Thanks Gene,

I agree it seems unlikely there are event dependences.
As it shows up in the adc data too, it must be an issue in FEE/RDO
Flemming



On 2024-10-18 17:00, Van Buren, Gene wrote:
Hi, all

Flemming asked me to look for any observable event-by-event
differences, so I looked at 6000 events from run 25292005 (taken this
morning), file st_physics_25292005_raw_2000014. I then counted hits,
sampling from 4 patches:
Patches 1 & 2: rows (25-28) by 9 pads in RDO 3 and RDO 4
Patches 3 & 4: rows (37-40-) by 19 pads in RDO 3 and RDO 4
I was conservative in my patch sizes because there are lots of dead
areas in both of these RDOs. Unfortunately, there's no function in
StEvent that says what RDO a hit was in (probably because hits can
span RDOs for some padrows). What I see:

First: no significant spans of events without hits in any of the RDO
patches. It should be noted that the events in the DAQ files are
however not immediately consecutive.



========= Comparisons on patches 1 & 2 =========

Of all the 207607 hits found in these patches, 111835 (53.9%) are in
the RDO 3 patch and 95772 (46.1%) are in the RDO 4 patch.

Event-wise....

1058 of 6000 events (17.6%) only had hits in the RDO 3 patch, none in
the RDO 4 patch
...and of these 1058, 182 (17.2%) had two events in a row like this

691 of 6000 events (11.5%) only had hits in the RDO 4 patch, none in
the RDO 3 patch
...and of these 691, 74 (10.7%) had two events in a row like this

3983 of 6000 events (66.4%) had hits in both patches, and in those
events the number of hits in the two patches was roughly 52.4% in RDO
3 and 47.6% in RDO 4



========= Comparisons on patches 3 & 4 =========

Of all the 229001 hits found in these patches, 156177 (68.2%) are in
the RDO 3 patch and 72824 (31.4%) are in the RDO 4 patch.

Event-wise....

1015 of 6000 events (16.9%) only had hits in the RDO 3 patch, none in
the RDO 4 patch
...and of these 1015, 169 (16.7%) had two events in a row like this

619 of 6000 events (10.3%) only had hits in the RDO 4 patch, none in
the RDO 3 patch
...and of these 619, 61 (9.9%) had two events in a row like this

4119 of 6000 events (68.7%) had hits in both patches, and in those
events the number of hits in the two patches was roughly 51.5% in RDO
3 and 48.5% in RDO 4



========= Conclusions =========

1) The probabilities for seeing zero hits in RDO 4 patches is
significantly higher than for RDO 3! This appears to be a real
difference.

2) The probabilities for two events in a row without hits in any of
the patches is reasonably consistent with their single event
probabilities repeating randomly, i.e. there is no clear sign in this
data of RDO 4 being dead for spans of time.

3) While the strongest bias appears to be seeing no hits in a patch in
RDO 4 vs. RDO 3, there is also a weaker bias towards more hits in RDO
3 when there are hits in both RDOs.

4) My over-arching conclusion is that it doesn't appear to be that RDO
4 is dead for certain events, but rather that RDO 4's efficiency for
finding hits is simply lower overall, for all events. That would be
consistent with the cut-off for zero suppression for RDO 4 being wrong
by a small amount for the entire run.



========= Offline cluster finder =========

With that final conclusion in mind, I repeated the above analysis
running the offline cluster finder on 198 events from file
st_physics_adc_25292005_raw_1800005. The poor statistics don't allow
for precise conclusions, but the percentages are consistent with the
findings from the online-found clusters shown above: the biases are
still very much present.



Hope that helps, and I hope I didn't make any typos with all those numbers!
-Gene



On Oct 16, 2024, at 5:04 PM, videbaek <videbaek AT bnl.gov> wrote:

Hi tonko for reference the normal sector 24 offline looks like
the attached plot


On 2024-10-16 16:53, videbaek wrote:
Hi Lanny
I can confirm your observation. The last run in the fill of day 089
289061 and that is bad,
but for the next fill as you sya 290008.
It turns out to be very hard to see the issue in the online plots
attach are the cluster plot for sector 24 in run289061, so we really
don't have any tools to see this online, but only after fast offline.
Tonko is there anything in the logs?
So in both cases the problem was fixed after a beam dump, but not between runs.
The short runs done were setRHICclock and emccheck , can this do something?
Flemming
On 2024-10-16 16:08, Ray, Robert L wrote:
Hi Flemming:
Unfortunately, the low number of hits in sector 24, RDO 4 is
back again starting with run 25289050 for the rest of day 289.
In the few runs I have looked at in today's data (290), the
problem seems to have gone away.
Thanks,
Lanny
-------------------------
From: videbaek <videbaek AT bnl.gov>
Sent: Tuesday, October 15, 2024 1:18 PM
To: Star-tpc L <Star-tpc-l AT lists.bnl.gov>; Ray, Robert L
<ray AT physics.utexas.edu>
Subject: Re: sector 24 RDO 4 issues.
Hi
Tonko pointed out that current runs don't have this problem. So I went
back at looked at the TPC offline QA
and see clearly that the issues on sector 24 itpc 4 RDO was present
over
many daya
first bad run 283030 last bad run 285046 since 286003 all good.
The most peculiar is that nothing special is reported in the elog from
last bad to first good
=no pedestal, no power cycling, no reboot on DAQ.
-single beam run 285047
-blue beam dumped 22:43 10/11
-Run 25286001 - setRHICClock 00:15 10/12
- Run 25286002 - emc-check and 56*56 injected
- run 25286003 Au-Au good run
Was anything done to tpc's during this time? Can one check in log
messages for tpc?
best Flemming
On 2024-10-11 17:03, videbaek wrote:
In the fast offline QA Lanny pointed out that it seems sector 4 has
fewer hits and this has been the case since Monday.
After today’s QA-meeting I got a root file with all the the
information with various information and looked at
1) charge per pad/row
2)cluster charge per pad/row
I focused on row 39 which has two region from rdo 3 and rdo 4. From
physics point of view they should have same #hits and charge per pad
,
in average.
The file is from run 25283030 in FastOffline,
The attached file has some additional comments , discussions, and
plots
Thanks to Gene for providing the complete QA root file.
best Flemming
--
Flemming Videbaek
senior scientist, emeritus
videbaek @ bnl.gov
Brookhaven National Lab
Physics Department
Bldg 510D
Upton, NY 11973
phone: 631-344-4106
cell : 631-681-1596

--
Flemming Videbaek
senior scientist, emeritus
videbaek @ bnl.gov
Brookhaven National Lab
Physics Department
Bldg 510D
Upton, NY 11973

phone: 631-344-4106
cell : 631-681-1596<Screenshot 2024-10-16 at 17.03.17.png>

--
Flemming Videbaek
senior scientist, emeritus
videbaek @ bnl.gov
Brookhaven National Lab
Physics Department
Bldg 510D
Upton, NY 11973

phone: 631-344-4106
cell : 631-681-1596



Archive powered by MHonArc 2.6.24.

Top of Page