Justin,

I just looked at some data that is not only clearly labeled in my notebook, but also has the test pulse. The data I attached was certainly normal (a.k.a. low) gain. The high gain pedestal widths (x16) are pretty easily distinguished from these. I should mention for full disclosure that the particular run I looked at was a little different (the largest width was maybe 50 instead of 65), but I'd have to dive into the temperature data at that level.

On 1/31/20 2:51 PM, Frantz, Justin wrote:
Hi John, Craig, or anyone....did we ever determine whether John's plot that
was attached to this email was for normal gain or high gain?


----------------------------------------------
Justin Frantz, Ph. D.
RHIC/AGS User Executive Committee
Brookhaven National Lab
Associate Professor
Ohio University Dept. Of Physics and Astronomy
frantz AT ohio.edu
646-228-2539


-----Original Message-----
From: sPHENIX-EMCal-l [mailto:sphenix-emcal-l-bounces AT lists.bnl.gov] On
Behalf Of John Haggerty
Sent: Saturday, January 18, 2020 11:28 AM
To: sphenix-emcal-l AT lists.bnl.gov
Subject: Re: [Sphenix-emcal-l] Radiation damage in the EMCAL SiPMs

Craig and Justin,

On 1/17/20 1:22 PM, woody wrote:
Dear All,
  We had a lively discussion at today's EMCAL meeting about how the
radiation damage in the SiPMs will affect the noise and our ability to
see MIPs in the EMCAL after we start running. While our paper on
radiation damage in the SiPMs doesn't answer this question directly, it
shows the effect we expect to see in the SiPMs currents as a function of
radiation dose, which I think we have now taken as a given. There are
also lots of other slides and talks about this subject, but we have not
really faced how this will affect the EMCAL calibration and performance,
especially after some period of running. I think some simulation of this
would be extremely helpful and I would therefore encourage anyone who
is
interested to join in the discussion and help us understand how we will
eventually deal with this problem.

For the purposes of the simulation, I think the best information we have
about how the detector will behave as it becomes radiation damaged comes
from the SiPM boards that we sent to Lowell to be radiation damaged in
their reactor:

https://urldefense.com/v3/__https://www.uml.edu/research/radlab/neutron-facilities.aspx__;!!P4SdNyxKAPE!U3rUxW1nDMLfF1NGByTkYWQg7f64iwwNBNa219YmKl7QokEzlXBGLPAcF_0p3N3o$

I like that exposure best because a) it's our exact devices, b) the
doses were chosen to reflect our best estimate of the dose we expect in
sPHENIX based on Jin's radiation maps, and c) Lowell is in the business
of radiation exposure, so I think we can trust their dosimetry. The
data we have is limited but consistent; Eric, Sean, and Craig selected 3
SiPM boards (== 4 towers == 16 SiPM's) and dosed each to about 1E10,
1E11, and 1E12 1 MeV equivalent n/cm^2. Some of what we know from that
is attached, which is the pedestal widths at the 3 doses taken through
the EMCAL electronics. I think that data can be codified into the
simulation, at least for a first pass.

There are some complications we have to account for to put this
carefully into our simulations. [The first is that I should make sure
that what I'm showing you is in a normal gain channel and not the
switchable high gain channel, but that should be just checking the
bookkeeping.] You need the energy scale in the channel to use it in the
Monte Carlo, and there is an additional effect due to the fact that the
SiPM's are drawing current which effectively reduces the bias applied
unless corrected for; I tried to do that and check it with the LED's,
but I had problems with the procedure, which is also a long story, but
boils down to the difficulty of making the LED's reproducible over time
and temperature and removal and insertion of the daughterboards. The
reduced gain due to current draw is not a negligible effect, and it's
part of what we have to keep under control to keep the EMCAL calibrated
in sPHENIX.

I think we can come up with some estimates for everything I've
described, and that can be the first pass at adding it to the
simulation. I think we should actually do a second round of this using
some sector 0 SiPM's where we dose more SiPM boards and do the
calibration and correction more carefully based on what we learned from
the first round. There is yet one more complication I should mention,
which are the dose estimates--Jin did a typically great job in his
estimates, but estimates are, well, estimates and maybe we should make
some measurements at higher doses to be prepared for the worst.

Finally, if anyone read this far, I think we do not want to divorce the
calibration from the detector group, which is why I encouraged Justin to
address the EWMCAL meeting. We have spent a loooong time developing
the
calorimeter concepts into a working detector, and the detector groups
are the repository of knowledge about how the detectors are built, how
they work, and what we know about their characteristics.

--
John Haggerty