hi Xu, Zhenyu,
Thanks; I didn't realize about this timebin dependence in the IST. By the way do we see any dependence of noise on the "cap id" (i.e. the "address" reported in the APV header)? I am assuming we are triggering at a low rate and asynchronous to the clock (ARC-II local clock) so we should be getting data in all values of capid. This could show some features (and perhaps can allow for some capid-dependent correction applied offline).
I make some estimate of the capacitances as follows: Neglecting the capacitance to neighbor pads, and all the capacitance of the routing line on the detector, let's consider only the infinite-parallel-plate capacitance in bulk of the detector:
IST: Pad size is 594 x 6275 um, thickness 300 um, k=11.7 ==> C=1.3 pF (I think this must certainly then be dominated by the other, neglected capacitances mentioned above).
FST: worst case outer pad size is (about) 1087 x 28750 um. (Right? If you have more precise info please say.) Also 300 um thick. ==> C=10.8 pF. (+ other again)
If you have some info about the gap in contact/metal between adjacent pads in the case of IST and FST I could try to roughly estimate the perimeter capacitance to neighbor pads.
If you have some measured capacitance info or real calculated capacitances from detector design, of course we could better think about those.
Anyway, my guess is that the capacitances in FST (outer at least) are probably 2-3x the capacitances in IST. This is probably responsible for the larger common mode noise and larger noise. If so unfortunately it probably means there is nothing that we can do about it.
If we have an IST stave with a defective detector and wanted to do further tests, we could bond some APV input pads to test capacitors of value similar to the FST detector and see how that looks. I think this could be a significant effort though.
Really the only question that we must answer is whether the PPB, purple cable, T-board, and hybrids are working well together to deliver clean supply voltages and clock/trigger signals to the APV chips (and clean bias to the sensors). This is probably so, I think all your noise plots look reasonable, but we could try to check more directly with low noise probing of the supply voltages on one of the prototype hybrids. We probably would have done this already if it weren't for the virus situation.
Sincerely,
Gerard
p.s. I suppose that, at least for lower trigger rates <3 kHz or so, we should consider to setup to read 4 timebins and ignore 0 in offline, use only 1-3. If 3 timebins is all that we really need, that is. What do you think?
On 8/10/2020 2:29 PM, Xu Sun wrote:
Hi Gerard,
Sorry for the late reply. Please find the FST & IST noise study in the attached file.
I see a similar behaviour for IST with a much smaller magnitude.
Best,
Xu
On Tue, Aug 4, 2020 at 1:12 PM Gerard Visser <gvisser AT indiana.edu <mailto:gvisser AT indiana.edu>> wrote:
Hi Xu, Zhenyu,
Do we see the timebin-dependece of noise as you show here
https://drupal.star.bnl.gov/STAR/event/2020/05/04/star-forward-silicon-tracker-meeting/prototype-module-assembly-and-test
in the IST data too?
Thanks,
Gerard
p.s. And, if possible to answer, there is also the question whether this
was
seen in the IST installed in STAR/HFT? I don't remember hearing about it
before.
On 8/4/2020 12:39 PM, Zhenyu Ye wrote:
> Hi Gerard,
>
>> On Aug 4, 2020, at 11:19 AM, Gerard Visser <gvisser AT indiana.edu
<mailto:gvisser AT indiana.edu>> wrote:
>>
>> hi Zhenyu,
>> That timebon dependence sounds definitely odd. Are you sure? Do
we
see that in IST too, only more mildly? I wasn't aware of this.
> ‘
> Please take a look at Xu’s presentation in FST meeting on May 4 (FST)
and
11 (FST and IST):
>
>
https://drupal.star.bnl.gov/STAR/event/2020/05/04/star-forward-silicon-tracker-meeting/prototype-module-assembly-and-test
>
https://drupal.star.bnl.gov/STAR/event/2020/05/11/star-forward-silicon-tracker-meeting/prototype-module-assembly-and-test
>
>> The capacitance of FST is much larger than IST, I think. This may
certainly be relevant. For sure it is relevant to CMN.
>> Anyway, I agree we should investigate these noise issues, I do
not
like to ignore them. However on the other hand it _may_ be an inherent
property of APV chips.
>
>> - Gerard
>>
>>
>>
>> On 8/4/2020 12:12 PM, Zhenyu Ye wrote:
>>> Hi Gerard,
>>> From the plots that Xu sent, there is a clear pattern where the
channels showing enhanced noise level after mounting the sensors, also
show
enhanced noise level before mounting the sensors.
>>> An independent topic, we see that the CMN in FST is much higher than
IST, and show a strong time-bin dependence, i.e., when we read in 9 time
bins, the 1st, 5th and 9th time bins have much higher CMN than the other
time bins. I don’t feel comfortable to ignore it w/o knowing the cause, as
it may get worse in the real experiment.
>>> Best,
>>> Zhenyu
_______________________________________________
Star-fst-l mailing list
Star-fst-l AT lists.bnl.gov <mailto:Star-fst-l AT lists.bnl.gov>
https://lists.bnl.gov/mailman/listinfo/star-fst-l
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Zhenyu Ye, 08/04/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Gerard Visser, 08/04/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Zhenyu Ye, 08/04/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Gerard Visser, 08/04/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Zhenyu Ye, 08/04/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Xu Sun, 08/10/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Gerard Visser, 08/10/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Zhenyu Ye, 08/11/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Xu Sun, 08/11/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Gerard Visser, 08/11/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Zhenyu Ye, 08/11/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Visser, Gerard, 08/11/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Xu Sun, 08/12/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Gerard Visser, 08/12/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Xu Sun, 08/12/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Gerard Visser, 08/12/2020
- Re: [Star-fst-l] noise vs bias for worst channel and typical channel?, Ye, Zhenyu, 08/13/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Zhenyu Ye, 08/11/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Gerard Visser, 08/10/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Gerard Visser, 08/04/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Zhenyu Ye, 08/04/2020
-
Re: [Star-fst-l] noise vs bias for worst channel and typical channel?,
Gerard Visser, 08/04/2020
Archive powered by MHonArc 2.6.24.