Minutes shift change 2023-05-03

general
—————
- Begin-Run party Friday noon. Come hungry.
- updated expert list
- any PLC work:  go through Joel, do not go directly to Lee
- still waiting for Alex Tulio for IRR approval
- Cindy forwarded documentation to Mike Lancy, which is next step
- magnet testing should finish late today

evening:  Xuan
- quiet shift
- INTT 2W1 and 2W3 work
- checked flow rates MVTX 1.5 l/m
- durachill notifications 
  - Lee is working on this
- Eric:  EMCal bias voltage on for stability test 

night:  Hanpu
- water leak in INTT chiller
  - with instructions from Joel and Rob turned off machine
  - caused by leak in flow meter
- LV was off, bias was off
  - turned off INTT racks
- 3 audio alarms from EMCAl chillers, old problem, no reason

day:  Sookhyun
- Rachid:  INTT flow meter fixed
  - tested interlock system, LV worked, bias didn’t 
- EMCal 1E1 was off by Cameron for work in MVTX right above
- chiller alarm now tells you where it comes from (thanks to Lee)

MBD (Abdul)
- timed in laser (FPGA card)
- next: fine delay scan to get signal into exact time window 
- then going to replace laser splitter

HCal (Hanpu)
- took cosmic run

EMcal (Sean)
- detector development work (Anthony, Tim): bias

TPOT (Hugo)
- waiting for vGTM (John K:  mapping problem in GTM)

TPC (Jin)
- new firmware and new driver version installed
- TPC pedestal taking, taking data now

INTT
- cooling issue has been fixed (the leaking flow meter)
- data taken with full bias
- interlock system tested and working (L1 interlock system for bias)
- data taken with RCDAQ yesterday
- JH: what fraction of the INTT is on? Rachid: all of it

MVTX 
- PLC system work, no spare for this one, getting it repaired
- broken connection in the cooling on DAQ-1, fixed
- persistent alarm, persistent false alarm (power)
- access request for Friday, not sure if this will be ready (scaffold etc)

DAQ 
- MLP working on a new version of the DCM2 readout (warning: this
will be disruptive! There is no way to make a gentle transition, so it
will take some hand-holding for emcal (I'm using HCal as guinea pig)


---------------------------------------------------------------------------------
Stefan Bathe
Professor of Physics
Baruch College, CUNY
and RIKEN Visiting Scientist

Baruch:                                     BNL:
17 Lexington Ave                      Bldg. 510
office 940                                  office 2-229
phone 646-660-6272                phone 631-344-8490
—————————————————————————————————————————





  • [Sphenix-run-l] Minutes shift change 2023-05-03, Stefan Bathe, 05/03/2023

Archive powered by MHonArc 2.6.24.

Top of Page