Looks like there was some discussion on the ops list just a bit ago too. I'm
forwarding here (TPC list) for the record.
Thanks,
-Gene
Begin forwarded message:
> From: David Tlustý <tlusty AT gmail.com>
> Subject: Re: [[Star-ops-l] ] Wiener/Iseg HV PS for Grid Leak Wall
> Date: September 24, 2024 at 2:21:17 PM EDT
> To: James H Thomas <jhthomas AT berkeley.edu>
> Cc: "Lebedev, Alexei" <alebedev AT bnl.gov>, STAR experiment operations
> mailing list <star-ops-l AT lists.bnl.gov>, videbaek <videbaek AT bnl.gov>
> Reply-To: star-ops-l AT lists.bnl.gov
>
> Hi Jim,
>
> That’s very useful. I tried that Emergency Off and then clearEvents command
> and it seems
> it fixed the problem. Shift crews are starting the new run (25268040) with
> the Grid Leak
> having all channels at reference values.
>
> David
>
>> On Sep 24, 2024, at 19:12:18, James H Thomas <jhthomas AT berkeley.edu> wrote:
>>
>> Dear David,
>>
>> I believe that we are using a Wiener/ISEG HV supply to provide 150 and 450
>> volts to the Grid Leak Wall. As I recall, these supplies have several
>> reset options. Do you think it would be possible to try some of the
>> other resets on the bad channels using CDEV commands?
>>
>> To be precise, the ISEG manual from 2013 is enclosed. Pages 26, 27, 28
>> describe some special commands for these modules. In particular, the ISEG
>> modules seem to have more than On, Off, and tripped modes. They also have
>> Emergency On and Emergency OFF, and ClearEvents. As I recall from prior
>> use of these modules on the SSD, they would trip and sometimes fall into
>> the Emergency Off mode and so would not respond to the normal clear trips
>> command. I think (but don't really remember accurately) that toggling
>> ClearEvents and resetEmergency OFF a few times for the bad channels was
>> required.
>>
>> If you can see clearly through the fog of these words, do you think it
>> would be worth a quick try to clear these channels before a more invasive
>> access this afternoon?
>>
>> Cheers,
>> Jim
>>
>
-