hi Xu, Zhenyu,
I put the plot from IST that Xu sent the other day on a slide (attached) together with the plot from FST03 with no detector that he sent today, at the same scale. I think it looks fairly comparable, there is no obvious reason I can see that these results indicate a problem.
I wish we had some understanding why the noise would be larger every 4th timebin. I have no clue about that. It is especially odd since things (readout, in particular) happen in the APV chip in groups of three timebins, but I don't know of anything except some "black-box" internal feature that happens in groups of four.
It is very likely related to some internal feature of the APV. That doesn't mean we can conclude that it is not affected by some factor under our control though, of course.
Sincerely,
Gerard
On 8/11/2020 5:56 PM, Xu Sun wrote:
Hi Gerard,
As Zhenyu mentioned, please find the plot of noise before sensor mounting (APV chips only) of FST03 attached. Please ignore the APV 7 which shows an abnormal behavior, it is from the cable we used to take the noise run.
We see the large CMN with strong time-bin dependence with the noise from APV chips only. This time bin dependence is the same as noise with sensor mounted, therefore it is very likely from APV chips on FST.
Best,
Xu
On Tue, Aug 11, 2020 at 4:28 PM Zhenyu Ye <yezhenyu2003 AT gmail.com <mailto:yezhenyu2003 AT gmail.com>> wrote:
Hi Gerard,
The total capacitance of a readout strip for silicon strip detectors is
given by the sum of the strip-backplane capacitance, and the inter-strip
capacitance. The latter usually dominates.
As we know, the APV noise is linearly dependent on input capacitance,
i.e.,
the total capacitance of the readout strip. For FST, we have seen that
outer
(larger R, Rstrip3 for the inner sensor or Rstrip7 for the outer sensor in
Xu’s plots) strips have much smaller noises than the inner strips (Rstrip0
for the inner sensor or Rstrip4 for the outer sensor), despite the fact
that
outer strips have larger area and thus large strip-backplane capacitance.
This is consistent with the above statement that the interstrip
capacitance
dominates.
The CMN, by definition, affects a group of channels in a coherent way. It
is
usually caused by a common electromagnetic pick-up, or noise on the supply
voltage, etc. The fact that we see large CMN in FST, suggests to me that
we
should check FST inner cable/T-board/hybrid, such as grounding.
P.S. Xu will send some plots on FST prototype modules before sensors were
mounted. We see large CMN with strong time-bin dependence. I think this is
consistent with my above assessment.
Best,
Zhenyu
> On Aug 10, 2020, at 3:45 PM, Gerard Visser <gvisser AT indiana.edu
<mailto:gvisser AT indiana.edu>> wrote:
>
> hi Xu, Zhenyu,
> Thanks; I didn't realize about this timebin dependence in the
IST.
By the way do we see any dependence of noise on the "cap id" (i.e. the
"address" reported in the APV header)? I am assuming we are triggering at
a
low rate and asynchronous to the clock (ARC-II local clock) so we should
be
getting data in all values of capid. This could show some features (and
perhaps can allow for some capid-dependent correction applied offline).
>
> I make some estimate of the capacitances as follows: Neglecting
the
capacitance to neighbor pads, and all the capacitance of the routing line
on
the detector, let's consider only the infinite-parallel-plate capacitance
in
bulk of the detector:
>
> IST: Pad size is 594 x 6275 um, thickness 300 um, k=11.7 ==> C=1.3 pF
(I
think this must certainly then be dominated by the other, neglected
capacitances mentioned above).
>
> FST: worst case outer pad size is (about) 1087 x 28750 um. (Right? If
you
have more precise info please say.) Also 300 um thick. ==> C=10.8 pF. (+
other again)
>
> If you have some info about the gap in contact/metal between
adjacent pads in the case of IST and FST I could try to roughly estimate
the
perimeter capacitance to neighbor pads.
> If you have some measured capacitance info or real calculated
capacitances from detector design, of course we could better think about
those.
> Anyway, my guess is that the capacitances in FST (outer at least)
are probably 2-3x the capacitances in IST. This is probably responsible
for
the larger common mode noise and larger noise. If so unfortunately it
probably means there is nothing that we can do about it.
> If we have an IST stave with a defective detector and wanted to
do
further tests, we could bond some APV input pads to test capacitors of
value
similar to the FST detector and see how that looks. I think this could be
a
significant effort though.
> Really the only question that we must answer is whether the PPB,
purple cable, T-board, and hybrids are working well together to deliver
clean supply voltages and clock/trigger signals to the APV chips (and
clean
bias to the sensors). This is probably so, I think all your noise plots
look
reasonable, but we could try to check more directly with low noise probing
of the supply voltages on one of the prototype hybrids. We probably would
have done this already if it weren't for the virus situation.
> Sincerely,
>
> Gerard
>
>
> p.s. I suppose that, at least for lower trigger rates <3 kHz or so, we
should consider to setup to read 4 timebins and ignore 0 in offline, use
only 1-3. If 3 timebins is all that we really need, that is. What do you
think?
>
>
> On 8/10/2020 2:29 PM, Xu Sun wrote:
>> Hi Gerard,
>> Sorry for the late reply. Please find the FST & IST noise study in the
attached file.
>> I see a similar behaviour for IST with a much smaller magnitude.
>> Best,
>> Xu
>> On Tue, Aug 4, 2020 at 1:12 PM Gerard Visser <gvisser AT indiana.edu
<mailto:gvisser AT indiana.edu> <mailto:gvisser AT indiana.edu
<mailto:gvisser AT indiana.edu>>> wrote:
>> Hi Xu, Zhenyu,
>> Do we see the timebin-dependece of noise as you show here
>>
https://drupal.star.bnl.gov/STAR/event/2020/05/04/star-forward-silicon-tracker-meeting/prototype-module-assembly-and-test
>> in the IST data too?
>> Thanks,
>> Gerard
>> p.s. And, if possible to answer, there is also the question whether
this was
>> seen in the IST installed in STAR/HFT? I don't remember hearing
about
it before.
>> On 8/4/2020 12:39 PM, Zhenyu Ye wrote:
>> > Hi Gerard,
>> >
>> >> On Aug 4, 2020, at 11:19 AM, Gerard Visser <gvisser AT indiana.edu
<mailto:gvisser AT indiana.edu>
>> <mailto:gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>>> wrote:
>> >>
>> >> hi Zhenyu,
>> >> That timebon dependence sounds definitely odd. Are you
sure?
Do we
>> see that in IST too, only more mildly? I wasn't aware of this.
>> > ‘
>> > Please take a look at Xu’s presentation in FST meeting on May 4
(FST) and
>> 11 (FST and IST):
>> >
>> >
>>
https://drupal.star.bnl.gov/STAR/event/2020/05/04/star-forward-silicon-tracker-meeting/prototype-module-assembly-and-test
>> >
>>
https://drupal.star.bnl.gov/STAR/event/2020/05/11/star-forward-silicon-tracker-meeting/prototype-module-assembly-and-test
>> >
>> >> The capacitance of FST is much larger than IST, I think.
This may
>> certainly be relevant. For sure it is relevant to CMN.
>> >> Anyway, I agree we should investigate these noise issues,
I
do not
>> like to ignore them. However on the other hand it _may_ be an
inherent
>> property of APV chips.
>> >
>> >> - Gerard
>> >>
>> >>
>> >>
>> >> On 8/4/2020 12:12 PM, Zhenyu Ye wrote:
>> >>> Hi Gerard,
>> >>> From the plots that Xu sent, there is a clear pattern where
the
>> channels showing enhanced noise level after mounting the sensors,
also show
>> enhanced noise level before mounting the sensors.
>> >>> An independent topic, we see that the CMN in FST is much
higher than
>> IST, and show a strong time-bin dependence, i.e., when we read in
9 time
>> bins, the 1st, 5th and 9th time bins have much higher CMN than the
other
>> time bins. I don’t feel comfortable to ignore it w/o knowing the
cause, as
>> it may get worse in the real experiment.
>> >>> Best,
>> >>> Zhenyu
>> _______________________________________________
>> Star-fst-l mailing list
>> Star-fst-l AT lists.bnl.gov <mailto:Star-fst-l AT lists.bnl.gov>
<mailto:Star-fst-l AT lists.bnl.gov <mailto:Star-fst-l AT lists.bnl.gov>>
>> https://lists.bnl.gov/mailman/listinfo/star-fst-l
> _______________________________________________
> Star-fst-l mailing list
> Star-fst-l AT lists.bnl.gov <mailto:Star-fst-l AT lists.bnl.gov>
> https://lists.bnl.gov/mailman/listinfo/star-fst-l
Attachment:
cmn_noise_comparison.pdf
Description: Adobe PDF document
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Zhenyu Ye, 08/04/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Gerard Visser, 08/04/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Zhenyu Ye, 08/04/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Gerard Visser, 08/04/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Zhenyu Ye, 08/04/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Xu Sun, 08/10/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Gerard Visser, 08/10/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Zhenyu Ye, 08/11/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Xu Sun, 08/11/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Gerard Visser, 08/11/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Zhenyu Ye, 08/11/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Visser, Gerard, 08/11/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Xu Sun, 08/12/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Gerard Visser, 08/12/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Xu Sun, 08/12/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Gerard Visser, 08/12/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Ye, Zhenyu, 08/13/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Gerard Visser, 08/13/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Ye, Zhenyu, 08/13/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Xu Sun, 08/18/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Zhenyu Ye, 08/11/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Gerard Visser, 08/10/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Gerard Visser, 08/04/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Zhenyu Ye, 08/04/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Gerard Visser, 08/04/2020
Archive powered by MHonArc 2.6.24.