sphenix-magnet-l AT lists.bnl.gov
Subject: sPHENIX discussion of the superconducting solenoid
List archive
Re: [Sphenix-magnet-l] Can we do hipot at 90-100K ? ---- How about Jan. 3 ?
- From: "Muratore, Joseph" <muratore AT bnl.gov>
- To: "Yip, Kin" <kinyip AT bnl.gov>
- Cc: "sphenix-magnet-l AT lists.bnl.gov" <sphenix-magnet-l AT lists.bnl.gov>
- Subject: Re: [Sphenix-magnet-l] Can we do hipot at 90-100K ? ---- How about Jan. 3 ?
- Date: Tue, 2 Jan 2018 15:10:18 +0000
Hi Kin –
Normally we would wait to hipot at the test temperature (4.5K) since the hipot of a magnet in He gas is problematic. We hipot in liquid He to avoid the problems with gas. But since this magnet never touches He (conductive cooling, I believe), it should be OK. That being said, it takes maybe 10 min to do a hipot so you are not saving much time.
Joe
From: Yip, Kin
Hello Carl,
For the record, I’ve just attached another screenshot of the magnet temperatures (~102 K at this moment). You can see that the temperature should fall below 100K sometime today.
We need to give time for Mike Lenz to install the box/barrier around the current leads (at the Valvebox). In order to install the box, we need to connect the current leads with MCM cables. For that to happen, we need to finish the hipot.
Since the cool-down to 4K won’t happen this week, I hope to make use of the time a bit more productively. You don’t need to come in on Jan. 2, 2018. But can you please come in on Jan. 3 (as you originally plan anyway) to do the hipot at 90-something K ? You have all the hours to do so as nothing else would happen around the Magnet (while the Cryo group is working on the Plant).
Kin
From: Schultheiss, Carl
Hi, From: Kin Yip [kinyip AT bnl.gov] Hi,
Hi,
Roberto reported (email below) about a false ODH alarm which closed the LN2 supply valve and cause other damages. The “vacuum skid” didn’t concern us as we are not doing 2K Helium cool-down but only ~4K. I attach the temperature profile from a sensor at the superconducting bus and you could see when this “accident” happened --- the warm of a few K.
In any case, the Cryo group has shut down the (He) plant whereas the LN2 continues to cool to LN2 temperature. They’re warming up the Plant and hopefully, they’ll clear the “plug” (blockage) next week and cool-down the plant again. The original plan of going to cool from ~90/100 K to 4K on Jan. 2, 2017 will not happen. Since it’ll take a couple days to cool-down the plant itself, it’s not very likely we’ll see 4K next week.
I wonder whether we can do the “hipot” at ~90/100 K or so as there isn’t much contraction between 100K and 4 K anyway.
I also talked with Jonathan Reich and actually told what phone to “catch” Roberto. There were actually two alarms. He’s already replaced the sensor in the “compressor room” which has nothing to do with the LN2 valve closing. He’ll try to see whether he can replace the other ODH alarm in 912 (which has been giving a lot of false alarms) next week while we’re still having cryo operation in 912.
Kin
From: Than, Yatming (Roberto)
Kin,
We had an ODH alarm event in 912 around overnight. Shutdown the LN2 supply valve [automatic] into the bldg. to cryo plant and sPHENIX LN2 cooler.
The fans kicked on in the compressor room and temperature dropped sufficient and the water lines in the compressor room and water/oil heatexchangers on the vacuum skid cracked. The water tower eventually tripped on low water since the water was not returning to the tower. Sullair helium compressor for the cryoplant tripped. System is restarted but the cryoplant is not performing. I have inlet pressure but the cold end of the plant and expander have to high inlet pressure. Appears I have plug in the plant. So shut down plant.
We have next week to get the plant back up.
Roberto
|
-
Re: [Sphenix-magnet-l] Can we do hipot at 90-100K ? ---- How about Jan. 3 ?,
Yip, Kin, 01/02/2018
-
Re: [Sphenix-magnet-l] Can we do hipot at 90-100K ? ---- How about Jan. 3 ?,
Muratore, Joseph, 01/02/2018
-
Re: [Sphenix-magnet-l] Can we do hipot at 90-100K ? ---- How about Jan. 3 ?,
Kin Yip, 01/02/2018
-
[Sphenix-magnet-l] Hipot tomorrow ... Box installation (around the current leads) after that ...,
Kin Yip, 01/02/2018
- Re: [Sphenix-magnet-l] Hipot tomorrow ... Box installation (around the current leads) after that ..., Phillips, David B, 01/02/2018
-
[Sphenix-magnet-l] Hipot tomorrow ... Box installation (around the current leads) after that ...,
Kin Yip, 01/02/2018
-
Re: [Sphenix-magnet-l] Can we do hipot at 90-100K ? ---- How about Jan. 3 ?,
Kin Yip, 01/02/2018
-
Re: [Sphenix-magnet-l] Can we do hipot at 90-100K ? ---- How about Jan. 3 ?,
Muratore, Joseph, 01/02/2018