let me add to Dave’s list:

stick with mm or cm but not both, e.g. 82-84 and 119-120

9 do you need both “system”

39 is it clear what a 3x3 cluster is

122 do we need the 1x8 ?

146 is it clear what a "asymmetric shift” is ?

caption Fig. 5 four blocks or four modules ?

180 ‘ganged” or better "summed"

caption Fig 7 “scintillating”

Fig 8 not sure those black numbers are going to be visible or informative

and “scintillation tile” ?

292 “scintillating"

more later ….

Achim

> On Jan 18, 2017, at 14:33 , David Morrison <morrison AT bnl.gov> wrote:
>
> Dear collaborators,
>
> Let me also encourage anyone and everyone to read the draft and comment
> on it. It's printed with line numbers in the margins – thanks! - which
> makes typing up comments pretty straightforward.
>
> To get the ball rolling, below are my comments on Sections I-III.
>
> Cheers,
> Dave
>
> abstract: s/preformance/performance
>
> 11: |eta| < 1 or 1.1 (you say 1.1 on line 90)
>
> 15: "... which sit inside the solenoid, and ..." (we have a specific magnet)
>
> 26: s/principle/principal/
>
> 40: s/from a typical/from an/ (you use typical again later in the sentence)
>
> 43: s/be no greater/be no worse/ (always confusing with resolution –
> greater can mean larger or it can mean better – choose "better" or "worse")
>
> 45: s/and another/and the other/ (otherwise sounds like there are lots
> of radial segments to choose from)
>
> 52: "... sqrt(E)$, which corresponds to an energy resolution for single
> hadrons in the full calorimeter system of ..." (was an awful lot of
> "requiring" before)
>
> 63: s/3/three/ (small counting numbers should be spelled out)
>
> 64-66: awkward - how about, "... to provide as many interaction lengths
> of material as a particle would encounter traversing the sPHENIX solenoid."
>
> 70: s/basic// (not needed)
>
> 71: s/is/consists of/
>
> 75-75: shouldn't the references be consolidated? I.e., "[7-11]".
>
> 90: cf line 11
>
> 99: "... were produced at two sites, ..."
>
> 101: s/Chaimpaign/Champaign/
>
> 104-106: how about, "The screens are then separated longitudinally,
> placed in the mold, and tilted to form the taper in one dimension."
>
> 107: relocate the "then": "Tungsten powder is poured uniformly into the
> mold and then epoxy is poured into the tungsten-fiber matrix."
>
> 107: s/In order to/To/
>
> 111-113: how about, "We trim the blocks first with carbide tipped
> cutters and then with diamond tipped ones."
>
> Table I is too wide.
>
> 124: s/Silicones/Silicone/
>
> 126-127: how about, "... coupled to a PMT with a 2in window."
>
> 139: s/map out/map/
>
> 141: s/readout/read out/ also s/the array/an array/
>
> 147-151: how about, "A ~30% relative variation in light collection is
> observed across the light guide, leading to a ~20% ..."
>
> 149: btw, what does "indenting position dependent" mean? Is that a typo?
>
> 154: s/finally// (yes, it seemed like a long time, but don't write it
> that way)
>
> 156: s/8/eight/
>
> 160: Why "(PCB)" as an initialism for pre-amplifier board? Yes, they
> are printed circuit boards, but I'd leave that off. I.e., "The
> pre-amplifier board, which carried four ..."
>
> 163: s/PCB/board/
>
> 168: s/amplifyer/amplifier/
>
> Fig. 6 caption: s/4/four/
>
> 192: s/prevent excess/limit/
>
> 201: weird spacing between the "x" and the second "3".
>
> 204: how about, "The air gap between the fiber ends and the face of the
> SiPM allows the emitted light to spread over the face of the SiPM,
> reducing the probability of optical ..." and then "A gap of 0.75mm
> balances the requirements that there be no more than a 5% variation ..."
>
> 216cf: how about, "A dry mix ... is melted and extruded, producing a
> continuous band of scintillating plastic 25cm wide."
>
> 224: s/a mix of//
>
> 227: s/removes/also removes/
>
> 251: "During the FNAL test beam, ..."
>
> 253: s/4/four/
>
> 257, 258: s/10/ten/ (this could go either way, I admit)
>
> 262, 263: no period after "nm"
>
> 275: s/5/five/
>
> 281: s/5/five/
>
> 287: s/then/than/
>
> Table III is too wide.
>
> 309: s/4/four/
>
> 324: s/board/boards/ (singular/plural agreement)
>
> 330: pick one: "comprises" or "consists of", but not "comprised of"
>
> 333: s/4/four/
>
>
> On 1/18/17 12:38 PM, Ron Belmont wrote:
>> Dear All,
>>
>> We'd like to remind you that the first comment period will close one
>> week from today.
>>
>> With best regards,
>>
>> Abhisek, Jin, Megan, Vera, and Ron
>>
>>
>>
>>
>> On Wed, Jan 11, 2017 at 12:05 PM, Ron Belmont <ron.belmont AT colorado.edu
>> <mailto:ron.belmont AT colorado.edu>> wrote:
>>
>> Dear sPHENIX collaborators,
>>
>> We are pleased to release to the collaboration the first draft of
>> the T-1044 2016 calorimeter beam test paper, which you can find
>> here:
>> https://wiki.bnl.gov/sPHENIX/images/7/75/Beam-test-results_first-collab-release.pdf
>>
>> <https://wiki.bnl.gov/sPHENIX/images/7/75/Beam-test-results_first-collab-release.pdf>
>> .
>> We greatly look forward to and appreciate your comments within two
>> week's time, by the end of the day on Wednesday, January 25. For
>> those who have made comments on the previous version released to the
>> calorimeter lists, you can find responses
>> here: https://wiki.bnl.gov/sPHENIX/index.php/T-1044_publication_comments
>> <https://wiki.bnl.gov/sPHENIX/index.php/T-1044_publication_comments>
>> The author list and acknowledgements are not yet finalized. If you
>> think you should be an author, please fill out the poll sent out by
>> Dave and
>> Gunther:
>> https://docs.google.com/forms/d/e/1FAIpQLSdyrXUoNoiS65D6VqVPNPyGZJQYk_5IiIYqcUW9G4ak5c19gw/viewform?c=0&w=1
>>
>> <https://docs.google.com/forms/d/e/1FAIpQLSdyrXUoNoiS65D6VqVPNPyGZJQYk_5IiIYqcUW9G4ak5c19gw/viewform?c=0&w=1>
>> .
>> If you have any additional requests for the acknowledgements, please
>> email one of us privately.
>>
>> With best regards,
>>
>> The T-1044 paper committee
>> (Abhisek, Jin, Megan, Vera, and Ron)
>
> --
> David Morrison Brookhaven National Laboratory phone: 631-344-5840
> Physics Department, Bldg 510 C fax: 631-344-3253
> Upton, NY 11973-5000 email: dave AT bnl.gov
>
> _______________________________________________