Re: [Sphenix-magnet-l] Magnet reached top current (4596 A) ... Tiny magnetic field at the top platform ... He-Gas-flow needs to be adjusted
- From: Edward O'Brien <eobrien AT bnl.gov>
- To: "Yip, Kin" <kinyip AT bnl.gov>, "sphenix-magnet-l AT lists.bnl.gov" <sphenix-magnet-l AT lists.bnl.gov>
- Cc: "Michnoff, Robert J" <michnoff AT bnl.gov>, "Sandberg, Jon N" <jsandberg AT bnl.gov>, "Albanese, Marianna" <malbanese AT bnl.gov>
- Subject: Re: [Sphenix-magnet-l] Magnet reached top current (4596 A) ... Tiny magnetic field at the top platform ... He-Gas-flow needs to be adjusted
- Date: Mon, 24 Oct 2022 23:50:40 -0400
Hi Kin, Thank you for the summary. I wouldn't describe the sound that we heard at the time of the forced-quench to be a "bang" but instead a loud "whoosh". I was surprised that we could hear the sound in the Assembly Hall because the gas was venting to the outside of the building, but we could definitely hear the noise of the gas discharge. Ed On 10/24/22 6:45 PM, Yip, Kin wrote:
Hi,
in the DCCT (which measures 0 to 10 V for current of 0 to 6000 A). We’ve found that we needed to set the “targetCurrentS” at 4620 A in order to get 4596 A in DCCT. I show the current plot in “current.gif”.
Tim Costanzo, Ed O’Brien, Chris O’Shea and I were in the 1008 and we heard the big bang/noise. But there was no smoke in the IR !
The thermal sensor for the magnet went from 4.8 K to >35 K due to the quench😊
about 0.17 G and if you walked
across it a few steps, it could be about 0.32 G. saw that it almost went away to above 40 mV --- 50 mV is our quench threshold. You may see this in the plot of “gas-cooled-lead.gif”. So, Carl immediately set the current down to 3500 A and called “Cryo”. ( This is the “dip” in “current.gif” after 4 pm. ) “Cryo” has had wrong current reading from DCCT today. When we read 4000 A, they might be reading between 1700 or 1800 A or so. They have tried to adjust the flow rate and at times, they thought it’s too high ---- I took a picture of the Current Leads (“20221024_170849.jpg”) during the magnetic field measurement and you can see a lot frost --- they tried to adjust it lower … Too high gas flow is no good because it might block or damage the gas pipe. The wrong reading may be due to the fact that it should be voltage reading instead of current reading as it’s assumed and they may need to change a wire or two. After “Cryo” told us that they temporarily fixed the problem (for now) and got steady, we stopped the down-ramping and went back to 4000 A and …. then 4596 A. Tomorrow, Carl will work the Cryo Group (Roberto and Brian) so that the Cryo group can optimize at each step (1000 A, 2000 A, …. ).
Overall, it’s a successful day and we ramped to the top without any uninitiated quench. Pablo doesn’t seem to make any “compensation” adjustment as the parameters used in 2018 seem good enough.
Kin
_______________________________________________ sPHENIX-magnet-l mailing list sPHENIX-magnet-l AT lists.bnl.gov https://lists.bnl.gov/mailman/listinfo/sphenix-magnet-l |
-
[Sphenix-magnet-l] Magnet reached top current (4596 A) ... Tiny magnetic field at the top platform ... He-Gas-flow needs to be adjusted,
Yip, Kin, 10/24/2022
- Re: [Sphenix-magnet-l] Magnet reached top current (4596 A) ... Tiny magnetic field at the top platform ... He-Gas-flow needs to be adjusted, Sondheim, Walter E, 10/24/2022
- Re: [Sphenix-magnet-l] Magnet reached top current (4596 A) ... Tiny magnetic field at the top platform ... He-Gas-flow needs to be adjusted, Edward O'Brien, 10/24/2022
-
Message not available
-
Message not available
- [Sphenix-magnet-l] Cryo seems to recover within in an hour ..., Yip, Kin, 10/25/2022
-
Message not available
Archive powered by MHonArc 2.6.24.