Skip to Content.
Sympa Menu

sphenix-magnet-l - [Sphenix-magnet-l] pressure testing done for the LN2 line around 1008 by the Cryo group team

sphenix-magnet-l AT lists.bnl.gov

Subject: sPHENIX discussion of the superconducting solenoid

List archive

Chronological Thread  
  • From: "Yip, Kin" <kinyip AT bnl.gov>
  • To: "sphenix-magnet-l AT lists.bnl.gov" <sphenix-magnet-l AT lists.bnl.gov>
  • Subject: [Sphenix-magnet-l] pressure testing done for the LN2 line around 1008 by the Cryo group team
  • Date: Thu, 12 May 2022 19:51:35 +0000

What you can read below  consist of a series of email communication, ending with pressure testing done for the welds / joints of the LN2 line around 1008.

 

I’ve include all the drawings that people have included in the conversations.

 

{

31050056.pdf, 105-0508-050-4-2.pdf and PhenixGasPad_R2.pdf  are from Paul Orfin.

205-0800-0101 LN2.pdf is from Richard Ruggiero.

IMG_8557.jpeg is from John Haggerty.

NitrogenGas.pdf and 1008LN2 Dewar.pdf is from Robert P. Pisani.

}

 

Kin

 

From: Orfin, Paul <porfin AT bnl.gov>
Sent: Thursday, May 12, 2022 10:28 AM
To: Pisani, Robert <pisani AT bnl.gov>
Cc: Haggerty, John <haggerty AT bnl.gov>; Yip, Kin <kinyip AT bnl.gov>; Biggs, John C <biggs AT bnl.gov>
Subject: RE: Print - 205-0800-0101

 

All,

 

Pressure testing of the welds and solder joints were completed this morning with no issue.  Vaporizer line was depressurized at the end of the test.

 

Thank you,

Paul Orfin
Cryogenic Systems, Collider-Accelerator Department
Brookhaven National Laboratory
Telephone: (631) 344-2899

 

From: Robert P. Pisani <pisani AT bnl.gov>
Sent: Thursday, May 12, 2022 9:23 AM
To: Orfin, Paul <porfin AT bnl.gov>
Cc: Haggerty, John <haggerty AT bnl.gov>; Yip, Kin <kinyip AT bnl.gov>; Biggs, John C <biggs AT bnl.gov>; Hamblen, Peter <phamblen AT bnl.gov>
Subject: Re: Print - 205-0800-0101

 

Here is a quick layout of how it's isolated.  

 

 

On Thu, May 12, 2022 at 8:16 AM Robert P. Pisani <pisani AT bnl.gov> wrote:

Line is isolated on the gas pad and in gas house.  You can test whenever ready.    The main n2 gas line that we use has a 350psi relief valve on it located on the south side of the tank.    So that line can go higher if needed.    The new line parts that were installed near the tank has a 300psi relief.

 

Ill send around a few pict when i get back ro the office

 

On Wed, May 11, 2022, 3:44 PM Robert P. Pisani <pisani AT bnl.gov> wrote:

Im off site today.  I would like to confirm the outside valve at the mixing house us closed.  It should be, but i havent looked at it in a while   i can check tomorrow first thing and let you know.  

 

Rob

 

On Wed, May 11, 2022, 3:33 PM Orfin, Paul <porfin AT bnl.gov> wrote:

Rob,

 

I will make the pressure test to 150 PSI for that one solder joint and do a 10 minute hold per ASME.  We will also do a helium mass spectrometer test since we will have all the equipment out there anyway for the weld portion. 

 

If the system can not hold pressure and my solder joint is good as determined by helium mass spectrometer, I will have to defer to your team to make any repairs down stream of our alteration.  The balance of that system is not part of our scope nor is it adequately documented for us to troubleshoot.  This was the agreement when we first tapped into the system to provide the 80K hold to the magnet since Physics owns all that equipment. 

 

I want to make this pressure and scope limitation clear before I apply any pressure to your system in case there are components downstream that are not rated for 150PSI.  If you are not comfortable with this please let me now so you can take more time to examine the system and the down stream components to ensure isolation or pressure ratings.  I need to be clear on this so we are not responsible for damage at the 150 PSI level or any down stream safety concerns once pressure is applied. 

 

By this email and your previous replies, you are certifying that it is safe to pressurize the system to 150 PIS and there are no down stream concerns.

 

We plan to start the test tomorrow morning.  If you need more time to check items let me know.  We can talk if that is preferred.

 

Thank you,

Paul Orfin
Cryogenic Systems, Collider-Accelerator Department
Brookhaven National Laboratory
Telephone: (631) 344-2899

 

From: Robert P. Pisani <pisani AT bnl.gov>
Sent: Wednesday, May 11, 2022 3:07 PM
To: Orfin, Paul <porfin AT bnl.gov>
Cc: Haggerty, John <haggerty AT bnl.gov>; Yip, Kin <kinyip AT bnl.gov>; Biggs, John C <biggs AT bnl.gov>
Subject: Re: Print - 205-0800-0101

 

We usually test to 50% operating pressure (100psi for us) when testing a new joint. 150psi would be a better test point.

 

   In theory, that joint can see 200psi from the tank if someone shuts off the valve at the evaporator. We will not operate there.  

 

 

 

 

On Wed, May 11, 2022, 2:26 PM Orfin, Paul <porfin AT bnl.gov> wrote:

We will be isolating the tank and the reliefs for the test.  I don’t think we should go to 450PSI for the vaporizer lines.

 

Is 100 PSI adequate for you? 

 

We are only test the first solder joint from our connection point, not the balance of this existing and presumed certified system. 

 

Thank you,

Paul Orfin
Cryogenic Systems, Collider-Accelerator Department
Brookhaven National Laboratory
Telephone: (631) 344-2899

 

From: Robert P. Pisani <pisani AT bnl.gov>
Sent: Wednesday, May 11, 2022 12:24 PM
To: Orfin, Paul <porfin AT bnl.gov>
Cc: Haggerty, John <haggerty AT bnl.gov>; Yip, Kin <kinyip AT bnl.gov>; Biggs, John C <biggs AT bnl.gov>
Subject: Re: Print - 205-0800-0101

 

The tank's max is 216psi.   There should be pressure reliefs on it, but I would have to go to the pad to see what they are. We usually operate it at 90spi in the gas house so the Dry gas is probably 100psi or slightly higher at the evaporator. 

 

rob

 

On Wed, May 11, 2022 at 12:15 PM Orfin, Paul <porfin AT bnl.gov> wrote:

Thanks. I'm getting out there too.  If the label is right we have 450 psi max.  I'm not sure the the line connected to it are rated for or if there are any pressure relief valves on that's system.

 

Do you know the normal operating pressure of the vaporizer line?

 

If not, I'll just go with 100psi leak check to call the solder joint passing for ASME code.

 

Thank you,

Paul Orfin
Cryogenic Systems, Collider-Accelerator Department
Brookhaven National Laboratory
Building 1005S, MS 1005S, Room 220
Upton, NY 11973-5000
Telephone: (631) 344-2899
Fax: (631) 344-3329
E-mail: 
porfin AT bnl.gov


From: John Haggerty <haggerty AT bnl.gov>
Sent: Wednesday, May 11, 2022 12:10:12 PM
To: Pisani, Robert <pisani AT bnl.gov>
Cc: Ruggiero, Richard <ruggiero AT bnl.gov>; Orfin, Paul <porfin AT bnl.gov>; Hamblen, Peter <phamblen AT bnl.gov>; Yip, Kin <kinyip AT bnl.gov>; Biggs, John C <biggs AT bnl.gov>
Subject: Re: Print - 205-0800-0101

 

Paul and Rob,

Yes, I walked over there.  This is the tall vaporizer east of the LN2
dewar.  I didn't know there were two of them.

On 2022-05-11 11:59, Robert P.  Pisani wrote:
> Our PO system only goes back to 2009, so I couldnt get any info from
> there.  But I attached the paperwork we got when the unit was
> delivered and installed. Figured I circulate it to a few more people.
>  Paul, I sent this to you back in 2019.   The tank working pressure is
> 216psi and was tested to 330psi (tank).   I could not find the
> vaporizer on it.
>
> Rob
>
> On Wed, May 11, 2022 at 11:37 AM Robert P. Pisani <pisani AT bnl.gov>
> wrote:
>
>> John, is this the big one next to the LN2 tank?  there are 2.  the
>> smaller one is for the smaller LN2 buggies.
>>
>> On Wed, May 11, 2022 at 11:33 AM John Haggerty <haggerty AT bnl.gov>
>> wrote:
>>
>>> Rob,
>>>
>>> Is this what you want?
>>>
>>> On 2022-05-11 11:18, Robert P.  Pisani wrote:
>>>> The  vaporizer should have a plate on it =stating its ratings.
>>> I am
>>>> offsite, but copied Carter Biggs.  Carter was the one who
>>> ordered this
>>>> equipment.   I am going to dig up the PO and see if the spec is
>>> listed
>>>> on the order.
>>>>
>>>> For the tubing, we need the OD of the line and material and we
>>> can
>>>> look up the spec.  If its flex line, it should have a tag on one
>>> end.
>>>>
>>>>
>>>> Rob
>>>>
>>>> On Wed, May 11, 2022 at 10:13 AM Ruggiero, Richard
>>> <ruggiero AT bnl.gov>
>>>> wrote:
>>>>
>>>>> Hi Paul,
>>>>>
>>>>> I’ve attached the drawing we have in Vault. Maybe
>>>>> Rob can confirm it’s the latest? It’s still marked checking
>>>>> print.
>>>>>
>>>>> RIch
>>>>>
>>>>> From: Orfin, Paul <porfin AT bnl.gov>
>>>>> Sent: Wednesday, May 11, 2022 9:08 AM
>>>>> To: Ruggiero, Richard <ruggiero AT bnl.gov>; Hamblen, Peter
>>>>> <phamblen AT bnl.gov>; Pisani, Robert <pisani AT bnl.gov>; Haggerty,
>>> John
>>>>> <haggerty AT bnl.gov>; Yip, Kin <kinyip AT bnl.gov>
>>>>> Subject: Print - 205-0800-0101
>>>>>
>>>>> All,
>>>>>
>>>>> Can someone please send me the final version of the
>>> 205-0800-0101
>>>>> drawing for the LN2 Dewar at 1008?
>>>>>
>>>>> We are looking to do the pressure test and want to confirm the
>>>>> valving after the ambient vaporizer.  Does physics have a
>>> drawing
>>>>> that documents that vaporizer?
>>>>>
>>>>> 31050056 is our official drawing for the LN2 supply.  The other
>>> 2
>>>>> drawings I got from Physics awhile back.  Only the gas pad
>>> print
>>>>> shows the ambient vaporizer. Does anyone one know the pressure
>>>>> rating of the vaporizer or the lines attached to it to pressure
>>> test
>>>>> the one new solder joint to it?
>>>>>
>>>>> Thank you,
>>>>>
>>>>> Paul Orfin
>>>>> Cryogenic Systems, Collider-Accelerator Department
>>>>> Brookhaven National Laboratory
>>>>> Building 1005S, MS 1005S, Room 217
>>>>> Upton, NY 11973-5000
>>>>> Telephone: (631) 344-2899
>>>>> Fax: (631) 344-3329
>>>>> E-mail: porfin AT bnl.gov
>>>>
>>>> --
>>>>
>>>> \~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~
>>>> Robert Pisani
>>>> Senior Physics Associate
>>>> Brookhaven National Laboratory
>>>> Bldg 510C
>>>> Upton, NY 11973
>>>> Phone:631-344-5301  Room 2-206
>>>> \~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~
>>>
>>> ---
>>> John Haggerty
>>> haggerty AT bnl.gov
>>> cell: 631 741 3358
>>
>> --
>>
>> \~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~
>> Robert Pisani
>> Senior Physics Associate
>> Brookhaven National Laboratory
>> Bldg 510C
>> Upton, NY 11973
>> Phone:631-344-5301  Room 2-206
>> \~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~
>
> --
>
> \~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~
> Robert Pisani
> Senior Physics Associate
> Brookhaven National Laboratory
> Bldg 510C
> Upton, NY 11973
> Phone:631-344-5301  Room 2-206
> \~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~

---
John Haggerty
haggerty AT bnl.gov
cell: 631 741 3358


 

--

\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~

Robert Pisani
Senior Physics Associate 

Brookhaven National Laboratory
Bldg 510C

Upton, NY 11973

Phone:631-344-5301  Room 2-206
\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~


 

--

\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~

Robert Pisani
Senior Physics Associate 

Brookhaven National Laboratory
Bldg 510C

Upton, NY 11973

Phone:631-344-5301  Room 2-206
\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~\~/~

Attachment: NitrogenGas.pdf
Description: NitrogenGas.pdf

Attachment: 1008LN2 Dewar.pdf
Description: 1008LN2 Dewar.pdf

Attachment: IMG_8557.jpeg
Description: IMG_8557.jpeg

Attachment: 205-0800-0101 LN2.pdf
Description: 205-0800-0101 LN2.pdf

Attachment: 31050056.pdf
Description: 31050056.pdf

Attachment: 105-0508-050-4-2.pdf
Description: 105-0508-050-4-2.pdf

Attachment: PhenixGasPad_R2.pdf
Description: PhenixGasPad_R2.pdf



  • [Sphenix-magnet-l] pressure testing done for the LN2 line around 1008 by the Cryo group team, Yip, Kin, 05/12/2022

Archive powered by MHonArc 2.6.24.

Top of Page