star-fst-l AT lists.bnl.gov
Subject: Star-fst-l mailing list
List archive
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?
- From: "Ye, Zhenyu" <yezhenyu AT uic.edu>
- To: Gerard Visser <gvisser AT indiana.edu>
- Cc: "star-fst-l AT lists.bnl.gov" <star-fst-l AT lists.bnl.gov>
- Subject: Re: [Star-fst-l] noise vs bias for worst channel and typical channel?
- Date: Thu, 13 Aug 2020 04:41:39 +0000
> On Aug 12, 2020, at 11:21 PM, Gerard Visser <gvisser AT indiana.edu> wrote:
>
> hi Zhenyu,
> I can readily believe that. Did you guys try a plastic plate?
We got large CMN noise, comparable to that w/o grounding the Al plate
> BTW, I hope we don't have problems with noise pickup from the
> beampipe at STAR. I don't know how the beampipe is grounded but I guess not
> directly grounded to STAR clean ground which the PPB (and so the APV) is
> grounded to. Hadn't thought about that yet, but this may be a source of
> trouble. We need to look into it.
>
> Gerard
>
>
> On 8/13/2020 12:14 AM, Ye, Zhenyu wrote:
>> Hi Gerard,
>> Page 12
>> https://drupal.star.bnl.gov/STAR/system/files/FstReviewAug2020_v10.pdf
>> shows the test stand setup.
>> The plate with many holes is made from Aluminum. The FST module is placed
>> inside a plastic storage box, which is placed on top of the Al plate.
>> Underneath the plate are the IST staves.
>> We found that grounding the plate makes a large difference to the FST CMN
>> (i.e., the CMN is much smaller by electrically connecting the Al plate to
>> the grey cable shield than no connection), but not so much difference in
>> IST. What Xu shows in his slides are with grounding the Al plate.
>> Zhenyu
>>> On Aug 12, 2020, at 10:55 PM, Gerard Visser <gvisser AT indiana.edu> wrote:
>>>
>>> hi Xu, Zhenyu,
>>> So it is the two smallest radius (farthest from readout) rings of the
>>> sensors which are extra noisy in both the inner and outer sensor case.
>>> That is curious indeed.
>>> Capacitances through the bulk of the sensor should be smaller there.
>>> I thought that sort of capacitance would be the worst for CMN.
>>> Routing line capacitance and resistance is of course largest there.
>>> What is the routing line resistance, is there any estimae from Hamamatsu?
>>> That resistance adds noise voltage, I think with basically the same
>>> effect as noise voltage of the frontend amplifier. I guess one would want
>>> 1-2 nV/sqrt(Hz) at most (though the number is really a guess; for sure we
>>> want "not more than about the voltage noise of the APV frontend" whatever
>>> that is). 2 nV/sqrt(Hz) would be 240 Ohms -- how does it compare w/ the
>>> sensor design?
>>> I don't understand though how any noise from the routing line
>>> resistance would appear as CMN.
>>> A different theory is that these rings are farther from the bypass
>>> capacitors on the nybrid (inevitably). The bias voltage there might be
>>> picking up some external noise more than it is closer to these
>>> capacitors. (If so, it could maybe be improved in our design.)
>>> Sincerely,
>>>
>>> Gerard
>>>
>>> On 8/12/2020 11:35 PM, Xu Sun wrote:
>>>> Hi Gerard,
>>>> 0-3 is for the inner sector from smallest radius to largest radius. And
>>>> 4-7 is for the outer sector from smallest radius to largest radius.
>>>> Hope this helps.
>>>> Best,
>>>> Xu
>>>> On Wed, Aug 12, 2020 at 10:29 PM Gerard Visser <gvisser AT indiana.edu
>>>> <mailto:gvisser AT indiana.edu>> wrote:
>>>> hi Xu,
>>>> Thanks, that is very good to see these plots including also
>>>> the
>>>> case from
>>>> unbonded APV on the IST hybrids (+IST cable and T-board). I am glad
>>>> to see that
>>>> the case of FST and IST unbonded chips is about the same. We do have
>>>> to
>>>> decipher
>>>> still if the difference between FST and IST with detector indicates a
>>>> problem or
>>>> merely the normal effect of different detector capacitances. I don't
>>>> think
>>>> it is
>>>> clear; but of course it _could_ be a problem.
>>>> I don't know the definition for R strip number 0-7, but your
>>>> plots
>>>> show that
>>>> for R-strips 2, 3, 6, and 7 the CMN is about the same as the unbonded
>>>> case. So,
>>>> which strips are those?
>>>> Sincerely,
>>>> Gerard
>>>> On 8/12/2020 10:44 PM, Xu Sun wrote:
>>>> > Hi Gerard,
>>>> >
>>>> > We just had a quick look at 2013 IST data before and after sensor
>>>> mounting and
>>>> > compared it with current FST data. You could find a summary slides
>>>> in the
>>>> > attached file.
>>>> > The total noise and random noise of IST increased significantly
>>>> after sensor
>>>> > mounting, but the CMN from IST data didn't. This suggests that CMN
>>>> is only
>>>> > weakly dependent on detector capacitance in the case of IST.
>>>> >
>>>> > For FST, total noise, CMN and random noise increased after sensor
>>>> mounting.
>>>> >
>>>> > Best,
>>>> >
>>>> > Xu
>>>> >
>>>> > On Tue, Aug 11, 2020 at 6:27 PM Visser, Gerard <gvisser AT indiana.edu
>>>> <mailto:gvisser AT indiana.edu>
>>>> > <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>>> wrote:
>>>> >
>>>> > hi Zhenyu,
>>>> > All I am saying is that these plots do not indicate to me
>>>> evidence of a
>>>> > problem. If the IST capacitance is "negligible enough" for
>>>> instance
>>>> it would
>>>> > look similar to FST unbonded. I wish we had such plots for an
>>>> unbonded IST
>>>> > chip...
>>>> > The pattern of higher noise every four timebins for sure
>>>> looks like
>>>> > something we don't want to see, but I think explanation for
>>>> _that_
>>>> ties into
>>>> > some internal details of the APV chip that we have no choice
>>>> but to
>>>> live with.
>>>> > Anyway, I think we need to keep investigating, we agree.
>>>> At the
>>>> moment I
>>>> > would say I believe the FST noise looks a bit worse than IST
>>>> simply
>>>> because
>>>> > the capacitances are worse, but if we think the S/N ratio is
>>>> still
>>>> adequate
>>>> > I don't see a clear worry. Of course we should check
>>>> everything that
>>>> we can
>>>> > though.
>>>> >
>>>> > Gerard
>>>> >
>>>> >
>>>> > p.s. At least the capacitance of IST to the backside contact /
>>>> bias
>>>> supply
>>>> > is << than that of FST. This could be the capacitance that
>>>> matters
>>>> most for
>>>> > CMN. (I'd naively expect that capacitance to at least adjacent
>>>> channels of
>>>> > same chip, does not generate CMN, or not significantly.)
>>>> >
>>>> > ________________________________________
>>>> > From: Zhenyu Ye <yezhenyu2003 AT gmail.com
>>>> <mailto:yezhenyu2003 AT gmail.com> <mailto:yezhenyu2003 AT gmail.com
>>>> <mailto:yezhenyu2003 AT gmail.com>>>
>>>> > Sent: Tuesday, August 11, 2020 7:14 PM
>>>> > To: Visser, Gerard
>>>> > Cc: Xu Sun; star-fst-l AT lists.bnl.gov
>>>> <mailto:star-fst-l AT lists.bnl.gov> <mailto:star-fst-l AT lists.bnl.gov
>>>> <mailto:star-fst-l AT lists.bnl.gov>>
>>>> > Subject: Re: [Star-fst-l] noise vs bias for worst channel and
>>>> typical channel?
>>>> >
>>>> > Hi Gerard,
>>>> >
>>>> > In the other email, you argued that the CMN increases with
>>>> input
>>>> capacitance
>>>> > (I agree). But the argument you are making here, which is
>>>> based on
>>>> "CMN: IST
>>>> > with sensor ~ FST w/o sensor" and concluding there is no issue
>>>> with FST,
>>>> > seems to neglect the fact that the input capacitance is
>>>> different between
>>>> > IST with sensor and FST w/o sensor.
>>>> >
>>>> > CMN:
>>>> > FST with sensor >> FST w/o sensor ~ IST with sensor
>>>> >
>>>> > CMN/Total
>>>> > FST with sensor ~ FST w/o sensor >> IST with sensor
>>>> >
>>>> > Zhenyu
>>>> >
>>>> >
>>>> > > On Aug 11, 2020, at 5:59 PM, Gerard Visser
>>>> <gvisser AT indiana.edu
>>>> <mailto:gvisser AT indiana.edu>
>>>> > <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>>>
>>>> wrote:
>>>> > >
>>>> > > hi Xu, Zhenyu,
>>>> > > I put the plot from IST that Xu sent the other day on
>>>> a slide
>>>> > (attached) together with the plot from FST03 with no detector
>>>> that he
>>>> sent
>>>> > today, at the same scale. I think it looks fairly comparable,
>>>> there is no
>>>> > obvious reason I can see that these results indicate a problem.
>>>> > > I wish we had some understanding why the noise would
>>>> be larger
>>>> > every 4th timebin. I have no clue about that. It is especially
>>>> odd since
>>>> > things (readout, in particular) happen in the APV chip in
>>>> groups of three
>>>> > timebins, but I don't know of anything except some "black-box"
>>>> internal
>>>> > feature that happens in groups of four.
>>>> > > It is very likely related to some internal feature of
>>>> the
>>>> APV. That
>>>> > doesn't mean we can conclude that it is not affected by some
>>>> factor under
>>>> > our control though, of course.
>>>> > > Sincerely,
>>>> > >
>>>> > > Gerard
>>>> > >
>>>> > >
>>>> > > On 8/11/2020 5:56 PM, Xu Sun wrote:
>>>> > >> Hi Gerard,
>>>> > >> As Zhenyu mentioned, please find the plot of noise before
>>>> sensor
>>>> > mounting (APV chips only) of FST03 attached. Please ignore the
>>>> APV 7
>>>> which
>>>> > shows an abnormal behavior, it is from the cable we used to
>>>> take the
>>>> noise run.
>>>> > >> We see the large CMN with strong time-bin dependence with
>>>> the
>>>> noise from
>>>> > APV chips only. This time bin dependence is the same as noise
>>>> with sensor
>>>> > mounted, therefore it is very likely from APV chips on FST.
>>>> > >> Best,
>>>> > >> Xu
>>>> > >> On Tue, Aug 11, 2020 at 4:28 PM Zhenyu Ye
>>>> <yezhenyu2003 AT gmail.com
>>>> <mailto:yezhenyu2003 AT gmail.com>
>>>> > <mailto:yezhenyu2003 AT gmail.com <mailto:yezhenyu2003 AT gmail.com>>
>>>> <mailto:yezhenyu2003 AT gmail.com <mailto:yezhenyu2003 AT gmail.com>
>>>> > <mailto:yezhenyu2003 AT gmail.com
>>>> <mailto:yezhenyu2003 AT gmail.com>>>> wrote:
>>>> > >> Hi Gerard,
>>>> > >> The total capacitance of a readout strip for silicon
>>>> strip
>>>> detectors is
>>>> > >> given by the sum of the strip-backplane capacitance,
>>>> and the
>>>> inter-strip
>>>> > >> capacitance. The latter usually dominates.
>>>> > >> As we know, the APV noise is linearly dependent on input
>>>> capacitance,
>>>> > i.e.,
>>>> > >> the total capacitance of the readout strip. For FST, we
>>>> have seen
>>>> > that outer
>>>> > >> (larger R, Rstrip3 for the inner sensor or Rstrip7 for
>>>> the outer
>>>> > sensor in
>>>> > >> Xu’s plots) strips have much smaller noises than the
>>>> inner strips
>>>> > (Rstrip0
>>>> > >> for the inner sensor or Rstrip4 for the outer sensor),
>>>> despite the
>>>> > fact that
>>>> > >> outer strips have larger area and thus large
>>>> strip-backplane
>>>> capacitance.
>>>> > >> This is consistent with the above statement that the
>>>> interstrip
>>>> > capacitance
>>>> > >> dominates.
>>>> > >> The CMN, by definition, affects a group of channels in
>>>> a coherent
>>>> > way. It is
>>>> > >> usually caused by a common electromagnetic pick-up, or
>>>> noise
>>>> on the
>>>> > supply
>>>> > >> voltage, etc. The fact that we see large CMN in FST,
>>>> suggests
>>>> to me
>>>> > that we
>>>> > >> should check FST inner cable/T-board/hybrid, such as
>>>> grounding.
>>>> > >> P.S. Xu will send some plots on FST prototype modules
>>>> before
>>>> sensors were
>>>> > >> mounted. We see large CMN with strong time-bin
>>>> dependence. I think
>>>> > this is
>>>> > >> consistent with my above assessment.
>>>> > >> Best,
>>>> > >> Zhenyu
>>>> > >> > On Aug 10, 2020, at 3:45 PM, Gerard Visser
>>>> <gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>
>>>> > <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>>
>>>> > >> <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>
>>>> <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>>>> wrote:
>>>> > >> >
>>>> > >> > hi Xu, Zhenyu,
>>>> > >> > Thanks; I didn't realize about this timebin
>>>> dependence in
>>>> > the IST.
>>>> > >> By the way do we see any dependence of noise on the
>>>> "cap id"
>>>> (i.e. the
>>>> > >> "address" reported in the APV header)? I am assuming we
>>>> are
>>>> > triggering at a
>>>> > >> low rate and asynchronous to the clock (ARC-II local
>>>> clock) so we
>>>> > should be
>>>> > >> getting data in all values of capid. This could show
>>>> some
>>>> features (and
>>>> > >> perhaps can allow for some capid-dependent correction
>>>> applied
>>>> offline).
>>>> > >> >
>>>> > >> > I make some estimate of the capacitances as
>>>> follows:
>>>> > Neglecting the
>>>> > >> capacitance to neighbor pads, and all the capacitance
>>>> of the
>>>> routing
>>>> > line on
>>>> > >> the detector, let's consider only the
>>>> infinite-parallel-plate
>>>> > capacitance in
>>>> > >> bulk of the detector:
>>>> > >> >
>>>> > >> > IST: Pad size is 594 x 6275 um, thickness 300 um,
>>>> k=11.7 ==>
>>>> > C=1.3 pF (I
>>>> > >> think this must certainly then be dominated by the
>>>> other,
>>>> neglected
>>>> > >> capacitances mentioned above).
>>>> > >> >
>>>> > >> > FST: worst case outer pad size is (about) 1087 x
>>>> 28750 um.
>>>> (Right?
>>>> > If you
>>>> > >> have more precise info please say.) Also 300 um thick.
>>>> ==>
>>>> C=10.8 pF. (+
>>>> > >> other again)
>>>> > >> >
>>>> > >> > If you have some info about the gap in
>>>> contact/metal
>>>> between
>>>> > >> adjacent pads in the case of IST and FST I could try to
>>>> roughly
>>>> > estimate the
>>>> > >> perimeter capacitance to neighbor pads.
>>>> > >> > If you have some measured capacitance info or
>>>> real
>>>> calculated
>>>> > >> capacitances from detector design, of course we could
>>>> better think
>>>> > about those.
>>>> > >> > Anyway, my guess is that the capacitances in
>>>> FST
>>>> (outer at
>>>> > least)
>>>> > >> are probably 2-3x the capacitances in IST. This is
>>>> probably
>>>> > responsible for
>>>> > >> the larger common mode noise and larger noise. If so
>>>> unfortunately it
>>>> > >> probably means there is nothing that we can do about it.
>>>> > >> > If we have an IST stave with a defective
>>>> detector and
>>>> wanted
>>>> > to do
>>>> > >> further tests, we could bond some APV input pads to test
>>>> capacitors
>>>> > of value
>>>> > >> similar to the FST detector and see how that looks. I
>>>> think this
>>>> > could be a
>>>> > >> significant effort though.
>>>> > >> > Really the only question that we must answer is
>>>> whether the PPB,
>>>> > >> purple cable, T-board, and hybrids are working well
>>>> together
>>>> to deliver
>>>> > >> clean supply voltages and clock/trigger signals to the
>>>> APV
>>>> chips (and
>>>> > clean
>>>> > >> bias to the sensors). This is probably so, I think all
>>>> your noise
>>>> > plots look
>>>> > >> reasonable, but we could try to check more directly
>>>> with low noise
>>>> > probing
>>>> > >> of the supply voltages on one of the prototype hybrids.
>>>> We
>>>> probably would
>>>> > >> have done this already if it weren't for the virus
>>>> situation.
>>>> > >> > Sincerely,
>>>> > >> >
>>>> > >> > Gerard
>>>> > >> >
>>>> > >> >
>>>> > >> > p.s. I suppose that, at least for lower trigger
>>>> rates <3
>>>> kHz or so, we
>>>> > >> should consider to setup to read 4 timebins and ignore
>>>> 0 in
>>>> offline, use
>>>> > >> only 1-3. If 3 timebins is all that we really need,
>>>> that is.
>>>> What do
>>>> > you think?
>>>> > >> >
>>>> > >> >
>>>> > >> > On 8/10/2020 2:29 PM, Xu Sun wrote:
>>>> > >> >> Hi Gerard,
>>>> > >> >> Sorry for the late reply. Please find the FST & IST
>>>> noise
>>>> study
>>>> > in the
>>>> > >> attached file.
>>>> > >> >> I see a similar behaviour for IST with a much
>>>> smaller
>>>> magnitude.
>>>> > >> >> Best,
>>>> > >> >> Xu
>>>> > >> >> On Tue, Aug 4, 2020 at 1:12 PM Gerard Visser
>>>> <gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>
>>>> > <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>>
>>>> > >> <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>
>>>> <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>>>
>>>> > <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>
>>>> <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>>
>>>> > >> <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>
>>>> <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>>>>> wrote:
>>>> > >> >> Hi Xu, Zhenyu,
>>>> > >> >> Do we see the timebin-dependece of
>>>> noise as
>>>> you show here
>>>> > >> >>
>>>> > >>
>>>> >
>>>>
>>>> https://drupal.star.bnl.gov/STAR/event/2020/05/04/star-forward-silicon-tracker-meeting/prototype-module-assembly-and-test
>>>> > >> >> in the IST data too?
>>>> > >> >> Thanks,
>>>> > >> >> Gerard
>>>> > >> >> p.s. And, if possible to answer, there is also
>>>> the question
>>>> > whether
>>>> > >> this was
>>>> > >> >> seen in the IST installed in STAR/HFT? I don't
>>>> remember
>>>> > hearing about
>>>> > >> it before.
>>>> > >> >> On 8/4/2020 12:39 PM, Zhenyu Ye wrote:
>>>> > >> >> > Hi Gerard,
>>>> > >> >> >
>>>> > >> >> >> On Aug 4, 2020, at 11:19 AM, Gerard Visser
>>>> > <gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>
>>>> <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>>
>>>> > >> <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>
>>>> <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>>>
>>>> > >> >> <mailto:gvisser AT indiana.edu
>>>> <mailto:gvisser AT indiana.edu> <mailto:gvisser AT indiana.edu
>>>> <mailto:gvisser AT indiana.edu>>
>>>> > <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>
>>>> <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>>>>> wrote:
>>>> > >> >> >>
>>>> > >> >> >> hi Zhenyu,
>>>> > >> >> >> That timebon dependence sounds
>>>> definitely odd. Are
>>>> > you sure?
>>>> > >> Do we
>>>> > >> >> see that in IST too, only more mildly? I wasn't
>>>> aware
>>>> of this.
>>>> > >> >> > ‘
>>>> > >> >> > Please take a look at Xu’s presentation in FST
>>>> meeting on May 4
>>>> > >> (FST) and
>>>> > >> >> 11 (FST and IST):
>>>> > >> >> >
>>>> > >> >> >
>>>> > >> >>
>>>> > >>
>>>> >
>>>>
>>>> https://drupal.star.bnl.gov/STAR/event/2020/05/04/star-forward-silicon-tracker-meeting/prototype-module-assembly-and-test
>>>> > >> >> >
>>>> > >> >>
>>>> > >>
>>>> >
>>>>
>>>> https://drupal.star.bnl.gov/STAR/event/2020/05/11/star-forward-silicon-tracker-meeting/prototype-module-assembly-and-test
>>>> > >> >> >
>>>> > >> >> >> The capacitance of FST is much larger
>>>> than
>>>> IST, I think.
>>>> > >> This may
>>>> > >> >> certainly be relevant. For sure it is relevant
>>>> to CMN.
>>>> > >> >> >> Anyway, I agree we should investigate
>>>> these noise
>>>> > issues, I
>>>> > >> do not
>>>> > >> >> like to ignore them. However on the other hand
>>>> it _may_
>>>> be an
>>>> > inherent
>>>> > >> >> property of APV chips.
>>>> > >> >> >
>>>> > >> >> >> - Gerard
>>>> > >> >> >>
>>>> > >> >> >>
>>>> > >> >> >>
>>>> > >> >> >> On 8/4/2020 12:12 PM, Zhenyu Ye wrote:
>>>> > >> >> >>> Hi Gerard,
>>>> > >> >> >>> From the plots that Xu sent, there is a
>>>> clear pattern
>>>> > where the
>>>> > >> >> channels showing enhanced noise level after
>>>> mounting
>>>> the sensors,
>>>> > >> also show
>>>> > >> >> enhanced noise level before mounting the sensors.
>>>> > >> >> >>> An independent topic, we see that the CMN
>>>> in FST
>>>> is much
>>>> > higher than
>>>> > >> >> IST, and show a strong time-bin dependence,
>>>> i.e., when
>>>> we read
>>>> > in 9 time
>>>> > >> >> bins, the 1st, 5th and 9th time bins have much
>>>> higher
>>>> CMN than
>>>> > the other
>>>> > >> >> time bins. I don’t feel comfortable to ignore it
>>>> w/o
>>>> knowing the
>>>> > >> cause, as
>>>> > >> >> it may get worse in the real experiment.
>>>> > >> >> >>> Best,
>>>> > >> >> >>> Zhenyu
>>>> > >> >> _______________________________________________
>>>> > >> >> Star-fst-l mailing list
>>>> > >> >> Star-fst-l AT lists.bnl.gov
>>>> <mailto:Star-fst-l AT lists.bnl.gov>
>>>> <mailto:Star-fst-l AT lists.bnl.gov <mailto:Star-fst-l AT lists.bnl.gov>>
>>>> > <mailto:Star-fst-l AT lists.bnl.gov
>>>> <mailto:Star-fst-l AT lists.bnl.gov>
>>>> <mailto:Star-fst-l AT lists.bnl.gov <mailto:Star-fst-l AT lists.bnl.gov>>>
>>>> > >> <mailto:Star-fst-l AT lists.bnl.gov
>>>> <mailto:Star-fst-l AT lists.bnl.gov> <mailto:Star-fst-l AT lists.bnl.gov
>>>> <mailto:Star-fst-l AT lists.bnl.gov>>
>>>> > <mailto:Star-fst-l AT lists.bnl.gov
>>>> <mailto:Star-fst-l AT lists.bnl.gov>
>>>> <mailto:Star-fst-l AT lists.bnl.gov <mailto:Star-fst-l AT lists.bnl.gov>>>>
>>>> > >> >> https://lists.bnl.gov/mailman/listinfo/star-fst-l
>>>> > >> > _______________________________________________
>>>> > >> > Star-fst-l mailing list
>>>> > >> > Star-fst-l AT lists.bnl.gov
>>>> <mailto:Star-fst-l AT lists.bnl.gov>
>>>> <mailto:Star-fst-l AT lists.bnl.gov <mailto:Star-fst-l AT lists.bnl.gov>>
>>>> > <mailto:Star-fst-l AT lists.bnl.gov
>>>> <mailto:Star-fst-l AT lists.bnl.gov>
>>>> <mailto:Star-fst-l AT lists.bnl.gov <mailto:Star-fst-l AT lists.bnl.gov>>>
>>>> > >> > https://lists.bnl.gov/mailman/listinfo/star-fst-l
>>>> > >
>>>>
>>>> <cmn_noise_comparison.pdf>_______________________________________________
>>>> > > Star-fst-l mailing list
>>>> > > Star-fst-l AT lists.bnl.gov <mailto:Star-fst-l AT lists.bnl.gov>
>>>> <mailto:Star-fst-l AT lists.bnl.gov <mailto:Star-fst-l AT lists.bnl.gov>>
>>>> > > https://lists.bnl.gov/mailman/listinfo/star-fst-l
>>>> >
>>> _______________________________________________
>>> Star-fst-l mailing list
>>> Star-fst-l AT lists.bnl.gov
>>> https://lists.bnl.gov/mailman/listinfo/star-fst-l
> _______________________________________________
> Star-fst-l mailing list
> Star-fst-l AT lists.bnl.gov
> https://lists.bnl.gov/mailman/listinfo/star-fst-l
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?
, (continued)
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Xu Sun, 08/11/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Gerard Visser, 08/11/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Zhenyu Ye, 08/11/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Visser, Gerard, 08/11/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Xu Sun, 08/12/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Gerard Visser, 08/12/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Xu Sun, 08/12/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Gerard Visser, 08/12/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Ye, Zhenyu, 08/13/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Gerard Visser, 08/13/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Ye, Zhenyu, 08/13/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Xu Sun, 08/18/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Gerard Visser, 08/11/2020
Archive powered by MHonArc 2.6.24.