0317 0.00 0.089669 0.909613 0.000718
321 1 10.50 3.0317 0.00 0.128506 0.867032 0.004462
321 1 11.50 3.0317 0.00 0.177425 0.809841 0.012734
321 1 12.50 3.0317 0.00 0.220943 0.760278 0.018779
321 1 13.50 3.0317 0.00 0.260014 0.703137 0.036849
321 1 14.50 3.0317 0.00 0.286295 0.654813 0.058893
321 1 0.75 2.9607 0.00 0.000000 0.000000 0.000000
321 1 0.85 2.9607 0.00 0.000000 0.000000 0.000000
321 1 0.95 2.9607 0.00 0.000000 0.000000 0.000000
321 1 1.50 2.9607 0.00 0.000000 0.000000 0.000000
321 1 2.50 2.9607 0.00 0.000000 0.000000 0.000000
321 1 3.50 2.9607 0.00 0.000006 0.999994 0.000000
321 1 4.50 2.9607 0.00 0.000000 1.000000 0.000000
321 1 5.50 2.9607 0.00 0.000037 0.938005 0.061958
321 1 6.50 2.9607 0.00 0.001532 0.994917 0.003552
321 1 7.50 2.9607 0.00 0.013945 0.985360 0.000695
321 1 8.50 2.9607 0.00 0.038657 0.961114 0.000229
321 1 9.50 2.9607 0.00 0.089322 0.909667 0.001012
321 1 10.50 2.9607 0.00 0.137247 0.860977 0.001776
321 1 11.50 2.9607 0.00 0.161448 0.832257 0.006295
321 1 12.50 2.9607 0.00 0.226336 0.758531 0.015133
321 1 13.50 2.9607 0.00 0.244823 0.719954 0.035223
321 1 14.50 2.9607 0.00 0.284113 0.660221 0.055666
321 1 0.75 2.7975 0.00 0.000000 0.000000 0.000000
321 1 0.85 2.7975 0.00 0.000000 0.000000 0.000000
321 1 0.95 2.7975 0.00 0.000000 0.000000 0.000000
321 1 1.50 2.7975 0.00 0.000000 0.000000 0.000000
321 1 2.50 2.7975 0.00 0.000000 0.000000 0.000000
321 1 3.50 2.7975 0.00 0.000005 0.999995 0.000000
321 1 4.50 2.7975 0.00 0.000001 0.999999 0.000000
321 1 5.50 2.7975 0.00 0.000142 0.957201 0.042658
321 1 6.50 2.7975 0.00 0.003496 0.988391 0.008114
321 1 7.50 2.7975 0.00 0.023802 0.974646 0.001552
321 1 8.50 2.7975 0.00 0.055747 0.943688 0.000566
321 1 9.50 2.7975 0.00 0.101626 0.896064 0.002311
321 1 10.50 2.7975 0.00 0.157611 0.837160 0.005230
321 1 11.50 2.7975 0.00 0.193004 0.796842 0.010154
321 1 12.50 2.7975 0.00 0.239251 0.741388 0.019361
321 1 13.50 2.7975 0.00 0.270648 0.684750 0.044602
321 1 14.50 2.7975 0.00 0.301466 0.625336 0.073198
2212 1 0.75 3.0641 0.00 0.000000 0.000000 0.000000
2212 1 0.85 3.0641 0.00 0.000000 0.000000 0.000000
2212 1 0.95 3.0641 0.00 0.000000 0.000000 0.000000
2212 1 1.50 3.0641 0.00 0.000000 0.000000 0.000000
2212 1 2.50 3.0641 0.00 0.000000 0.000000 0.000000
2212 1 3.50 3.0641 0.00 0.000000 0.000000 0.000000
2212 1 4.50 3.0641 0.00 0.000000 0.000000 0.000000
2212 1 5.50 3.0641 0.00 0.000000 0.000000 1.000000
2212 1 6.50 3.0641 0.00 0.000000 0.000000 1.000000
2212 1 7.50 3.0641 0.00 0.000000 0.000000 1.000000
2212 1 8.50 3.0641 0.00 0.000000 0.000002 0.999998
2212 1 9.50 3.0641 0.00 0.000000 0.000071 0.999929
2212 1 10.50 3.0641 0.00 0.000000 0.000805 0.999195
2212 1 11.50 3.0641 0.00 0.000000 0.003907 0.996093
2212 1 12.50 3.0641 0.00 0.000000 0.016109 0.983891
2212 1 13.50 3.0641 0.00 0.000000 0.029652 0.970348
2212 1 14.50 3.0641 0.00 0.000000 0.055654 0.944346
2212 1 0.75 3.0317 0.00 0.000000 0.000000 0.000000
2212 1 0.85 3.0317 0.00 0.000000 0.000000 0.000000
2212 1 0.95 3.0317 0.00 0.000000 0.000000 0.000000
2212 1 1.50 3.0317 0.00 0.000000 0.000000 0.000000
2212 1 2.50 3.0317 0.00 0.000000 0.000000 0.000000
2212 1 3.50 3.0317 0.00 0.000000 0.000000 0.000000
2212 1 4.50 3.0317 0.00 0.000000 0.000000 0.000000
2212 1 5.50 3.0317 0.00 0.000000 0.000000 1.000000
2212 1 6.50 3.0317 0.00 0.000000 0.000000 1.000000
2212 1 7.50 3.0317 0.00 0.000000 0.000000 1.000000
2212 1 8.50 3.0317 0.00 0.000000 0.000002 0.999998
2212 1 9.50 3.0317 0.00 0.000000 0.000123 0.999877
2212 1 10.50 3.0317 0.00 0.000000 0.000888 0.999112
2212 1 11.50 3.0317 0.00 0.000000 0.006060 0.993940
2212 1 12.50 3.0317 0.00 0.000000 0.016144 0.983856
2212 1 13.50 3.0317 0.00 0.000000 0.036023 0.963977
2212 1 14.50 3.0317 0.00 0.000000 0.054647 0.945353
2212 1 0.75 2.9607 0.00 0.000000 0.000000 0.000000
2212 1 0.85 2.9607 0.00 0.000000 0.000000 0.000000
2212 1 0.95 2.9607 0.00 0.000000 0.000000 0.000000
2212 1 1.50 2.9607 0.00 0.000000 0.000000 0.000000
2212 1 2.50 2.9607 0.00 0.000000 0.000000 0.000000
2212 1 3.50 2.9607 0.00 0.000000 0.000000 0.000000
2212 1 4.50 2.9607 0.00 0.000000 0.000000 0.000000
2212 1 5.50 2.9607 0.00 0.000000 0.000000 1.000000
2212 1 6.50 2.9607 0.00 0.000000 0.000000 1.000000
2212 1 7.50 2.9607 0.00 0.000000 0.000000 1.000000
2212 1 8.50 2.9607 0.00 0.000000 0.000004 0.999996
2212 1 9.50 2.9607 0.00 0.000000 0.000049 0.999951
2212 1 10.50 2.9607 0.00 0.000000 0.000777 0.999223
2212 1 11.50 2.9607 0.00 0.000000 0.005093 0.994907
2212 1 12.50 2.9607 0.00 0.000000 0.013229 0.986771
2212 1 13.50 2.9607 0.00 0.000000 0.032800 0.967200
2212 1 14.50 2.9607 0.00 0.000000 0.057789 0.942211
2212 1 0.75 2.7975 0.00 0.000000 0.000000 0.000000
2212 1 0.85 2.7975 0.00 0.000000 0.000000 0.000000
2212 1 0.95 2.7975 0.00 0.000000 0.000000 0.000000
2212 1 1.50 2.7975 0.00 0.000000 0.000000 0.000000
2212 1 2.50 2.7975 0.00 0.000000 0.000000 0.000000
2212 1 3.50 2.7975 0.00 0.000000 0.000000 0.000000
2212 1 4.50 2.7975 0.00 0.000000 0.000000 0.000000
2212 1 5.50 2.7975 0.00 0.000000 0.000000 1.000000
2212 1 6.50 2.7975 0.00 0.000000 0.000000 1.000000
2212 1 7.50 2.7975 0.00 0.000000 0.000001 0.999999
2212 1 8.50 2.7975 0.00 0.000000 0.000012 0.999988
2212 1 9.50 2.7975 0.00 0.000000 0.000295 0.999705
2212 1 10.50 2.7975 0.00 0.000000 0.003253 0.996747
2212 1 11.50 2.7975 0.00 0.000000 0.012285 0.987715
2212 1 12.50 2.7975 0.00 0.000000 0.023402 0.976598
2212 1 13.50 2.7975 0.00 0.000000 0.039440 0.960560
2212 1 14.50 2.7975 0.00 0.000000 0.062346 0.937654