Dear sPHENIX

Here is the shift change meeting minutes on August 4.

***************
General (Stefan/Kin)
• It is clear now that we will not return to beam operations this run
• Nevertheless, the commissioning continues full force
• Switching to two-person shifts at midnight: SL and Detector Operator
Current Priority Items (PC)
• Make use of magnetic field, which we’ll have until Friday +/- 1 day:
take data with tracking detectors: TPC, MVTX, INTT, TPOT
Work Control Coordinators (Chris/Joel)
• No Updates from WCC’s
Plan of the Day (Stefan/PC/all–to be revisited at end of meeting)
• TPC magnetic field test by 5:30 to 6:00pm
• Then take cosmics data with coincidence trigger for INTT and TPOT

========================================================================

Evening (Ross)
• Magnet lost telemetry, Kin investigated.
• Kin: Lee Flader has needed to change the script/save the project in
order to solve the problem.
• Red caution light in IR will be on as long as magnet is 'ON', even
if at zero amps. Check with Shift Leader and monitors, then proceed.
• MVTX worked to add to big partition, no success yet. Ran on GTM1
successfully.
• one gas checklist was submitted with a mis-click, unable to delete.
• TPC spark detection development
Night (Joey)
• Confusion about magnet state; call to CAS and Kin clarified status
• ControlRoom1 computer froze and needed hard reboot
• DO computer instructions were mostly sufficient to recover GUIs
• Exception: MBD timing scope proved difficult due to a “firefox is
already running” error
• Had to kill existing firefox process which was using the same
profile
• Replaced ADC board in rack 1E1 crate 3 slot 3 - replacement board was
tested prior to swap and worked OK
Day (Tristan)
• HCal + EMCal + TPOT cosmics in the morning
• DAQ development by Jaebeom starting around 10:45
• IR swept for tools/magnetic material ahead of ramping up magnet
• Magnet ramping for TPC tests started ~1:00

========================================================================

Magnet (Kin)
• Magnet is being ramped up for the TPC studies ( stopping at 0 A, 500
A, and 2000 A and then full current).

MBD (Mickey, Lameck)
• Nothing new to update.
Trigger (Dan)
• Working fine
GTM/GL1 (Martin/Dan/John K)
• We have version 47. We held off on installing it last night. We are
continuing with 46 through the weekend in order not to interrupt current
testing (TPC, MVTX, DAQ). Also see the MVTX report - apparent behavior
difference between gtm(0) and gtm1. Now that I thought about this - I
disabled the fiducial tick now on the main gtm, maybe that had to do with it?
Worth trying again.
DAQ (Martin/John H)
• DAQ testing/debugging is continuing, kicking into even-higher gear next
week. After the imminent run23 end, we have a clear mission. Just for the
record, I’m using seb01 (Emcal 1W3) as my go-to system.
• I replaced the failed power supply on the intt3 Felix server PC.
MVTX (Cameron)
• We were able to run the MVTX in Global mode on GTM1. We weren’t able to
run it on the main GTM though. Jaebeom has been looking into it and we’re
trying the main GTM again now. Still didn’t work when running rcdaq as phnxrc
on operator1
TPC (Tom Hemmick, Jin, Takao, Evgeny, Charles, David, Bob, Ross, Nick, Luke,
John K.)
• Work on spark monitoring calibration finished yesterday 08/03
• Thanks to David from Debrecen!
• Using the spark monitoring system now
• Today 08/04:
• Decided to do Magnet Ramping test (thanks to Kin for coordinating):
• Using diffuse laser as “standard candle” to get qualitative sense
of gain
• Raise one HV module to 4650 V, all others 4200 V
• 0 A - done - 2nA response

PNG image


• 500 A - done - 2-3 nA response (i.e. stable relative to zero field)

PNG image


• 2000 A - done 3-4nA response (possibly higher GEM gain with higher
field)

PNG image


• 4620 A (full field) - final measurement later today
• Also testing spark monitor with module raised up to 4650 V
• Saw one spark in U709 (red trace)

PNG image


• Weekend: 08/05 & 08/06
• Full field running primarily for gathering stats for spark monitor
• (David + Jin + Ross)
• Secondarily, can take cosmics runs
• Spark monitor takes priority, otherwise at cost of ~1 GEM strip
per hour at operational HV
• Once spark protection works, start cosmic run with HCal cosmic
trigger
• Need to move back to GTM0 from GTM1 for cosmic run
• Front end configured, under 24*7 monitoring run when not
used in big partition
• Normal cosmic big partition procedure to bring in TPC
• Next week,
• Full field cosmic running
• Best be done with all tracking detectors for global tracker
alignment (on top of studying TPC response alone)
• Possibly mix in some diffuse laser shots, depending on laser
debug status
• B-Field scan may be requested again, after the understanding of
diffuse laser with full GEM gain: aim is to observe changes in the laser
pattern in data with B-field

HCal (Silas)
• Replacement of pre-amp board to fix hot tower
• (JaeBeom) Still hot.. Run of this morning 25134 (cosmics)

PNG image


• New HCal coincidence cosmics trigger: should be utilized when taking
cosmics with tracking detectors
• Otherwise use standard cosmics trigger (any single tower)
• In the past we were anxious to collect enough statistics and therefore
not included EMCal
• Now that there’s no more beam, we can include EMCal
EMCal (Sean)
• Nothing new to report
TPOT (Hugo)
• Nothing to report. Smooth cosmics data taking
INTT (Genki/Rachid)
• The power supply module in INTT3 was replaced by Martin. Thank you so
much!
• We can try to measure cosmic ray this evening.
sEPD(Rosi)
• Checking out south EPD electronics - 6 out of 6 interface boards have
signal
• Install North box after magnet off (? tbd)
Gas/Cooling ()
ZDC (Ejiro)
• With no beam, immediate plans are:
• Finish ZDC calibration using MBD mip information
• Correlate MBD timing with ZDC timing (offline analysis)
• Nothing new on the SMD
Background Counters ()

Online Monitoring (Chris)

*************end of minutes ******
Best,

Maya
###################