star-qaboard-l@lists.bnl.gov
Subject: STAR QA Board
List archive
Re: [STAR-QAboard] 回覆: 回覆: Weekly QA Board Meeting - 02/June/2023 at noon 12 EDT
- From: "Tang, Aihong" <aihong@bnl.gov>
- To: Han-sheng Li <li3924@purdue.edu>, 李鸿灿 via STAR-QAboard-l <star-qaboard-l@lists.bnl.gov>
- Subject: Re: [STAR-QAboard] 回覆: 回覆: Weekly QA Board Meeting - 02/June/2023 at noon 12 EDT
- Date: Sat, 3 Jun 2023 11:45:50 +0000
Hi, Han-sheng :
Thanks for the message. So you rely on daily fast offline production to do this ? Note for this purpose, it does not have to be minbias, any collision trigger can serve the purpose.
For a single run, could you post example(s) somewhere ? 
If it is convenient for you, could you join today's 10 o'clock meeting and talk about your plots ?
Thanks,
Aihong
On Jun 3, 2023, at 6:43 AM, Han-sheng Li <li3924@purdue.edu> wrote:
Hi Aihong,
I want to correct one thing. the run ID was supposed to be up to 24147033 not 24149034 on my previous slides.
For your first request, due to the lack of minbias zdc trigger rootfiles I probably could not give the update today.For your second request, I have those plots for a single run but it not obvious to spot the issue with a single run.
Best regards,Han-Sheng
寄件者: Tang, Aihong <aihong@bnl.gov>
寄件日期: 2023年6月2日 下午 04:30
收件者: Han-sheng Li <li3924@purdue.edu>
副本: EsumiShinIchi <esumi.shinichi.gn@u.tsukuba.ac.jp>; 李鸿灿 via STAR-QAboard-l <star-qaboard-l@lists.bnl.gov>
主旨: Re: [STAR-QAboard] 回覆: Weekly QA Board Meeting - 02/June/2023 at noon 12 EDT
---- External Email: Use caution with attachments, links, or sharing data ---- 
Hi, Han-sheng :
That is great ! I have asked the shift to power cycle the QTD crate for ZDCSMD. The first meaningful run after the power cycling is 24153074. If you could repeat your study on tonight's data from that run and onwards, what will be helpful. I will bring up this for discussion at tomorrow's 10 o'clock meeting.
Thanks,Aihong
On Jun 2, 2023, at 9:42 PM, Han-sheng Li <li3924@purdue.edu> wrote:
Hi Aihong,
After power recycling I can repeat my study within couple hours.
Best regards,Han-Sheng
From: Tang, Aihong <aihong@bnl.gov>
Sent: Friday, June 2, 2023 9:34:13 PM
To: Han-sheng Li <li3924@purdue.edu>
Cc: EsumiShinIchi <esumi.shinichi.gn@u.tsukuba.ac.jp>; 李鸿灿 via STAR-QAboard-l <star-qaboard-l@lists.bnl.gov>; Tang, Aihong <aihong@bnl.gov>
Subject: Re: 回覆: 回覆: 回覆: 回覆: [STAR-QAboard] 回覆: Weekly QA Board Meeting - 02/June/2023 at noon 12 EDT
---- External Email: Use caution with attachments, links, or sharing data ---- 
Hi, Han-sheng :
Thanks for fast response !
It is not exactly stuck bits but I do see generally that when the distribution for lower position bits is smooth, the adc spectrum look normal. This is mostly obvious by looking at east vertical slats.
I will 1) forward your study to star-operation list, 2) ask the shift crew to power recycle the crate for ZDCSMD in a hour or so. I am not sure if it will fix the problem but that is what we can try for now. 3) talk about this at tomorrow's 10'clock meeting.
How soon can you repeat your study if we take runs after power recycling ? The shift QA histogram is not sensitive to this problem because it has 200 bins, not 2000 bins, for adc range from 0 to 2000.
Thanks,Aihong
On Jun 2, 2023, at 8:41 PM, Han-sheng Li <li3924@purdue.edu> wrote:
Hi Aihong,
I just updated the 12-bit binary plots in the updated slides:
The X-axis is the 12-bit positions, and Y-axis is the total counts.
The 12-bit binary plots are on pages 4, 6, 8, and 10, and the previous plots are on pages 3, 5, 7, and 9 for comparison.
Best regards,Han-Sheng
寄件者: Tang, Aihong <aihong@bnl.gov>
寄件日期: 2023年6月2日 上午 11:17
收件者: Han-sheng Li <li3924@purdue.edu>
副本: EsumiShinIchi <esumi.shinichi.gn@u.tsukuba.ac.jp>; 李鸿灿 via STAR-QAboard-l <star-qaboard-l@lists.bnl.gov>; Tang, Aihong <aihong@bnl.gov>
主旨: Re: 回覆: 回覆: 回覆: [STAR-QAboard] 回覆: Weekly QA Board Meeting - 02/June/2023 at noon 12 EDT
---- External Email: Use caution with attachments, links, or sharing data ---- 
Hi, Han-sheng :
Thank you very much for checking this. Since you have the information handy, I hope you don't mind doing a further check. To determine the presence of any stuck bits, would you please convert the ADC readings corresponding to an abnormal channel into a 12-bit binary representation ? You can then make a histogram to show the distribution of these binary bits. The X-axis should represent the 12-bit positions, and Y-axis should display the frequency (or total counts depending on your prefernce) for each bit being set to 1. In the event of a stuck bit issue, one particular bit will consistently remain either on (1) or off (0), and on the histogram it will have either the maximum possible counts or just zero.
You may also make a similar histogram for a normal channel, just for comparison/reference.
Thanks,Aihong
On Jun 2, 2023, at 4:25 PM, Han-sheng Li <li3924@purdue.edu> wrote:
Hi Aihong,
I checked that the adjacent bins are not zero, and doubled the bin width but the double entry still there.
Best regards,Han-Sheng
寄件者: Tang, Aihong <aihong@bnl.gov>
寄件日期: 2023年6月2日 上午 09:42
收件者: Han-sheng Li <li3924@purdue.edu>
副本: EsumiShinIchi <esumi.shinichi.gn@u.tsukuba.ac.jp>; 李鸿灿 via STAR-QAboard-l <star-qaboard-l@lists.bnl.gov>
主旨: Re: 回覆: 回覆: [STAR-QAboard] 回覆: Weekly QA Board Meeting - 02/June/2023 at noon 12 EDT
---- External Email: Use caution with attachments, links, or sharing data ---- 
Hi, Han-Sheng :
Thanks. I just randomly checked a few from your list, and I don't see this from online plots. For example, see page 29-32 of this file :
To me it looks like a digital rounding issue. For those bins with double entry, could you check if the adjacent bins got zero entry ?You have bin width corresponding to one adc count, what if you double the bin width ?
Thanks,
Aihong
On Jun 2, 2023, at 3:21 PM, Han-sheng Li <li3924@purdue.edu> wrote:
Hi Aihong,
The attachment is the runnumber list I used.
Best regards,Han-Sheng
寄件者: Tang, Aihong <aihong@bnl.gov>
寄件日期: 2023年6月2日 上午 09:06
收件者: Han-sheng Li <li3924@purdue.edu>
副本: EsumiShinIchi <esumi.shinichi.gn@u.tsukuba.ac.jp>; 李鸿灿 via STAR-QAboard-l <star-qaboard-l@lists.bnl.gov>
主旨: Re: 回覆: [STAR-QAboard] 回覆: Weekly QA Board Meeting - 02/June/2023 at noon 12 EDT<run.all.list>
---- External Email: Use caution with attachments, links, or sharing data ---- 
Hi, Han-sheng :
That is interesting. I have not noticed this in shift QA histograms. Could you give me an example runnumber so I can double check it in the JeffPlots ?
Thanks,
Aihong
On Jun 2, 2023, at 2:18 PM, Han-sheng Li <li3924@purdue.edu> wrote:
Dear ShinIchi,
Yes, our histograms have the number of bin per one adc count and this issue should come from the data itself.Does Aihong know how to deal with this issue?
Best regards,Han-Sheng
寄件者: EsumiShinIchi <esumi.shinichi.gn@u.tsukuba.ac.jp>
寄件日期: 2023年6月2日 上午 07:20
收件者: star-qaboard-l@lists.bnl.gov <star-qaboard-l@lists.bnl.gov>
副本: Han-sheng Li <li3924@purdue.edu>
主旨: Re: [STAR-QAboard] 回覆: Weekly QA Board Meeting - 02/June/2023 at noon 12 EDT
---- External Email: Use caution with attachments, links, or sharing data ---- 
Dear Han-ShengI would first like to make sure all your histograms have the number of bin per one adc count, right?If you count which bin number have double event entry counts, you might be able to tell which bithas problem in the ADC digitization depending on the SMD slat channel. Please talk to the read-outelectronics experts, maybe Aihong as Prithwish says (or maybe talk to Akio?) for ZDC/SMD. Theymight just need to power recycle the electronics or plug in/out the ADC board etc…Best regards, ShinIchi
On Jun 3, 2023, at 0:42, Han-sheng Li via STAR-QAboard-l <star-qaboard-l@lists.bnl.gov> wrote:
Hi Ashik and all,
I would like to give a report about ZDCSMD.Please find my slides in the following link:
Best regards,Han-Sheng
寄件者: STAR-QAboard-l <star-qaboard-l-bounces@lists.bnl.gov> 代表 Ray, Robert L via STAR-QAboard-l <star-qaboard-l@lists.bnl.gov>
寄件日期: 2023年6月2日 上午 05:34
收件者: ASHIK IKBAL <ashikhep@gmail.com>; star-qaboard-l <star-qaboard-l@lists.bnl.gov>; 李鸿灿 <lihc@mails.ccnu.edu.cn>
主旨: Re: [STAR-QAboard] Weekly QA Board Meeting - 02/June/2023 at noon 12 EDT_______________________________________________
---- External Email: Use caution with attachments, links, or sharing data ---- 
Hi Ashik and QA folks:
My slides for today's meeting are posted at:
Cheers,Lanny
From: STAR-QAboard-l <star-qaboard-l-bounces@lists.bnl.gov> on behalf of 李鸿灿 via STAR-QAboard-l <star-qaboard-l@lists.bnl.gov>
Sent: Friday, June 2, 2023 9:40 AM
To: ASHIK IKBAL <ashikhep@gmail.com>; star-qaboard-l <star-qaboard-l@lists.bnl.gov>
Subject: Re: [STAR-QAboard] Weekly QA Board Meeting - 02/June/2023 at noon 12 EDTHi Ashik and all,Please find my slides here: https://drupal.star.bnl.gov/STAR/blog/lhc/LFSUPC-QA-20230602
Best,Hongcan
------------------ Original ------------------From: "ASHIK IKBAL via STAR-QAboard-l"<star-qaboard-l@lists.bnl.gov>;Date: Thu, Jun 1, 2023 05:58 AMSubject: [STAR-QAboard] Weekly QA Board Meeting - 02/June/2023 at noon 12 EDTDear All,
We will have our weekly QA board meeting this Friday
(02/June/2023) at noon 12 EDT. Please let me know if you would like to present.
The agenda can be found here: https://drupal.star.bnl.gov/STAR/blog/ashik/QA-Board
Zoom connection details:
Join ZoomGov Meeting
https://bnl.zoomgov.com/j/1618907709?pwd=YitQYzFUdWFYU1ppTjYxTjRYbDdsZz09
Meeting ID: 161 890 7709
Passcode: 775538
Thanks,Ashik
--
Ashik Ikbal
Research AssociateKent State UniversityDepartment Of Physics
Kent, OH 44242, USA
Contact no.s: +1 631-504-7647 (USA)
+91 8250216234 (India)
Alternate e-mails: asheikh2@kent.edu
STAR-QAboard-l mailing list
STAR-QAboard-l@lists.bnl.gov
https://lists.bnl.gov/mailman/listinfo/star-qaboard-l
- 
            
            [STAR-QAboard] 回覆:  回覆:  Weekly QA Board Meeting - 02/June/2023 at noon	12 EDT,
            Han-sheng Li, 06/03/2023
- 
        Re: [STAR-QAboard] 回覆:  回覆:  Weekly QA Board Meeting - 02/June/2023 at noon	12 EDT,
        Tang, Aihong, 06/03/2023
- 
            
            Re: [STAR-QAboard] 回覆:  回覆:  Weekly QA Board Meeting - 02/June/2023 at noon	12 EDT,
            Tang, Aihong, 06/03/2023
- 
            
            [STAR-QAboard] 回覆: 回覆:  回覆:  Weekly QA Board Meeting - 02/June/2023 at noon	12 EDT,
            Han-sheng Li, 06/03/2023
- Re: [STAR-QAboard] 回覆: 回覆: 回覆: Weekly QA Board Meeting - 02/June/2023 at noon 12 EDT, Tang, Aihong, 06/03/2023
- 
            
            Re: [STAR-QAboard] 回覆: 回覆:  回覆:  Weekly QA Board Meeting - 02/June/2023 at noon	12 EDT,
            Tang, Aihong, 06/03/2023
- 
            
            Re: [STAR-QAboard] 回覆: 回覆:  回覆:  Weekly QA Board Meeting - 02/June/2023 at noon	12 EDT,
            EsumiShinIchi, 06/03/2023
- 
            
            Re: [STAR-QAboard] 回覆: 回覆:  回覆:  Weekly QA Board Meeting - 02/June/2023 at noon	12 EDT,
            Tang, Aihong, 06/03/2023
- Re: [STAR-QAboard] 回覆: 回覆: 回覆: Weekly QA Board Meeting - 02/June/2023 at noon 12 EDT, EsumiShinIchi, 06/03/2023
- Re: [STAR-QAboard] 回覆: 回覆: 回覆: Weekly QA Board Meeting - 02/June/2023 at noon 12 EDT, Tang, Aihong, 06/05/2023
- [STAR-QAboard] 回覆: 回覆: 回覆: 回覆: Weekly QA Board Meeting - 02/June/2023 at noon 12 EDT, Han-sheng Li, 06/05/2023
- [STAR-QAboard] 回覆: 回覆: 回覆: 回覆: Weekly QA Board Meeting - 02/June/2023 at noon 12 EDT, Han-sheng Li, 06/06/2023
 
 
- 
            
            Re: [STAR-QAboard] 回覆: 回覆:  回覆:  Weekly QA Board Meeting - 02/June/2023 at noon	12 EDT,
            Tang, Aihong, 06/03/2023
 
- 
            
            Re: [STAR-QAboard] 回覆: 回覆:  回覆:  Weekly QA Board Meeting - 02/June/2023 at noon	12 EDT,
            EsumiShinIchi, 06/03/2023
 
 
- 
            
            [STAR-QAboard] 回覆: 回覆:  回覆:  Weekly QA Board Meeting - 02/June/2023 at noon	12 EDT,
            Han-sheng Li, 06/03/2023
 
- 
            
            Re: [STAR-QAboard] 回覆:  回覆:  Weekly QA Board Meeting - 02/June/2023 at noon	12 EDT,
            Tang, Aihong, 06/03/2023
 
- 
        Re: [STAR-QAboard] 回覆:  回覆:  Weekly QA Board Meeting - 02/June/2023 at noon	12 EDT,
        Tang, Aihong, 06/03/2023
Archive powered by MHonArc 2.6.24.
