Hi
Tonko pointed out that current runs don't have this problem. So I went back at looked at the TPC offline QA
and see clearly that the issues on sector 24 itpc 4 RDO was present over many daya
first bad run 283030 last bad run 285046 since 286003 all good.

The most peculiar is that nothing special is reported in the elog from last bad to first good
=no pedestal, no power cycling, no reboot on DAQ.

-single beam run 285047
-blue beam dumped 22:43 10/11
-Run 25286001 - setRHICClock 00:15 10/12
- Run 25286002 - emc-check and 56*56 injected
- run 25286003 Au-Au good run

Was anything done to tpc's during this time? Can one check in log messages for tpc?


best Flemming



On 2024-10-11 17:03, videbaek wrote:
In the fast offline QA Lanny pointed out that it seems sector 4 has
fewer hits and this has been the case since Monday.
After today’s QA-meeting I got a root file with all the the
information with various information and looked at
1) charge per pad/row
2)cluster charge per pad/row
I focused on row 39 which has two region from rdo 3 and rdo 4. From
physics point of view they should have same #hits and charge per pad ,
in average.
The file is from run 25283030 in FastOffline,

The attached file has some additional comments , discussions, and plots

Thanks to Gene for providing the complete QA root file.

best Flemming

--