Craig and Justin,

On 1/17/20 1:22 PM, woody wrote:
Dear All,
  We had a lively discussion at today's EMCAL meeting about how the radiation damage in the SiPMs will affect the noise and our ability to see MIPs in the EMCAL after we start running. While our paper on radiation damage in the SiPMs doesn't answer this question directly, it shows the effect we expect to see in the SiPMs currents as a function of radiation dose, which I think we have now taken as a given. There are also lots of other slides and talks about this subject, but we have not really faced how this will affect the EMCAL calibration and performance, especially after some period of running. I think some simulation of this would be extremely helpful and I would therefore encourage anyone who is interested to join in the discussion and help us understand how we will eventually deal with this problem.

For the purposes of the simulation, I think the best information we have about how the detector will behave as it becomes radiation damaged comes from the SiPM boards that we sent to Lowell to be radiation damaged in their reactor:

https://www.uml.edu/research/radlab/neutron-facilities.aspx

I like that exposure best because a) it's our exact devices, b) the doses were chosen to reflect our best estimate of the dose we expect in sPHENIX based on Jin's radiation maps, and c) Lowell is in the business of radiation exposure, so I think we can trust their dosimetry. The data we have is limited but consistent; Eric, Sean, and Craig selected 3 SiPM boards (== 4 towers == 16 SiPM's) and dosed each to about 1E10, 1E11, and 1E12 1 MeV equivalent n/cm^2. Some of what we know from that is attached, which is the pedestal widths at the 3 doses taken through the EMCAL electronics. I think that data can be codified into the simulation, at least for a first pass.

There are some complications we have to account for to put this carefully into our simulations. [The first is that I should make sure that what I'm showing you is in a normal gain channel and not the switchable high gain channel, but that should be just checking the bookkeeping.] You need the energy scale in the channel to use it in the Monte Carlo, and there is an additional effect due to the fact that the SiPM's are drawing current which effectively reduces the bias applied unless corrected for; I tried to do that and check it with the LED's, but I had problems with the procedure, which is also a long story, but boils down to the difficulty of making the LED's reproducible over time and temperature and removal and insertion of the daughterboards. The reduced gain due to current draw is not a negligible effect, and it's part of what we have to keep under control to keep the EMCAL calibrated in sPHENIX.

I think we can come up with some estimates for everything I've described, and that can be the first pass at adding it to the simulation. I think we should actually do a second round of this using some sector 0 SiPM's where we dose more SiPM boards and do the calibration and correction more carefully based on what we learned from the first round. There is yet one more complication I should mention, which are the dose estimates--Jin did a typically great job in his estimates, but estimates are, well, estimates and maybe we should make some measurements at higher doses to be prepared for the worst.

Finally, if anyone read this far, I think we do not want to divorce the calibration from the detector group, which is why I encouraged Justin to address the EWMCAL meeting. We have spent a loooong time developing the calorimeter concepts into a working detector, and the detector groups are the repository of knowledge about how the detectors are built, how they work, and what we know about their characteristics.

--
John Haggerty